From: Interference by new-generation mobile phones on critical care medical equipment
Type of device or incident | Number of devices | Distancea (cm) | Type of incident per signalb | |||
---|---|---|---|---|---|---|
 | Tested | Influenced |  | GPRS-1 | GPRS-2 | UMTS |
Intensive care unit ventilator | 9 | 7 | 1.5 [0.1–300] | 6H, 1L | 2H, 1S, 1L | 1H, 2S, 1L |
Critical care monitor | 13 | 7 | 3 [0.1–500] | 4S, 3L | 2S, 4L |  |
Syringe pump | 7 | 3 | 5 [0.1–50] | 2H, 1S | S | S |
Volumetric infusion pump | 4 | 1 | 30 | S | S | S |
Intra-aortic balloon pump | 2 | 1 | 0.1 | L | Â | Â |
Haemofiltration/dialysis | 5 | 1 | 15 | H | Â | Â |
External pacemaker | 4 | 1 | 3 | H | Â | Â |
Defibrillator | 3 | 1 | 0.1 | Â | Â | L |
12-lead EKG | 1 | 1 | 150 | S | S | S |
Fluid warmer | 2 | 1 | 6 | S | S | Â |
Enteral feeding pump | 2 | 1 | 30 | H | H | Â |
Air humidifier | 1 | 1 | 5 | H | Â | Â |
EKG telemetry | 1 | 0 | Â | Â | Â | Â |
Forced-air warming unit | 3 | 0 | Â | Â | Â | Â |
Mobile suction unit | 1 | 0 | Â | Â | Â | Â |
Critical care bed | 2 | 0 | Â | Â | Â | Â |
Continuous-airflow mattress | 1 | 0 | Â | Â | Â | Â |
Type of incidentb | Â | Â | Â | Â | Â | Â |
   Hazardous |  |  | 3.5 [0.1–300] |  |  |  |
   Significant |  |  | 25 [0.1–500] |  |  |  |
   Light |  |  | 0.1 [0.1–3] |  |  |  |
Total | 61 | 26 (43%) | 3 [0.1–500] | 25 (41%) | 15 (25%) | 8 (13%) |