Table 1 Specific reasons and multiplex series approach mitigation: The SCCM/ASA/AARC/AACN/ASPF/CHEST consensus statement [1] strongly recommends against parallel breathing and ventilation of patients; this poses a wide range of valid criticisms. We address these criticisms in terms of the serial ventilation approach/concept presented VC volume-controlled, PC pressure-controlled ventilation, PEEP positive end-expiratory pressure, PIP peak inspiratory pressure, Vt tidal volume

From: Safe doubling of ventilator capacity: a last resort proposal for last resorts

Consensus statement specific critiques (SCCM/ASA/AARC/AACN/ASPF/CHEST):

Mitigation by serial ventilation approach vs parallel

Volumes would go to the most compliant lung segments.

Serial breathing ventilates a single lung at a time, and thus, using a volume-controlled (VC) mode, lung compliances are not ‘mixed’ and do not create this same, critical problem. A pressure controlled (PC) mode will also be separated.

Critically, in all modes, each lung responds individually to the inputs.

Positive end-expiratory pressure, which is of critical importance in these patients, would be impossible to manage.

PEEP can be individually set using PEEP valves on the expiratory circuit and putting PEEP = 0 on the ventilator. These valves are commonly available and some come with multiple settings. Thus, PEEP may also be individualised

Monitoring patients and measuring pulmonary mechanics would be challenging, if not impossible.

Patients are split in serial breathing so inspiration does not overlap, and any monitoring present would monitor each inspiratory portion (at least) separately.

Monitoring mechanics would depend on the ventilator interface and monitoring algorithms used, thus the displayed patient-specific parameters would be averaged. However, clinicians could still examine breath by breath waveforms or PV loops. PIP or Vt alarm limits could still be used as these are based on safety settings determined for a population of patients, rather than individual patients. Again, these outcomes are enabled by separating inspiration for both patients.

Alarm monitoring and management would not be feasible.

See above, again by separating patient inspiration segments in serial ventilation this issue is mitigated.

Individualised management for clinical improvement or deterioration would be impossible.

PC driving pressure and VC tidal volume would have to be the same as ventilators currently do not have the capability to enable alternating breath settings. Clinical judgement would determine which one of the ventilation is most appropriate for this situation. Where there are significant differences in compliance, a volume-controlled mode may be preferable.

However, PEEP would be individualised via separate PEEP valves. These PEEP valves could also be made active if desired, or set manually similarly to changing PEEP on a ventilator, but for each patient.

In the case of a cardiac arrest, ventilation to all patients would need to be stopped to allow the change to bag ventilation without aerosolizing the virus and exposing healthcare workers. This circumstance also would alter breath delivery dynamics to the other patients.

In this case, the patient still on the ventilator can be restored to a 1 patient, 1 ventilator standard use, after the other patient is disconnected.

Alternatively, a rubber bag (test lung) could be swapped in while the arrested patient is being hand ventilated during CPR. This would not involve having to make changes to the ventilator settings, which would create cognitive overload in the event of a cardiac arrest.

The added circuit volume defeats the operational self-test (the test fails). The clinician would be required to operate the ventilator without a successful test, adding to errors in the measurement.

Self-testing can be carried out in the usual manner. There is no added circuit volume as individual breaths are within usual physiological limits and therefore not vulnerable to errors of extrapolation created by connecting patients in parallel.

Additional external monitoring would be required. The ventilator monitors the average pressures and volumes.

In serial breathing, each breath would be presented. The clinician would have to know to identify each patient by examining their breathing directly, to know which waveform or PV loop corresponds to a particular patient.

Even if all patients connected to a single ventilator have the same clinical features at initiation, they could deteriorate and recover at different rates, and distribution of gas to each patient would be unequal and unmonitored. The sickest patient would get the smallest tidal volume and the improving patient would get the largest tidal volume.

Since each patient is separated, there is less need for matching compliance or resistance, the latter of which would be similar. Specifically, for the following:

PC: driving pressure would have to be the same. However, a resistor with known pressure drop can be added to one of the two inspiratory circuits to reduce driving pressure for one patient. Tidal volume alarming would still be feasible and ventilator controlled to avoid injury or damage.

VC: tidal volume would be the same for both patients, where we would recommend setting tidal volume for the smaller of two patients in ml/kg; however, a vast difference could be problematic requiring some light matching by approximate size. PIP alarms and limits would still be applicable and ventilator controlled.

The greatest risks occur with sudden deterioration of a single patient (e.g. pneumothorax, kinked endotracheal tube), with the balance of ventilation distributed to the other patients.

Patients are ventilated separately so changes in patient condition, resulting in tidal volume changes (during PC) or peak pressure changes (during VC) would be notable on the monitor and ventilator set limits and alarms would still work be useful.

Finally, there are ethical issues. If the ventilator can be lifesaving for a single individual, using it on more than one patient at a time risks life-threatening treatment failure for all of them.

The best way to ventilate 2 patients on 1 ventilator is not to do it!

Given the exigency of no other alternative, we propose this method is currently the next best way.