From: Clinical review: Brain-body temperature differences in adults with severe traumatic brain injury
Reference | Study design, setting, country | Population and sample size | Age range (years) and GCS score | Brain device and method | Core device and method | Calibration | Deliberate thermoregulatory interventions | Results (°C) |
---|---|---|---|---|---|---|---|---|
Henker et al. (1998) [16] | Descriptive comparison ICU USA | Severe head injury n = 8 | Age: 18-58 GCS: 4-7 | Site: right lateral ventricle Device: type T thermocouple (model no. 12983 Ventricular Catheter with Thermocouple; Radionics, Burlington, MA, USA) | Site 1: rectum Device: Hewlett-Packard Model 21090A probe (Hewlett-Packard Company, Androver, MA, USA) Site 2: bladder Device: bladder catheter (Bardex Lubricath, model 909116; CR Bard, Inc., Convington, GA, USA) | Yes | Random assignment to groups Hypothermia ('target' 32°C for 48 h, n = 5) Normothermia group (n = 3) | • Mean (SD) of Tbr - Tr: 0.8 (0.8) °C • Mean (SD) of Tbr - Tbl: 1.0 (0.7) °C Brain temperature is higher than rectal and higher than bladder temperature Not possible to determine which patients reached the hypothermia 'target' |
Rumana et al. (1998) [18] | Descriptive study ICU USA | Severe head injury n = 30 | Mean (SD) age: 30.7 (13.0) GCS: ≤8 (n = 25); >8 (n = 5) | Site: intraparenchyma Device: Licox, Kiel-Mielkendorf, Germany | Site: rectum Device: 10-Fr thermistor probe (400 series, Sheridan Catheter Corporation, Argyle, NY, USA) | Yes | No | • Mean (SD) of Tbr- Tr: 1.1 (0.6) °C Brain temperature is higher than rectal temperature |
Zhang et al. (2002) [21] | Descriptive study Unclear China | Severe head injury n = 18 | Age: 20-73 GCS: ≤8 | Site: frontal white matter Device: flexible Clark-type microcatheter (Licox-II system, GSM Co. Ltd, Germany) | Site: rectum Device: not reported | Unclear | Mild hypothermia on all patients | • Mean of Tbr - Tr at 0 hours after cooling: 0.8°C • Mean of Tbr - Tr at 24 hours after cooling: 1.1°C • Mean of Tbr - Tr at 72 hours after cooling: 1.4°C Brain temperature is consistently higher at 0, 24 and 72 hours after cooling |
Soukup et al. (2002) [12] | Descriptive study ICU USA | Severe head injury n = 58 | Age: ≥ 16 GCS: ≤8 | Site: parenchyma Device: Paratrend 7, Diametrics Medical Inc., Roseville, MN, USA | Site: rectum Device: Hewlett Packard | Unclear | Mild hypothermia (n = 33); Control (n = 25) | • Mean (SD) of Tbr -Tr in the normothermic group: 0.0 (0.5)°C • Mean (SD) of Tbr-Tr in the hyperthermic group: 0.3 (0.5)°C • Mean (SD) of Tbr-Tr in the therapeutic 'cooled' group: -0.2 (0.6)°C • Mean (SD) of Tbr-Tr of spontaneous hypothermia group: -0.8 (1.4)C |
 |  |  |  |  |  |  |  | Brain temperature is higher than rectal temperature at normothermia and hyperthermia conditions, but lower during therapeutic cooling and spontaneous hypothermia |
Tokutomi et al. (2003) [13] | Descriptive study ICU Japan | Severe brain injury n = 31 | Age: 15-69 GCS: 3-5 | Site: subdural Device: Camino Direct Pressure monitor (Camino Laboratories, San Diego, CA, USA) | Site: rectum Device: Bi-Temp (RSP, Boston, MA, USA) Site 2: Jugular vein Device: fibreoptic catheter (manufacturer type not stated) | Unclear | Hypothermia therapy on all patients | • Mean (SD) of Tbr- Tr: 0.5 (0.3) °C • Mean (SD) of Tbr- Tjv: 0.3(0.3) °C Brain temperature is higher than both rectal and jugular vein temperature |
