How I approach weaning from venoarterial ECMO
Critical Care volume 24, Article number: 307 (2020)
Introduction
Venoarterial extracorporeal membrane oxygenation (VA-ECMO) is a temporary mechanical circulatory support device capable of providing robust cardiopulmonary support for days to weeks, most commonly in the setting of refractory cardiogenic shock or cardiac arrest as a bridge to recovery or heart replacement therapies (HRT, e.g., durable left ventricular assist device or cardiac transplantation) [1]. Given the complications associated with prolonged use of VA-ECMO, it is imperative to institute systematic weaning strategies. In this article, we review our standardized approach to VA-ECMO weaning in adult patients.
Considerations prior to weaning
VA-ECMO supports circulation while treatment of the underlying pathology is prioritized to facilitate successful weaning or bridge to HRT. Because the potential for ventricular recovery is often difficult to ascertain, we advocate that all patients where HRT is a realistic consideration undergo early evaluation even if the intended goal is recovery [2,3,4,5].
Our center favors the use of invasive hemodynamics whenever possible to guide therapy in VA-ECMO. A pulmonary artery catheter provides valuable information about left ventricular (LV) loading conditions [6]. A right upper extremity arterial line should be maintained in all patients with femoral cannulation to facilitate monitoring of pulsatility and oxygenated blood flow to the arch of the aorta [7]. We generally target a cardiac index > 2.2 L/min/m2, mean arterial pressure (MAP) 65–80 mmHg, central venous pressure 8–12 mmHg, and pulmonary capillary wedge pressure < 18 mmHg. Arterial blood gas (ABG), lactate, hepatic, and renal function are followed in serial laboratory measurements to assess adequacy of end-organ perfusion.
In the event of concomitant use of a LV venting device [intra-aortic balloon pump (IABP) or percutaneous left ventricular assist device (pLVAD)], it is our approach to prioritize weaning and removing VA-ECMO first, given its propensity to increase afterload on a failing myocardium and relatively high complication rate [8]. One exception is when complications arise directly related to the ancillary device necessitating its removal. When used as a LV venting device, we maintain the IABP at 1:1 for unloading and ensuring adequate washing of the aortic root. Concomitant pLVADs, which in our institution is most often an Impella (Abiomed, Danvers, MA), are generally maintained at relatively low levels of flow (1.5–2 LPM) as its primary use in this setting is LV decompression rather than circulatory support [8,9,10].
Prior to weaning VA-ECMO flow, we wean vasoactive medications to low levels, given the deleterious effects associated with these medications, including arrythmia, renal injury, and limb ischemia. The following criteria should be met prior to weaning VA-ECMO: first, the patient phenotype is compatible with recovery; second, end-organ function is recovering; third, Pa02/Fi02 > 100; and fourth, vasopressors and inotropes are at reasonably low levels (for instance norepinephrine ≤ 4 μg/min or dobutamine < 5 mcg/kg/min). When these criteria are met, we initiate a 3-part approach to weaning that includes the following: (1) daily weaning study, (2) bedside assessment for decannulation, and (3) final assessment (Fig. 1).
VA-ECMO weaning process. *Criteria required to initiate weaning trial: (1) Phenotype is compatible with recovery (2) End-organ function is improving (3) Pa02/Fi02 > 100 (4) Vasopressors and intropes at low levels (norepinephrine ≤ 4 μg/min, dobutamine < 5 mcg/kg/min) **Any of the following criteria constitutes failure of a weaning trial: (1) MAP falls below 65–70 mmHg or decreases by more than 10 mmHg from baseline (2) Significant increase in intracardiac filling pressures (3) Deterioration in respiratory status
Daily weaning study
We perform daily transient reductions in ECMO flow rate in all patients to assess suitability for weaning, analogous to the strategy of daily spontaneous breathing trials in mechanically ventilated patients. We reduce flow rates in increments of 0.5 LPM. At each incremental drop in blood flow to a minimum of 2 LPM, we wait approximately 1 min to assess the effect on MAP and intracardiac pressures. If the MAP falls more than 10–15 mmHg or below 65 mmHg, the patient is not yet ready to wean to that level. Significant increases in right-sided filling pressures during weaning may also constitute failure especially in the setting of predominant right ventricular failure. At the end of the study, the blood flow rate is set at the lowest level achieved in the weaning study required to maintain stable MAP and intracardiac pressures. ABG, lactate, and a full set of invasive hemodynamics are obtained at this new flow rate to detect any impact on hemodynamics, tissue perfusion, or respiratory status [7]. Because a subset of patients who tolerate reduced support transiently are unable to tolerate the decreased support over a prolonged duration, lower flow rates are maintained for a minimum of 8 h prior to further weaning attempts. Weaning trials are performed at least every 24 h. Bedside echocardiography is used to provide additional information about native cardiac function as blood flow rates are reduced, particularly if the patient has failed prior weaning attempts. Prior literature has suggested the following parameters to be associated with successful weaning: aortic VTI ≥ 10 cm, LVEF > 20–25%, and lateral mitral annulus peak systolic velocity > 6 cm/s [11].
