From: A narrative review on trans-nasal pulmonary aerosol delivery
Techniques for aerosol delivery with HFNC | Recommendations | Evidence resource |
---|---|---|
Aerosol generator | VMN is more efficient than jet nebulizer when placed in-line with HFNC | |
Discontinue HFNC treatment to deliver conventional aerosol treatment | Not recommended. | In vitro pediatric [33, 37, 38] In vitro adult [32]. |
Use conventional aerosol device with concurrent HFNC | Not recommended. | Adult in vitro [32] Pediatric in vitro [34] |
Nebulizer placement | VMN should be placed at the inlet of humidifier, except when gas flow is extremely low, such as ≤ 0.25 L/kg/min for infants | |
Gas flow setting during trans-nasal aerosol delivery | If possible, titrate HFNC gas flow below the patient’s inspiratory flow | Adult in vivo [42] |
Open mouth breathing during trans-nasal aerosol delivery | When gas flow exceeds patient inspiratory flow, open mouth breathing reduces inhaled dose; when gas flow is below the patient’s inspiratory flow, open mouth breathing could generate higher inhaled dose. | Adult in vitro [26] Pediatric in vitro [37] |
Use heliox to deliver aerosol via HFNC | Might be considered for pediatric patients | Adult in vitro study [27] Pediatric in vitro [23] |
Use dry gas to deliver aerosol via HFNC | Not recommended | adult in vivo [42] |
Using frequent unit doses or infusion pump to deliver continuous albuterol for asthma exacerbation | If possible, use unit dose to deliver albuterol and decrease gas flow during nebulization; return flow to original setting when nebulization is completed. Titrate FIO2 to maintain SpO2 during the periods of flow reduction. If infusion pump has to be used, relative low gas flow and a higher nominal dose could be considered. | Pediatric in vitro [37] |
Stable COPD | Standard dose (2.5 mg) of albuterol is sufficient to elicit bronchodilation responses with HFNC gas flow set at 15–20 L/min. | |
COPD exacerbation | Standard dose (2.5 mg) of albuterol as a starting dose with HFNC flow set at 20–30 L/min is recommended during trans-nasal aerosol delivery. | |
Pulmonary hypertension without hypoxemia | HFNC flow set at 5–10 L/min is recommended | Adult in vivo [15] |
Pulmonary hypertension with refractory hypoxemia | Titrating HFNC flow at bedside based on patient’s response in order to determine the optimal flow for each individual patient is recommended | Adult in vivo [15] |