Childs et al. (2005) [19] | Descriptive study ICU UK | Severe TBI n = 19 | Age: 19-70 GCS: 3-13 | Site: brain parenchyma Device: Camino 110-4BBT, 4 Fr, fibre-optic, transducer tipped sensor (Integra Neurosciences, Andover, UK) | Site: rectum Device: Mon-a- therm 400 series, 9 Fr, thermistor (Mallinckrodt Medical, Tyco Healthcare, Gosport, UK) | Yes | No | • Mean (SD) of Tbr- Tr: -0.04 (0.9) °C No evidence of systematic difference between brain and core temperature. Brain temperature is lower than rectal temperature |
Kirk et al. (2009) [14] | Descriptive study ICU UK | Severe TBI n = 20 | Age: 17-76 GCS: ≤8 | Site: intraparenchyma Device: Neurovent-PTemp™, Raumedic AG, Münchberg, Germany | Site 1: tympanic membrane Device: Core-Check thermometer (model 2090 IVAC Corporation, San Diego, California, USA) Site 2: temporal artery, Device: model TAT5000, Exergen, Watertown, MA, USA | Unclear | No | • Mean (SD) of Tbr - Ttm: 0.9 (0.7) °C • Mean (SD) of Tbr - Tta: 0.3 (0.4) °C Brain temperature is higher than both tympanic membrane and temporal artery temperature Temporal artery temperature is closer to brain than tympanic membrane temperature is |
Puccio et al. (2009) [17] | Descriptive study Neurotrauma ICU USA | Severe TBI n = 42 | Mean (SD) Age: 36.4 (14.8) GCS: 3-8 | Site: intraventricular Device: Licox® (Integra LifeSciences Corp., Plainsboro, NJ, USA) | Site: rectum Device: not reported | Unclear | Induced normothermia (n = 21) Control (n = 21) | • Mean of Tbr - Tr in the control group: -0.4°C • Mean of Tbr - Tr in the induced normothermic group: -0.1°C Brain temperature is lower than rectal temperature in both groups of patients |
Kuo et al. (2011) [15] | Descriptive study ICU Taiwan | TBI n = 28 | Age: 16-80 GCS: 6 (median) | Site: intraparenchyma Device: 110-4BT, Pressure-Temperature Monitoring Kit (Integra Camino, San Diego, CA, USA) | Site 1: rectum Device: Medi-Therm II Hyper-Hypothermia Machine (Gaymar Industries, Orchard Park, NY, USA) Site 2: temporal artery Device: Temporal Scanner™ TAT-5000 (EXERGEN, Boston, MA, USA) | Yes | No | • Mean (SD) of Tbr - Tr: 0.23 (0.45) °C • Mean (SD) of Tbr - Tta: 0.64 (0.60) °C Brain temperature is higher than both rectal temperature and temporal artery temperature |
Suehiro et al. (2011) [20] | Descriptive study Did not specify Japan | Severe TBI n = 11 | Age: 15-73 GCS: ≤8 | Site: intraparenchyma Device: Camino Laboratories, San Diego, California, USA | Site: bladder Device: thermocouple (Terumo, Tokyo, Japan) | Yes | Mild hypothermia on all patients | • Mean (SD) of Tbr - Tbl: -0.17 (0.02) °C Brain temperature is lower than bladder temperature |
Fischer et al. (2011) [11] | Descriptive study Neuro-ICU Austria | Severe TBI n = 7 | Age: 21-69 GCS: 3-8 | Site: intraparenchyma Device: Neurovent-Temp-P (Raumedic AG, Muenchberg, Germany) | Site: bladder Device: Kendall Curity, Mansfield, Massachusetts, USA | Unclear | Induced normothermia (36.5°) on all patients | • Mean of Tbr - Tbl: 0.1°C Brain temperature is higher than bladder temperature during induced normothermia |