Table 1 The patient test

From: A new device for 100 per cent humidification of inspired air

 

Humid-Heat

Fisher & Paykel MR730

P

Duration of the test (h)

24 ± 0.5

22± 4.0

0.27

Change in the patient's body temperature (°C)

-0.28 ± 0.36

+0.01 ± 1.1

0.56

Change in the temperature at the Y-piece or the flex-tube (°C)

-0.13 ± 0.64

0.0 ± 0.54

0.71

Suctions (n)

10.4 ± 3.9

9.2 ± 3.2

0.75

Saline instillations (n)

5.9 ± 3.1

4.6 ± 2.9

0.43

Total volume of saline instilled (ml)

30 ± 20

24 ± 23

0.41

No of nebulizations (salbutamol 2.5 mg plus 200 mg N-acetylcystein

3.6 ± 1.8

3.9 ± 2.0

0.75

diluted with saline to a total of 5 ml)

   

Quality of secretions at the start of the experiment (thick, normal, thin) (n)

thick 2, normal 2, thin 4

thick 4, normal 3, thin 1

0.26

Quality of secretions at the end of the experiment, patients (n)

thick 0, normal 5, thin 3

thick 1, normal 5, thin 2

0.18

Change in thickness of secretions (i = increase, d = decrease,

d = 3, i = 3, O = 2

D = 3, i = 1, O = 4

0.44

O = no change), patients (n)

   

Change in airway pressure at a prolonged expiratory pause

0

0

0.99

Water consumed/h (ml)

5.6 ± 1.2

30 ± 15

0.0008

Water consumed/l inspired gas (ml)

0.009 ± 0.003

0.043 ± 0.018

0.0008

Number of times that the water traps were emptied

0

8.3 ± 3.2

0.0008

Total amount of fluid in the water traps (ml)

0

213 ± 84

0.0008

Number of changes of machine filters due to excessive water condensation

0

4.7 ± 1.6

0.0008

  1. The numbers are presented as mean ± standard deviation whenapplicable.