Table 3 Bland-Altman analysis for the comparison between cardiac index measured by the Nexfin device and transpulmonary thermodilution depending upon the absence of atrial fibrillation and norepinephrine infusion.

From: The estimation of cardiac output by the Nexfin device is of poor reliability for tracking the effects of a fluid challenge

Subgroup

Bias (L/min/m2)

Upper limit of agreement (L/min/m2)

Lower limit of agreement (L/min/m2)

Percentage error

Whole population (n = 38)

0.2

-1.8

2.2

57%

Patients without atrial fibrillation (n = 33)

0.3

-1.8

2.4

57%

Patients without norepinephrine (n = 21)

0.3

-1.8

2.3

62%