Skip to main content

Table 2 Glucose metrics

From: Manual versus Automated moNitoring Accuracy of GlucosE II (MANAGE II)

Variable

YSIa

OptiScanner®b

Number of measurements in the ICU

944

929

Number of measurements per patient, median (IQR)

9 (7–12)

9 (7–12)

Blood glucose level (mg/dl), median (IQR)

140.5 (122–169)

145 (123–169.5)

Blood glucose level (mg/dl), mean ± SD

150.6 ± 44.5

150.5 ± 42.2

Range of glucose (mg/dl)

59.7–484

57–436

Measurements of severe hypoglycemia (<40 mg/dl), number (%)

0

0

Patients with severe hypoglycemia, number (%)

0

0

Measurements of mild hypoglycemia (41–70 mg/dl), number (%)

5 (0.5%)

6 (0.6%)

Patients with mild hypoglycemia, number (%)

4 (5%)

3 (3%)

Measurements of severe hyperglycemia (>180 mg/dl), number (%)

167 (18%)

157 (17%)

Patients with severe hyperglycemia, number (%)

46 (52%)

47 (53%)

  1. OptiScanner® glucose metrics are based on the measurements that have a corresponding value for YSI
  2. aYSI 2300 STAT Plus analyzer (Yellow Spring Instruments, Yellow Spring, OH, USA)
  3. bOptiscan Biomedical Corporation, Hayward, CA, USA
  4. ICU intensive care unit, IQR interquartile range, SD standard deviation