Table 3 Main features of the studies reporting drift

From: Accuracy of intracranial pressure monitoring: systematic review and meta-analysis

 

Author and year of publication

Probe type

Readings (n)a

Test duration (days)

ICP range (mmHg)

Clinical studies

1

Statham 1993 [22]

Fiberoptic

11

1–11

0/+4

2

Bavetta 1997 [27]

Fiberoptic

83

1–12

−12/+14

3

Münch1998 [28]

Fiberoptic

95

1–17

−15/+12

4

Martines-Mañas 2000 [29]

Fiberoptic

56

1–12

−24/+35

5

Poca 2002 [19]

Fiberoptic

126

1–11

−12/+7

6

Stendel 2003-1 [17]

Fiberoptic

50

1–32

0/+12

7

Gelabert-González 2006 [9]

Fiberoptic

624

1–5

−17/+21

8

Gopinath 1995 [23]

Microstrain gauge

25

2–7

−2/+2

9

Stendel 2003-2 [17]

Microstrain gauge

98

3–28

−2/+3

10

Koskinen 2005 [8]

Microstrain gauge

128

1–16

−5/+4

11

Citerio 2008 [30]

Microstrain gauge

89

1–10

−4/+8

12

Al-Tamimi 2009 [20]

Microstrain gauge

88

3; 6b

NA

13

Lang 2003 [31]

Microstrain gauge

84

3–28

−2/+2

Laboratory studies

14

Czosnyka 1996-1 [32]

Fiberoptic

1

3

−0.8/+0.8

15

Czosnyka 1996-2 [32]

Fiberoptic

1

3

−0.4/+0.4

16

Piper 2001 [33]

Fiberoptic

34

1–12

−13/+22

17

Sundbӓrg 1987 [12]

Microstrain gauge

1

3

−2/+2

18

Czosnyka 1996-3 [32]

Microstrain gauge

1

3

−0.8/+0.8

19

Morgalla 1999 [10]

Microstrain gauge

7

10

−4/+3

20

Citerio 2004 [11]

Microstrain gauge

10

5

0/+2

  1. Studies reporting drift were classified according to the setting (clinical or laboratory). Some papers reported more than one study; one paper is cited three times. Since different probes were analyzed in some studies, a single reference may appear more than once
  2. aIndicates how many readings of drift were used to calculate the range, mean and/or median (see text for more details)
  3. bThe study by Al-Tamimi et al. included data from two centers and reported the median observation time for both
  4. ICP Intracranial pressure, NA Not available