- Poster presentation
- Open access
- Published:
Positive end-expiratory pressure improves oxygenation inducing ventral-to-dorsal tidal ventilation redistribution: an electrical impedance tomography study
Critical Care volume 15, Article number: P190 (2011)
Introduction
Positive end-expiratory pressure (PEEP) improves oxygenation in acute lung injury (ALI) patients by increasing end-expiratory lung volume (EELV). Electrical impedance tomography (EIT) is a relatively new non-invasive bedside method to monitor regional distribution of tidal ventilation and EELV changes, validated in preclinical studies. We tested EIT as a monitor of PEEP-induced tidal redistribution and EELV changes in ALI patients, and the relationship between EIT parameters and oxygenation.
Methods
We enrolled 14 consecutive ALI patients admitted to our ICU, intubated and undergoing mechanical ventilation. We monitored the regional tidal ventilation distribution by means of a new EIT system (PulmoVista 500®; Dräger Medical GmbH, Lübeck, Germany) dividing the lung imaging field into four contiguous same-size regions of interest (ROIs): ventral right (ROI 1) and left (ROI 2) and dorsal right (ROI 3) and left (ROI 4). EIT allowed us to measure changes in EELV at different PEEP levels by measuring differences in end-expiratory total lung electrical impedance. We randomly performed the following two steps for 15 minutes, leaving tidal volume and FiO2 unchanged: PEEPlow (clinical) and PEEPhigh (PEEPlow +5 cmH2O). At the end of each step, we recorded: ventilation parameters, arterial blood gas analysis, percentage of tidal ventilation distribution in the four ROIs and EELV change. Analyses were performed by paired t test and linear regression.
Results
Patients were 55 ± 12 years old and seven were women. ALI etiology was: trauma (14%), septic shock (21%), pneumonia (37%) and postoperative respiratory failure (28%). PEEPlow was 7 ± 2 cmH2O and PEEPhigh 12 ± 3 cmH2O. At PEEPhigh, PaO2/FiO2 significantly ameliorated (266 ± 98 vs. 287 ± 102 mmHg, P = 0.0003), the proportional distribution of tidal ventilation changed in all four ROIs (ROI 1 to ROI 4: 34 ± 14 vs. 29 ± 9%, P = 0.03; 33 ± 13 vs. 30 ± 11%, P = 0.12; 16 ± 9 vs. 20 ± 10%, P = 0.002; 17 ± 7 vs. 21 ± 6%, P = 0.002), moving from ventral to dorsal, and EELV increased by 349 ± 121 ml. Changes in PaO2/FiO2 correlated better with ventral-to-dorsal shifts of tidal ventilation than with EELV changes (r = 0.499, P = 0.08; r = -0.399, P = 0.18).
Conclusions
EIT allowed us to detect ventral-to-dorsal tidal ventilation redistribution at higher PEEP levels. This mechanism may be a key determinant of PEEP-induced oxygenation improvement.
Author information
Authors and Affiliations
Rights and permissions
This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article's Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article's Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.
About this article
Cite this article
Mauri, T., Bellani, G., Pradella, A. et al. Positive end-expiratory pressure improves oxygenation inducing ventral-to-dorsal tidal ventilation redistribution: an electrical impedance tomography study. Crit Care 15 (Suppl 1), P190 (2011). https://doi.org/10.1186/cc9610
Published:
DOI: https://doi.org/10.1186/cc9610