Bedside assessment for decannulation
After a patient has tolerated a trial of 2 LPM for a minimum of 8 h with stable end-organ function, a bedside assessment for decannulation is performed. Flow is gradually decreased to 1 LPM over the course of approximately 1 min to detect hemodynamic instability with minimal ECMO support. If tolerated, blood flow is then returned to 2 LPM, and plans are made for decannulation if the underlying cause of initial decompensation has been sufficiently addressed such that liberation from VA-ECMO is possible. If the patient fails the bedside assessment, then flow is returned to 2 LPM with plans to reassess every 24 h.
Final assessment
When a patient has tolerated a trial of 2 LPM of blood flow for a minimum of 8 h with stable hemodynamics and end-organ function and tolerates transient flow reduction to 1 LPM at the bedside, the patient is typically taken to the operating room for ECMO decannulation. The final assessment is performed at this time, wherein blood flows are gradually decreased and the cannulae are clamped; hemodynamic and ABG parameters are checked. Focused echocardiography is often undertaken to assess the impact of flow reduction on biventricular function. If acceptable, decannulation is performed.
Conclusion
Weaning from VA-ECMO remains a challenging but critically important step in device management. The key is to balance minimizing complications associated with device support with the potential for hemodynamic deterioration if support is prematurely discontinued. We advocate for a standardized and systematic approach to weaning but also recognize that deviation from the above approach may be required in specific clinical scenarios.
Availability of data and materials
Not applicable
References
Guglin M, Zucker MJ, Bazan VM, Bozkurt B, El Banayosy A, Estep JD, et al. Venoarterial ECMO for adults: JACC scientific expert panel. J Am Coll Cardiol. 2019;73(6):698–716.
Cheng YT, Garan AR, Sanchez J, Kurlansky P, Ando M, Cevasco M, et al. Midterm outcomes of bridge-to-recovery patients after short-term mechanical circulatory support. Ann Thorac Surg. 2019;108(2):524–30.
Garan AR, Malick WA, Habal M, Topkara VK, Fried J, Masoumi A, et al. Predictors of survival for patients with acute decompensated heart failure requiring extra-corporeal membrane oxygenation therapy. ASAIO J. 2019;65(8):781–7.
Schmidt M, Burrell A, Roberts L, Bailey M, Sheldrake J, Rycus PT, et al. Predicting survival after ECMO for refractory cardiogenic shock: the survival after veno-arterial-ECMO (SAVE)-score. Eur Heart J. 2015;36(33):2246–56.
Garan AR, Eckhardt C, Takeda K, Topkara VK, Clerkin K, Fried J, et al. Predictors of survival and ability to wean from short-term mechanical circulatory support device following acute myocardial infarction complicated by cardiogenic shock. Eur Heart J Acute Cardiovasc Care. 2018;7(8):755–65.
Truby LK, Takeda K, Mauro C, Yuzefpolskaya M, Garan AR, Kirtane AJ, et al. Incidence and implications of left ventricular distention during venoarterial extracorporeal membrane oxygenation support. ASAIO J. 2017;63(3):257–65.
Brodie D, Slutsky AS, Combes A. Extracorporeal life support for adults with respiratory failure and related indications: a review. JAMA. 2019;322(6):557–68.
Cevasco M, Takayama H, Ando M, Garan AR, Naka Y, Takeda K. Left ventricular distension and venting strategies for patients on venoarterial extracorporeal membrane oxygenation. J Thorac Dis. 2019;11(4):1676–83.
Russo JJ, Aleksova N, Pitcher I, Couture E, Parlow S, Faraz M, et al. Left ventricular unloading during extracorporeal membrane oxygenation in patients with cardiogenic shock. J Am Coll Cardiol. 2019;73(6):654–62.
Akanni OJ, Takeda K, Truby LK, Kurlansky PA, Chiuzan C, Han J, et al. EC-VAD: combined use of extracorporeal membrane oxygenation and percutaneous microaxial pump left ventricular assist device. ASAIO J. 2019;65(3):219–26.
Aissaoui N, Luyt CE, Leprince P, Trouillet JL, Leger P, Pavie A, et al. Predictors of successful extracorporeal membrane oxygenation (ECMO) weaning after assistance for refractory cardiogenic shock. Intensive Care Med. 2011;37(11):1738–45.
Acknowledgements
Susan Tanzi Pfeifer helped with the creation of the figure.
Funding
None
Author information
Authors and Affiliations
Contributions
All authors were a part of drafting/revision of the manuscript and approved the final manuscript.
Corresponding author
Ethics declarations
Ethics approval and consent to participate
Not applicable
Consent for publication
Not applicable
Competing interests
Dr. Brodie receives research support from ALung Technologies; he was previously on their medical advisory board. He has been on the medical advisory boards for Baxter, BREETHE, Xenios, and Hemovent. Dr. Fried, Dr. Masoumi, and Dr. Takeda have no disclosures.
Additional information
Publisher’s Note
Springer Nature remains neutral with regard to jurisdictional claims in published maps and institutional affiliations.
Rights and permissions
Open Access This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article's Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article's Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.
About this article
Cite this article
Fried, J.A., Masoumi, A., Takeda, K. et al. How I approach weaning from venoarterial ECMO. Crit Care 24, 307 (2020). https://doi.org/10.1186/s13054-020-03010-5
Received:
Accepted:
Published:
DOI: https://doi.org/10.1186/s13054-020-03010-5
Keywords
- Venoarterial ECMO weaning
- VA-ECMO weaning
- ECMO weaning
- Cardiogenic shock ECMO
- Extracorporeal life support
- ECLS