- Poster presentation
- Open access
- Published:
Central line change in potential catheter-related bloodstream infection: target for intervention to reduce harm
Critical Care volume 15, Article number: P230 (2011)
Introduction
Central venous catheterization is routine in critical care, but a potential source of harm. Forty-two per cent of bloodstream infections in England are central-line related [1], at a substantial cost to the health service. Early catheter removal is vital for source control where catheter-related bloodstream infection (CRBSI) is suspected. Furthermore, a model encompassing daily review and removal of unnecessary catheters has been shown to reduce the risk [2]. We studied the time from decision to removal of existing central venous catheters (CVCs), and evaluated potential reasons for delay.
Methods
This is a retrospective review of practice at a 43-bed medical/surgical ICU at a London teaching hospital, using computerized patient records. All patients requiring a change of CVC over a 2-month period in 2010 were included. Change of CVC was defined as the time from decision to removal of the old CVC, incorporating new CVC insertion. Sepsis was defined as rising inflammatory markers, an impression of local/systemic infection, or emergency (unsterile) insertion. Routine was defined as no signs of infection, usually at 5 to 7 days or if accidentally dislodged/blocked.
Results
Seventy-eight CVC changes were performed, 45 (57.7%) for sepsis and 33 (42.3%) as routine. The median time to change a septic CVC was 742.5 minutes (106 to 2,038 minutes). The median time for a routine change was 611 minutes (130 to 1,759 minutes). On average, 70% of the time taken to change a CVC involved new catheter insertion. Where the tip position was confirmed with a chest X-ray scan, it took a median of 182 minutes longer (-97 to 946 minutes) to change the CVC. Check X-ray review was documented in 28 (45.1%) of 62 internal jugular/subclavian CVCs and only five X-ray scans resulted in repositioning. Where inotropes/vasopressors were administered, it took a median of 209 minutes longer (106 to 599 minutes) for CVC change. Where coagulation products were administered, it took a median of 168.5 minutes longer (209 to 279 minutes) for CVC change.
Conclusions
Our data suggest that in our unit the duration of catheter change in the critically ill is a prolonged process, and took longer where potential harm is greatest. Check X-ray scans infrequently result in CVC repositioning, contribute to delays and could be performed after old-CVC removal. We plan to audit the changes we have made, and believe that the timely exchange of old CVCs should be incorporated into models aiming to reduce the impact of CRBSI.
References
Smyth ETM, et al.: J Hosp Infect. 2006, 69: 230-248. 10.1016/j.jhin.2008.04.020
Pronovost PJ, et al.: N Engl J Med. 2006, 335: 2725-2732. 10.1056/NEJMoa061115
Author information
Authors and Affiliations
Rights and permissions
This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article's Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article's Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.
About this article
Cite this article
Davies, R., Lowings, M., Jones, A. et al. Central line change in potential catheter-related bloodstream infection: target for intervention to reduce harm. Crit Care 15 (Suppl 1), P230 (2011). https://doi.org/10.1186/cc9650
Published:
DOI: https://doi.org/10.1186/cc9650