Table 4 Comparison of PEEP titration in supine versus prone position, for patients with coronavirus disease-19-related acute respiratory distress syndrome

From: Electrical impedance tomography to titrate positive end-expiratory pressure in COVID-19 acute respiratory distress syndrome

 

C-ARDS in supine position (n = 14)

C-ARDS in prone position (n = 14)

p value

PaO2/FiO2 (mmHg)

107 [93, 136]

179 [127, 190]

< 0.01

EIT-PEEP titration

 Best PEEPa (cmH2O)

12 [9, 12]

12 [10, 15]

0.59

 AUC for compliance

344 [316, 439]

455 [332, 494]

0.04

 Hyperdistension at PEEP 18 cm H2O (%)

24 [17, 30]

17 [15, 41]

0.63

 AUC for hyperdistension

113 [68, 134]

80 [40, 194]

0.82

 Collapse at PEEP 6 cm H2O (%)

23 [19, 37]

35 [23, 42]

0.07

 AUC for collapse

86 [58, 139]

173 [135, 261]

< 0.01

  1. aDefined using EIT as the smallest sum of hyperdistension and collapse. Continuous variables are expressed as median [interquartile range]
  2. PEEP positive end-expiratory pressure, EIT electrical impedance tomography, AUC area under the curve in supine and prone position (see Fig. 4)