Ref. (year) | Number of patients and category | Blood glucose target (mmol/l) | Method | Used BG meter | Frequency of measurements | Hypoglycaemia (mmol/l) | Results (mmol/l) |
---|---|---|---|---|---|---|---|
[1] | 765 Thoracosurgical ICU patients | 4.4–6.1 | Start at BG >6.1 mmol/l. Insulin ± 0.1–2 IU/hour depending on last two BG values. Glucose infusion or feeding | ICU based ABL700® bloodgas/BG analyzer. Arterial blood samples | 1–4 hours | 5.1% of patients <2.2 | Mean morning BG 5.7 ± 1.1 |
[4] (2004) | 27 Mixed ICU patients | 4–7 | Bath IIP: insulin ± 0.5–4 IU/hour, depending on last two BG values | Accu Check® Advantage 2. Mostly arterial samples | 1–2 hours | Three BG values <2.2 | Median BG 6.6 |
[2] (2004) | 800 Mixed ICU patients | <7.7 | Insulin sc. If two BG values >11.1 mmol/l: insulin iv, sliding scale. Glucose infusion/feeding | Finger stick samples or plasma BG (Vitros® lab analyzer) | Every 3 hours if sc, hourly if iv | 0.34% of patients <2.2 | Mean BG 7.2 70% of BG <7.7 |
[5] (2001) | 20 Critically ill diabetic patients | 6.7 | Insulin iv between -1.5 and +1.5 IU/hours depending on last two BG values. Glucose-potassium infusions | OneTouch® II. Capillary samples | 4 hours | No BG <2.5 | Mean BG 7.8 ± 0.2 |
[6] (2004) | 118 Cardiothoracic ICU patients | 5.5–7.7 | Yale IIP. Insulin ± 0.5–10 IU, depending on last two BG values and infusion rate | OneTouch® Surestep Flex | 1 hour | 0.2% of BG <3.3 | 73% between 4.4 and 7.7 |
[7] (2004) | 52 Medical ICU patients | 5.5–7.7 | Yale IIP. Insulin ± 0.5–10 IU, depending on last two BG values and infusion rate | OneTouch® Surestep Flex | 1–4 hours | 0.3% of BG <3.3 | 66% between 4.4 and 7.7 |
[3] (2004) | 50 Mixed ICU patients | 4.5–6.1 | Insulin ± 0.5–2 IU/hour, depending on last two BG values. Dextrose infusion or feeding | Accu Check® Inform. Capillary samples | 1–2 hours | 4% of patients <2.2 | 11.5 ± 3.7 hours/day between 4.5 and 6.1 |
[8] (2004) | 168 Cardiothoracic ICU patients | 4.4–8.3 | Insulin 1–16 IU/hour + bolus, sliding scale, depending on last BG value | Accu Check® Inform. Venous samples | 1–4 hours | 7.1% of BG <2.2 | 61% of BG between 4.4 and 8.3 |
[9] (2003) | 17 Diabetic patients, acute medical diseases | 6–7 | GIK + bolus, insulin 1–4 IU/h, dynamic scale, depending on last BG value | Hemocue® meter. Capillary samples | 1 hour | Four times | Mean BG 10.1 |
[10] (2002) | 37 Postsurgical NIDDM patients |  | 19 patients sc, 18 patients iv, sliding scale, 5% glucose infusion | Glucometer®. Capillary samples | 4 hours | 5.6% of patients in iv group | sc mean: 7.2 ± 1.2 iv mean: 7.3± 1.1 |
[11] (2004) | 72 Cardiothoracic diabetic patients | 6.9–11.1 | GIK protocol. Continuous GIK infusion + insulin bolus if BG >15 mmol/l | ? | 1 hour |  | Mean BG 7.7 ± 0.2 |
[31] (1996) | 60 Surgical NIDDM patients | 3.3–11.1 | Insulin bolus if BG >11.1 mmol/l. Group 1: saline. Group 2: glucose-insulin. Group 3: bolus every 2 hours | Capillary samples | 15 min | No | Mean BG <9 within all groups |
