Table 1 Haemodynamic data at each experimental step

From: Reliability of continuous cardiac output measurement during intra-abdominal hypertension relies on repeated calibrations: an experimental animal study

Haemodynamic variable

Baseline

Fluid loading

IAH

IAH + Fluid

Mean arterial pressure (mmHg)

77 ± 20

99 ± 18*

109 ± 13*

112 ± 16*

Heart rate (beats/min)

87 ± 18

92 ± 18

92 ± 13

99 ± 12*

Central venous pressure (mmHg)

5 ± 2

7 ± 2

12 ± 4*

12 ± 3*

Pulmonary artery occlusion pressure (mmHg)

7 ± 2

11 ± 3

16 ± 6*

16 ± 3*

Intra-abdominal pressure (mmHg)

7 ± 2

7 ± 3

25 ± 3*

27 ± 4*

Chest wall compliance (ml/cmH2O)

58 ± 24

65 ± 22

22 ± 5*

21 ± 4*

Systemic vascular resistance (dyn·s/cm5)

963 ± 392

1006 ± 356

1038 ± 192

994 ± 303

Global end-diastolic volume (ml)

767 ± 164

883 ± 214*

928 ± 172*

997 ± 217*

PulseCO (l/min)

6.4 ± 2.0

7.1 ± 2.7*

6.8 ± 2.5

7.0 ± 2.8

PCCOpre (l/min)

n.a.

7.1 ± 1.1*

6.8 ± 2.0

7.9 ± 1.9

COTCP (l/min)

6.2 ± 1.6

7.5 ± 2.1*

8.0 ± 1.9*

9.4 ± 2.4*†‡

PCCOrecal (l/min)

6.2 ± 1.6

7.5 ± 2.1*

8.0 ± 1.9*

9.4 ± 2.4*†‡

CCOPAC (l/min)

6.4 ± 1.4

7.6 ± 1.4*

7.7 ± 1.5*

9.1 ± 2.1*†‡

  1. Data presented as the mean ± standard deviation. IAH, intra-abdominal hypertension; IAH + Fluid, second fluid load at IAH; PulseCO, continuous cardiac output by pulse power analysis; PCCOpre, continuous cardiac output by pulse contour analysis before calibration; COTCP, bolus transcardiopulmonary thermodilution cardiac output; PCCOrecal, continuous cardiac output by pulse contour analysis after recalibration; CCOPAC, continuous cardiac output by pulmonary artery catheter thermodilution. *P < 0.05 versus BL; P < 0.05 versus + Fluid; P < 0.05 versus IAH (PCCOpre versus previous PCCOrecal). n.a., not applicable.