Table 2 Biases and limits of agreement between blood pressure measurement techniques for each level of blood pressure: effect of BMI. Parenthesis indicate upper and lower limits of agreement

From: Arterial blood pressure monitoring in overweight critically ill patients: invasive or noninvasive?

Arterial blood pressure

BMI <30 kg/m2 (n = 24)

BMI >30 kg/m2 (n = 30)

Systolic (mmHg)

   Auscultatory

-16.0 ± 8.1 (+61.8 to -93.8)

0.1 ± 3.8 (+40.8 to -40.7)

   Oscillometric

-16.7 ± 7.3 (+53.0 to -86.4)

-13.3 ± 3.4 (+23.1 to -49.6)

Diastolic (mmHg)

   Auscultatory

10.0 ± 3.3 (+41.6 to -21.5)

11.9 ± 2.9 (+43.7 to -20.0)

   Oscillometric

1.3 ± 2.3 (+23.4 to -20.9)

-7.8 ± 2.9 (+24.2 to -39.8)

Mean (mmHg)

   Auscultatory

3.7 ± 4.7 (+48.7 to -44.4)

11.2 ± 3.0 (+43.7 to -21.2)

   Oscillometric

-2.3 ± 3.2 (+32.5 to -37.5)

-12.3 ± 1.7 (+6.7 to -31.3)

  1. Values are expressed as mean ± standard error.