Ventilatory ratio: validation in an ex vivo model and analysis in ARDS/ALI patients

Introduction

Several indices exist to monitor adequate oxygenation, but no such index exists for ventilatory efficiency. The ventilatory ratio (VR) is a simple tool to monitor changes in ventilatory efficiency using variables commonly measured at the bedside [1]:

VR = V ˙ E m e a s u r e d × P a c o 2 m e a s u r e d V ˙ E p r e d i c t e d × P a c o 2 p r e d i c t e d

See Figure 1 overleaf (where predicted values are VE 100 ml/kg/minute and PaCO2 5 kPa).

Figure 1
figure 1

Chi-squared test for trends P = 0.0015.

Methods

The Nottingham Physiology Simulator (NPS), a validated computational model of cardiopulmonary physiology [2], was used to validate the ability of VR to reflect ventilatory efficiency ex vivo. Three virtual patients were configured, representing healthy lung, ARDS and COPD. VR was calculated while minute ventilation, ventilation rate and VCO2 were each varied in isolation. The clinical uses of VR were then examined in a database comprising 122 patients with ALI and ARDS [3]. Standard respiratory data and VR values were analysed in all patients.

Results

The NPS model showed significant correlation between VR and physiological deadspace fraction (Vd/Vtphys) at constant VCO2 (P < 0.001, r = 0.99). Similarly, VCO2 had a linear relationship with VR at constant Vd/Vtphys. Across the various ventilatory configurations the median values and ranges of calculated VR for the three patients were as follows: normal patient VR 0.89 (0.61 to 1.36), COPD 1.36 (0.95 to 1.89) and ARDS 1.73 (1.2 to 2.62). In the ALI/ARDS database the range of values for VR was 0.56 to 3.93 (median 1.36). Patients with ARDS had a significantly higher VR in comparison with patients with ALI (1.44, 1.25 to 1.77 vs. 1.25, 0.94 to 1.6, P = 0.02). VR was significantly higher in nonsurvivors as compared with survivors (1.7 ± 0.64 vs. 1.45 ± 0.56, P < 0.03). There was poor correlation between PaO2/FiO2 ratio and VR in the population (r = -0.32, 95% CI = -0.47 to -0.15).

Conclusions

Ex vivo modling shows that VR can be simply and reliably used to monitor ventilatory efficiency at the bedside. VR is influenced by changing CO2 production and deadspace ventilation. As a clinical tool it is a predictor of outcome and is independent to PaO2/FiO2 ratio.

References

  1. Sinha , et al.: Br J Anaesth. 2009, 102: 692-697. 10.1093/bja/aep054

    Article  CAS  PubMed  Google Scholar 

  2. Hardman , et al.: Br J Anaesth. 1998, 81: 327-329.

    Article  CAS  PubMed  Google Scholar 

  3. Bersten , et al.: Am J Respir Crit Care Med. 2002, 165: 443-448.

    Article  PubMed  Google Scholar 

Download references

Author information

Authors and Affiliations

Authors

Rights and permissions

This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article's Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article's Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.

Reprints and permissions

About this article

Cite this article

Sinha, P., Corrie, K., Bersten, A. et al. Ventilatory ratio: validation in an ex vivo model and analysis in ARDS/ALI patients. Crit Care 15 (Suppl 1), P179 (2011). https://doi.org/10.1186/cc9599

Download citation

  • Published:

  • DOI: https://doi.org/10.1186/cc9599

Keywords