Patient

PS level

PS rise time (s)

PS cycle off

NAVA level

NAVA

PEEP

FiO2

(cmH2O)

(% peak flow)

(cmH2O/μV)

trigger (μV)

(cmH2O)

(%)

1

6

0.20

30

0.1

0.5

6

70

2

10

0.20

50

0.8

0.5

8

50

3

8

0.20

70

0.4

0.5

8

50

4

5

0.20

50

0.1

1.0

7

60

5

6

0.20

50

0.1

0.5

4

40

6

5

0.05

50

5.0

0.5

5

30

7

10

0.00

50

0.2

0.5

6

55

8

6

0.20

50

0.2

0.5

6

45

9

6

0.20

50

0.2

0.5

6

70

10

8

0.00

60

0.1

0.5

5

35

11

6

0.20

50

0.2

0.5

6

40

  1. Pressure support (PS) levels and rise time hold for both ventilators, whereas cycle-off criteria is only set for NIV-PSVServo-I. The BiPAP Vision uses the Auto-Trak Sensitivity algorithm to trigger and cycle off the ventilator and cannot be set individually. PEEP and FiO2 were similar for all three ventilatory modes. FiO2: inspired oxygen fraction; NAVA: neurally adjusted ventilatory assist; PEEP: positive end-expiratory pressure.