Table 1 Spontaneous breathing simulation and ventilator settings

From: Imposed work of breathing during high-frequency oscillatory ventilation: a bench study

 

Newborn

Infant

Small child

Large child

Adult

Assumed weight (kg)

3.5

10

25

40

70

Spontaneous breathing simulation

     

   Respiratory rate (/min)

35

30

25

20

12

   Tidal volume (ml/kg)/peak inspiratory flow (l/min)

5/2.6

5/6.3

5/13

5/17

5/18

 

7/3.6

7/8.9

7/19

7/24

7/25

 

10/5.1

10/12

10/27

10/34

10/36

   Inspiratory/expiratory ratio

1:2

    

Ventilator settings

     

   HFOV ventilator

3100 A

3100 A

3100 B

3100 B

3100 B

   Tube inner diameter (mm)

3.0, 3.5, 4.0

4.0, 4.5, 5.0

5.5, 6.0, 6.5

6.5, 7.0, 7.5

7.5, 8.0, 8.5

   Fresh gas flow rate (l/min)

15/20

20/40

20/60

20/60

20/60

   Continuous distending pressure (cmH2O)

18

25

25

25

25

   Proximal pressure amplitude (cmH2O)

35

50

50

50

50

   Oscillation frequency (Hz)

10

8

8

6

6

  1. HFOV ventilator, SensorMedics 3100 A or SensorMedics 3100 B high-frequency oscillatory ventilation ventilator.