Glucometer accuracy and implications for clinical studies
Critical Care volume 16, Article number: P173 (2012)
Introduction
Elucidating links between glycemic control and clinical outcome requires reliable discrimination between groups with different target blood glucose (BG) cut-offs. Point-of-care glucometers are commonly used, but lower accuracy means BG errors will impact classification and thus outcome analyses. This study reanalyses a BG control trial with an error model of a typical glucometer to assess the impact of sensor errors on interpretation of trial results.
Methods
BG profiles from 301 patients (stay >24 hours) from the SPRINT trial with BG measurements (n = 25,000) using the Arkray SuperGlucocard II GT-1630. A model of sensor bias and variance (CV 2.7 to 3.5%, regression: y = 3.92 + 0.97x) was used to estimate possible 'true' BG profiles from measured BG and repeated 100 times for each patient. The defined cut-off for 'good' control for a patient was ≥70% of BG in 72 to 126 mg/dl (cTIB ≥0.7), and 'poor' as <70% (cTIB <0.7), based on original observed clinical BG. The number of true BG profiles that resulted in misclassification between 'good' and 'poor' control for a patient was recorded over all Monte-Carlo runs. The maximum change in true and observed BG mean and standard deviation were used to evaluate potential worst-case scenarios.
Results
Good control was clinically measured in 76% of patients (24% with cTIB <0.7). Of these, 83% of 'good' and 64% of 'poor' control would never be misclassified over all 100 runs due to sensor error. A total of 91% (good) and 87.5% (poor) could be misclassified 10% of the time. Patients with cTIB near 0.7 were more likely to be misclassified when accounting for glucometer error. Hence, a deadband around the cut-off would reduce this misclassification. If 'good' cut-off was cTIB ≥0.5 (95% of clinical patients) then 100% correct classification was 97% for good control patients, but fell to 40% of poor control patients. The median largest difference in observed and true mean BG across patients was -54 mg/dl (90th percentile: -21 mg/dl) and the standard deviation was 3.2 mg/dl (90th percentile: 1.8 mg/dl).
Conclusion
Glucometers can distinguish between patients that received good and poor BG control but risk of misclassification rises for patients nearer cut-offs. Reliable classification to associate with outcomes relies on the control protocol and cut-off choice to achieve sufficient separation between groups so that device errors do not result in significant misclassification confounding the results. A deadband around cut-off values to eliminate patients at high risk of misclassification may be required.
Author information
Authors and Affiliations
Rights and permissions
This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article's Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article's Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.
About this article
Cite this article
Le Compte, A., Pretty, C., Shaw, G. et al. Glucometer accuracy and implications for clinical studies. Crit Care 16 (Suppl 1), P173 (2012). https://doi.org/10.1186/cc10780
Published:
DOI: https://doi.org/10.1186/cc10780