From: The future of mechanical ventilation: lessons from the present and the past
Brochard et al. [45] | Stewart et al. [46] | Brower et al. [47] | ARDS Network [35] | Esteban et al. [62] | Brower et al. [75] | Meade et al. [76] | Briel et al. [77] | Villar et al. [63] | Guerin et al. [73] | Bellani et al. [64] | ||||||||
---|---|---|---|---|---|---|---|---|---|---|---|---|---|---|---|---|---|---|
Type | RCT | RCT | RCT | RCT | Observation | RCT | RCT | Meta-analysis | Observation | RCT | Observation | |||||||
Year | 1998 | 1998 | 1999 | 2000 | 2002 | 2004 | 2008 | 2010 | 2011 | 2013 | 2016 | |||||||
Number of patients | 58 | 58 | 60 | 60 | 100 | 100 | 387 | 405 | 231 | 236 | 258 | 508 | 475 | 1163 | 1136 | 255 | 229a | 2377 |
Vt/PBW (ml/kg) | 7.1 | 10.3 | 7.0 | 10.7 | 10.2 | 7.3 | 6.2 | 11.8 | 8.7 | 6.1 | 6 | 6.8 | 6.8 | 6.3 | 6.3 | 7.2 | 6.1 | 7.6 |
RR (bpm) | 22.1 | 15.6 | 29 | 16 | 20 | 29 | 29 | 26 | 25.2 | 27 | 20.8 | |||||||
Peak pressure (cmH2O) | 24.2 | 32.1 | 32 | 39 | 34 | 27 | ||||||||||||
Plateau pressure (cmH2O) | 25.7 | 31.7 | 22.3 | 26.8 | 30.6 | 24.9 | 25 | 33 | 28 | 24 | 27 | 24.9 | 30.2 | 23 | 29 | 26 | 23 | 23 |
PEEP (cmH2O) | 10.7 | 10.7 | 8.6 | 7.2 | ~8.2 | ~9.5 | 9.4 | 8.6 | 8 | 8.9 | 14.7 | 10.1 | 15.6 | 9 | 15 | 9.3 | 10 | 8.5 |
Mortality (%) | 46.6b | 37.9b | 50c | 47c | 46c | 50c | 31c | 39.8c | 52c | 24.9c | 27.5c | 32.3d | 28.4d | 36.6c | 30.3c | 42.7c | 32,8d | 35.3c |