Skip to main content

Heat and moisture exchangers (HMEs) and heated humidifiers (HHs) in adult critically ill patients: a systematic review, meta-analysis and meta-regression of randomized controlled trials

Abstract

Background

The aims of this systematic review and meta-analysis of randomized controlled trials are to evaluate the effects of active heated humidifiers (HHs) and moisture exchangers (HMEs) in preventing artificial airway occlusion and pneumonia, and on mortality in adult critically ill patients. In addition, we planned to perform a meta-regression analysis to evaluate the relationship between the incidence of artificial airway occlusion, pneumonia and mortality and clinical features of adult critically ill patients.

Methods

Computerized databases were searched for randomized controlled trials (RCTs) comparing HHs and HMEs and reporting artificial airway occlusion, pneumonia and mortality as predefined outcomes. Relative risk (RR), 95% confidence interval for each outcome and I 2 were estimated for each outcome. Furthermore, weighted random-effect meta-regression analysis was performed to test the relationship between the effect size on each considered outcome and covariates.

Results

Eighteen RCTs and 2442 adult critically ill patients were included in the analysis. The incidence of artificial airway occlusion (RR = 1.853; 95% CI 0.792–4.338), pneumonia (RR = 932; 95% CI 0.730–1.190) and mortality (RR = 1.023; 95% CI 0.878–1.192) were not different in patients treated with HMEs and HHs. However, in the subgroup analyses the incidence of airway occlusion was higher in HMEs compared with HHs with non-heated wire (RR = 3.776; 95% CI 1.560–9.143). According to the meta-regression, the effect size in the treatment group on artificial airway occlusion was influenced by the percentage of patients with pneumonia (β = -0.058; p = 0.027; favors HMEs in studies with high prevalence of pneumonia), and a trend was observed for an effect of the duration of mechanical ventilation (MV) (β = -0.108; p = 0.054; favors HMEs in studies with longer MV time).

Conclusions

In this meta-analysis we found no superiority of HMEs and HHs, in terms of artificial airway occlusion, pneumonia and mortality. A trend favoring HMEs was observed in studies including a high percentage of patients with pneumonia diagnosis at admission and those with prolonged MV. However, the choice of humidifiers should be made according to the clinical context, trying to avoid possible complications and reaching the appropriate performance at lower costs.

Background

Mechanical ventilation (MV) suppresses the mechanisms that heat and moisturize inhaled air. As a consequence, the lack of adequate conditioning may thicken airway secretions, which increases the airway resistance, reduces the gas exchange effectiveness and increases the risk of respiratory infections [1]. For these reasons, gas delivered during MV must be warmed and humidified to avoid serious complications related to dry gases [2]. To date, humidification devices can be divided into active heated humidifiers (HHs), which are devices heated by warm water, and passive devices such as heat and moisture exchangers (HMEs), which capture the heat of exhaled air and release it at the next inspiration [3]. HHs may result in increased airway hydration, decreased incidence of bacterial infection and work of breathing, while HMEs may increase the risk of airway occlusion [4]. In clinical practice, humidification during MV is widely accepted and applied; however, there is lack of consensus on the optimal device to humidify the airways. The aims of this systematic review and meta-analysis of randomized controlled trials are to evaluate the effects of HMEs and HHs in preventing artificial airway occlusion and pneumonia, and on mortality in adult critically ill patients. We planned a priori a sub-analysis stratifying the studies according to the type of HH, hypothesizing that HHs with heated wire could perform differently from those with non-heated wire. In addition, we planned to perform a meta-regression analysis to evaluate the relationship between the incidence of artificial airway occlusion, pneumonia and mortality and clinical features of adult critically ill patients.

Methods

Data sources and searches

We aimed to identify all randomized controlled trials (RCTs) comparing HMEs and HHs in adult critically ill patients. We applied standard filters for the identification of RCTs using the MEDLINE and PUBMED search engines (from inception to June 2014), using English language restrictions. Our search included the following keywords: heat and moisture exchangers, heated humidifiers, airway humidification, artificial humidification, artificial airway occlusion, mortality, pneumonia and humans and randomized clinical trial.

Selection of studies

Trials comparing any type of HH, including systems with heated and non-heated wire, with HMEs in adult critically ill patients were included. We restricted the analysis to RCTs to guarantee control of selection bias. We included only published full papers and excluded abstracts. Study designs containing inadequately adjusted planned co-interventions and crossover trials were excluded. The intervention of interest was the use of HH and HME in reducing artificial airway occlusion, pneumonia and mortality. Studies were further divided according to the use of HH with heated and HH with non-heated wire to perform the subgroup analysis.

Outcome measures

The primary outcome was the incidence of artificial airway occlusion; the secondary outcomes were the incidences of pneumonia and mortality.

