From: Alarms in the intensive care unit: how can the number of false alarms be reduced?
Type of alarm | Alarm category | Note | Standard |
---|---|---|---|
Electric or pneumatic failure | High priority | Â | EN 794-1 [12] |
FIO2 high or low | At least medium priority | Is applicable as soon as O2 concentration is | EN 794-1 |
 |  | different from that of ambient air |  |
Paw high | High priority | Â | EN 794-1 |
VE low* or VT low* | At least medium priority | Â | EN 794-1 |
Apnoea | At least medium priority | Â | EN 794-1 |
Disconnection | At least medium priority | Could be detected for example from a low Paw, | EN 794-1 |
 |  | a low ETCO2 and a low tidal volume |  |
Continuous pressure | High priority | Is relative to a continuous pressure kept over a | EN 794-1 |
 |  | given limit during more than 15 ± 1.5 s |  |
ETCO2 | Â | Â | Â |
   High | Medium priority |  | EN 864 [14] |
   Low | Medium priority |  | EN 864 |
FICO2 high | Medium priority | Â | EN 864 |
SpO2 | Â | Â | Â |
   High | No priority indicated | For neonatology | EN 865 [15] |
   Low | No priority indicated |  | EN 865 |
   Sensor failure | Low or medium priority |  | EN 865 |