From: Ventilator-derived carbon dioxide production to assess energy expenditure in critically ill patients: proof of concept

Fig. 4

Accuracy and inaccuracy of the different methods quantified in less than 10 % and less than 15 % accuracy rates and greater than 25 % and greater than 30 % inaccuracy rates. a Less than 10 % and less than 15 % accuracy rates were defined as the proportion of patients for whom energy expenditure (EE) was predicted within 10 % and within 15 % of gold standard EE:Calorimetry. b Greater than 25 % and greater than 30 % inaccuracy rates were defined as the proportion of patients for whom EE differed by more than 25 % and more than 30 % from gold standard EE:Calorimetry. The x-axis shows the different methods used to assess EE. The y-axis represents the accuracy rates or inaccuracy rates in percentages. The error bars reflect upper bounds of 95 % confidence intervals. *Significantly different from EE:VCO2 (p values are shown in Table 3). EE:Esp25, Energy expenditure calculated with the European Society for Clinical Nutrition and Metabolism guideline equation of 25 kcal/kg/day; EE:Faisy, Energy expenditure calculated with the Faisy equation; EE:HB, Energy expenditure calculated with the Harris–Benedict equation; EE:PSU, Energy expenditure calculated with the Penn State University 2003b equation; EE:VCO2, Energy expenditure from ventilator-derived volume of carbon dioxide and nutritional respiratory quotient