Table 2 Limits of agreement of the studies included in the meta-analysis

From: Accuracy of intracranial pressure monitoring: systematic review and meta-analysis

 

Author and year of publication

Patients (n)

Lower limit of agreement

Mean difference

Upper limit of agreement

Group 1

1

Schickner 1992 [4]

10

−9.31

9.20

27.71

2

Chambers 1993 [21]

10

−3.92

1.43

6.78

3

Statham 1993 [22]

11

−7.09

2.73

12.55

4

Gopinath 1995 [23]

25

−4.97

0.50

5.97

5

Signorini 1998-2 [24]

5

−9.58

1.81

13.21

6

Chambers 2001 [25]

11

−11.51

0.10

11.71

7

Koskinen 2005 [8]

22

−5.86

1.20

8.26

8

Lescot 2011-1 [26]

15

−9.08

–0.60

7.88

9

Lescot 2011-2 [26]

15

−7.50

0.30

8.10

10

Eide 2012-2 [18]

5

−37.97

4.52

47.01

Group 2

11

Signorini 1998-1 [24]

5

−8.65

1.58

11.82

12

Sahuquillo 1999-1 [2]

33

–0.47

1.80

4.07

13

Sahuquillo 1999-2 [2]

16

−6.56

7.50

21.56

14

Eide 2012-1 [18]

5

−16.37

0.64

17.65

15

Eide 2012-3 [18]

7

−6.17

0.70

7.57