Table 1 Characteristics of the study population of N = 312 mechanically ventilated children. Data are presented as median (25–75 interquartile range) or percentage (%) of total. Ventilator-free days (VFD) through day 28 were defined as the number of days within 28 days that a patient was alive and free of mechanical ventilation [16]. Patients were assigned 0 VFD if they remained intubated or died prior to day 28 without remaining extubated for more than 24 h. *Denotes p < 0.005 (Kruskal Wallis test). PRISM Pediatric Risk of Mortality; OI oxygenation index

From: Energy transmission in mechanically ventilated children: a translational study

 

Patient stratification by Pediatric Mechanical Ventilation Consensus Conference category

Normal respiratory system mechanics

Restrictive respiratory system mechanics (i.e., reduced compliance)

Mixed respiratory system mechanics (i.e., reduced compliance and increased resistance)

Cardiac (combination of normal and altered respiratory system mechanics)

N = 84

N = 69

N = 78

N = 81

Baseline patient characteristics

 Age (months) *

49.3 (9.4–122.0)

23.1 (8.0–46.0)

2.0 (1.2–8.5)

2.2 (0.5–6.7)

  ≤ 12 months (%)

29.78

39.1

84.6

84.0

  13–60 months (%)

20.2

34.8

11.5

6.2

  > 60 months (%)

50.0

26.1

3.8

9.9

 Male gender (%)

61.9

58.0

62.8

63.0

 Weight (kg) *

16.5 (8.0–29.5)

12.4 (8.5–17.5)

5.0 (4.0–6.7)

4.2 (3.6–6.9)

 PRISM III 24 h score

8.5 (5.0–14.0)

10.0 (6.0–15.0)

11.0 (8.0–15.0)

12.0 (7.0–15.5)

 PARDS (%)

0.0

5.8

24.4

1.2

 Cstat (cmH2O/L/kg) *

0.59 (0.43–0.78)

0.55 (0.39–0.73)

0.34 (0.28–0.44)

0.5 (0.32–0.6)

Baseline ventilator settings

 PIP (cmH2O) *

18 (16–22)

20 (17–23)

27 (24–29)

20 (18–23)

 Pplat (cmH2O) *

16 (14–19)

17 (15–21)

24 (21–26)

18 (16–21)

 PEEP (cmH2O) *

5 (5–6)

5 (5–6)

6 (5–7)

5 (5–6)

 Vt-exp (mL/kg) *

7.6 (6.6–8.6)

7.6 (6.5–8.5)

6.6 (6.1–7.7)

7.5 (6.8–8.3)

 Inspiratory time (s) *

0.75 (0.62–0.9)

0.7 (0.6–0.75)

0.55 (0.5–0.6)

0.55 (0.5–0.6)

 Mandatory breath rate (/min) *

22 (17–30)

23 (20–30)

40 (32–40)

35 (29–40)

Outcome data

 Ventilation time (h) *

88.3 (42.5–163.4)

69.8 (25.3–179.3)

107.5 (78.0–152.2)

91.5 (32.6–163.1)

 VFD day 28

24 (20–26)

24 (19–26)

23 (21–24)

24 (21–26)

PICU mortality (%)

6.0

4.3

2.6

3.7