From: Accuracy of intracranial pressure monitoring: systematic review and meta-analysis
Author and year of publication | Probe type | Readings (n)a | Test duration (days) | ICP range (mmHg) | |
---|---|---|---|---|---|
Clinical studies | |||||
1 | Statham 1993 [22] | Fiberoptic | 11 | 1–11 | 0/+4 |
2 | Bavetta 1997 [27] | Fiberoptic | 83 | 1–12 | −12/+14 |
3 | Münch1998 [28] | Fiberoptic | 95 | 1–17 | −15/+12 |
4 | Martines-Mañas 2000 [29] | Fiberoptic | 56 | 1–12 | −24/+35 |
5 | Poca 2002 [19] | Fiberoptic | 126 | 1–11 | −12/+7 |
6 | Stendel 2003-1 [17] | Fiberoptic | 50 | 1–32 | 0/+12 |
7 | Gelabert-González 2006 [9] | Fiberoptic | 624 | 1–5 | −17/+21 |
8 | Gopinath 1995 [23] | Microstrain gauge | 25 | 2–7 | −2/+2 |
9 | Stendel 2003-2 [17] | Microstrain gauge | 98 | 3–28 | −2/+3 |
10 | Koskinen 2005 [8] | Microstrain gauge | 128 | 1–16 | −5/+4 |
11 | Citerio 2008 [30] | Microstrain gauge | 89 | 1–10 | −4/+8 |
12 | Al-Tamimi 2009 [20] | Microstrain gauge | 88 | 3; 6b | NA |
13 | Lang 2003 [31] | Microstrain gauge | 84 | 3–28 | −2/+2 |
Laboratory studies | |||||
14 | Czosnyka 1996-1 [32] | Fiberoptic | 1 | 3 | −0.8/+0.8 |
15 | Czosnyka 1996-2 [32] | Fiberoptic | 1 | 3 | −0.4/+0.4 |
16 | Piper 2001 [33] | Fiberoptic | 34 | 1–12 | −13/+22 |
17 | Sundbӓrg 1987 [12] | Microstrain gauge | 1 | 3 | −2/+2 |
18 | Czosnyka 1996-3 [32] | Microstrain gauge | 1 | 3 | −0.8/+0.8 |
19 | Morgalla 1999 [10] | Microstrain gauge | 7 | 10 | −4/+3 |
20 | Citerio 2004 [11] | Microstrain gauge | 10 | 5 | 0/+2 |