Skip to main content

Table 1 Examples of available methods to measure cardiac output

From: Clinical review: Update on hemodynamic monitoring - a consensus of 16

Method

System

Limitations

Thermodilution

PAC

Invasiveness - training required

Transpulmonary indicator dilution

PiCCO®

Decreased accuracy?

  

Need for dedicated arterial catheter

 

LiDCO™

Decreased accuracy?

  

Need for lithium injection

  

Interference by non-depolarizing muscle relaxants; inaccurate in case of hyponatremia

 

COstatus®

Decreased accuracy?

 

VolumeView™

Decreased accuracy?

  

Need for dedicated arterial catheter

Arterial-pressure waveform-derived

PiCCO®, LiDCO™, Vigileo™, MostCare™

Decreased accuracy, need for optimal arterial pressure tracing

Esophageal Doppler

CardioQ™, WAKIe TO

Training required, intermittent measurement

Suprasternal Doppler

USCOM®

Difficult in some patients

Echocardiography

Vivid™, Sonosite MicroMaxx®, Philips CX50™, and so on

Training required, intermittent measurement

Partial CO2 rebreathing

NiCO®

Less reliable in respiratory failure

Bioimpedance

Lifegard®, TEBCO®, Hotman®, BioZ®, and so on

Less reliable in critically ill patients, not applicable in cardiothoracic surgery

Bioreactance

NICOM®

Validated in only one study in critically ill patients

  1. PAC, pulmonary artery catheter.