[12] (2002) | 29 Diabetic patients, cardiac surgery, 5 days | 6.7–11.1 | Start if BG >7.8 mol/l. Sliding scale | Venous and capillary samples | Six measurements per day | 0.2% of BG <3.8 | Mean BG 9.5 |
[13] (1997) | 595 Diabetic patients, cardiac surgery | <11.1 l | Portland protocol: insulin depending on last two BG values and insulin infusion rate | Glucometer | 1–2 hourly |  | Mean BG 9.7 |
[14] (1987) | 24 Diabetic patients after surgery | 6.7–10 | Insulin ± 0.5 IU/hour depending on BG. Bolus if BG >13.3 mol/l. 5% dextrose infusion | Accu Check + strips | 2 hourly | 1.4% of measurements | Mean BG between 6.7 and 10 |
[15] (1988) | 30 Diabetic patients, perioperative | 5–10 | Group 1: iv, every 4 hours ± 0.5 IU/hour. Group 2: sc, every 4 hours ± 2 IU/4 hours. Glucose-potassium infusion | Glucometer | Hourly during surgery, 4 hourly after surgery | 0.6% of measurements <2.8 in iv group | 67% of iv group between 5.0 and 10; 40% in sc group |
[17] (2002) | 188 Patients, during cardiac surgery | 4.4–6.6 | Start (2 IU/hour) with BG >6.6 mmol/l. Double infusion rate until BG <6.6 mol/l. | ? | 20 min | 12% of patients with BG <3.8 | In 23% of patients BG <8.3 |
[18] (1994) | 77 Diabetic patients, surgery | 6.7–10 | Insulin ± 0.5–1.0 U/hr depending on BG | Reflolux S (+strips) and Glucose hexokinase (lab) | 4 hourly, hourly during surgery | Two patients | 62% of patients BG between 3.5 and 15.0 |
[19] (2000) | 24 Type 2 diabetic patients, acute myocardial infarction | 8.3–11.0 | Insulin ± 1–2 IU/h, depending on BG range. | Venous samples. Automatic analyzer (lab) | 30 min to 2 hours |  | Mean BG 6.9 ± 0.8 |
[20] (2002) | 25 Diabetic patients, acute coronary syndromes | 6.6–8.2 | Insulin change by -1 to +3 IU/hour, depending on BG | Beckman® glucose analyzer II | 1–3 hourly | Four patients with mild hypoglycaemia | Mean 7.2 ± 1.7 |
[21] (1999) | 25 Patients, acute stroke (during 24 hours) | 4–7 | GIK: 500 ml glucose 10% + 16 U insulin + 20 mmol KCl; 100 ml/hour. ± 4 U insulin/infusate, depending on BG | BM Glycemic strips | 2 hourly | One patient | Mean BG of 68% of patients <7 |
[22] (1992) | 29 Diabetic patients, acute myocardial infarction | 4–8, to reach within 4 hours | Sliding scale, more insulin with left ventricular failure and bodyweight >120 % of ideal | Capillary samples | 1–4 hourly | 1.2% of BG <3 | Mean BG 8.2 ± 1.3 |
[23] (1994) | 158 Diabetic patients, acute myocardial infarction | 7–10 | >15 mmol/l; bolus iv. Insulin ± 0.5–1 IU/hour depending on BG. Glucose infusion | Venous samples. Reflolux II | 1–2 hourly | 17.7% of patients with BG <3.0 mmol/l | Mean BG 9.2 ± 2.9 after 24 hours |
[24] (1991) | 35 Diabetic patients, acute myocardial infarction | 4–8 | Sliding scale. Dextrose 5% infusion | Glucometer II®. Capillary samples | 2–4 hourly |  | Mean BG 10.3 ± 2.1 |