Fig. 5 | Critical Care

Fig. 5

From: Reliability of standard pupillometry practice in neurocritical care: an observational, double-blinded study

Fig. 5

Assessment of PLR by nursing staff. Percentage agreement (green and light green) or discrepancy (red and orange) in the assessment of PLR either by standard visual examination or using the automated pupillometer. The presence (+) or absence (–) of PLR was evaluated by nurses using a penlight and by physicians using a calibrated light stimulus delivered by the pupillometer. A global rate of discordance (red and orange) of 18 % (72/406) was found between the two techniques when assessing the PLR. For measurements with small pupils (diameters <2 mm) the error rate was at its greatest: 39 % (24/61). (Color figure online)