Data extraction and quality assessment

Initial selection was performed by two pairs of independent reviewers (MV and DC, PP and YS) screening titles and abstracts. For detailed evaluation, a full-text copy of all studies of possible relevance was retrieved. Data from each study were extracted independently by paired and independent reviewers (LB and DC, PP and YS) using a standardized data abstraction form. Data extracted from the publications were independently checked for accuracy by two other reviewers (GS and AE). Quality assessment of these studies included: (1) use of randomization sequence generation, (2) reporting and type of allocation concealment, (3) blinding, (4) reporting of incomplete outcome data and (5) comparability of the groups at baseline. Quality assessment is reported in Additional file 1. Two reviewers (MV and LB) independently used these criteria to evaluate trial quality. We solved any possible disagreement by consensus in consultation with two other reviewers (GS and AE) if needed.

Qualitative analysis

A narrative summary approach was used to explore study characteristics and quality indicators in describing variation among studies and to consider possible implications for this in our understanding of the outcomes of the RCTs included in the Cochrane review [5, 6].

Quantitative analysis

The meta-analysis was conducted according to the Preferred Reporting Items for Systematic Reviews and Meta-analyses (PRISMA) guidelines [7]. Meta-analysis was performed with mixed random effect using the DerSimonian and Laird method. Results were graphically represented using forest plot graphs. The relative risk (RR) and 95% CI for each outcome were separately calculated for each trial, pooling data when needed, according to an intention-to-treat principle. The choice to use RRs was driven by the design of meta-analysis based on RCTs. Tau2 was used to define the variance between studies. The difference in the estimates of treatment effect between the treatment groups for each hypothesis was tested using the two-sided z test with statistical significance considered at p value <0.05. The homogeneity assumption was checked with the Q test with a degree of freedom (df) equal to the number of analyzed studies minus 1. The heterogeneity was measured by the I 2 metric, which describes the percentage of total variation across studies that is due to heterogeneity rather than chance. I 2 was calculated as:

I 2 = 100% Å ~ (Q − df)/Q

where Q is Cochran’s heterogeneity statistic and df is degrees of freedom. A value of 0% indicates no observed heterogeneity, and larger values show increasing heterogeneity. We decided a priori to analyze all the outcomes according to the following categories when possible: HME vs HH, HME vs HH with heated wire and HME vs HH with non-heated wire.

Weighted random-effect meta-regression analysis was performed to test the relationship between the effect size on each considered outcome and the following covariates, with each one analysed separately: duration of MV, pneumonia incidence, intensive care unit (ICU) length of stay, percentage of respiratory diagnoses at ICU admission, simplified acute physiology score (SAPS), age and acute physiology and chronic health evaluation II (APACHE II) score.

Analyses were conducted with OpenMetaAnalyst (version 6) and SPSS version 20 (IBM SPSS). Weighted linear regression was used to evaluate potential publication bias, with the natural log of the RR as the dependent variable and the inverse of the total sample size as the independent variable. This is a modified Macaskill test that gives more balanced type I error rates in the tail probability areas in comparison to other publication bias tests [8].

Results

Study selection

We identified 1349 references and excluded 1266 after screening titles and abstracts. We analyzed 77 articles in full paper format. We excluded 59 references [9,10,11,12,13,14,15,16,17,18,19,20,21,22,23,24,25,26,27,28,29,30,31,32,33,34,35,36,37,38,39,40,41,42,43,44,45,46,47,48,49,50,51,52,53,54,55,56,57,58,59,60,61,62,63,64,65] and 18 references fulfilled our search criteria [66,67,68,69,70,71,72,73,74,75,76,77,78,79,80,81,82,83]. Figure 1 shows the study selection process.

Fig. 1
figure 1

The study selection process

Characteristics of included studies

These 18 RCTs included 2442 adult critically ill patients. The main characteristics of the included studies are reported in Table 1.

Table 1 Main characteristics of the randomized controlled trials included in the meta-analysis (HME vs HH with heated and non-heated wire)

Systematic errors of included studies

None of the included trials had a low risk of bias. The random sequence generation was adequate in seven studies [75, 78,79,80,81,82,83], adequate allocation concealment was present in three studies [77,78,79] and adequate blinding was present in two studies [75, 76]. Complete outcome data were reported from 10 studies [66,67,68,69,70, 73,74,75,76, 80,81,82,83]. Ten studies reported no imbalance in baseline characteristics [66,67,68,69,70,71, 73, 76, 79, 81]. The quality assessment for each RCT is reported in Additional file 1.

Primary outcome

The incidence of artificial airway occlusion was not different in patients treated with HMEs and HHs (Fig. 2) (RR = 1.853; 95% CI 0.792–4.338). Artificial airway occlusion incidence was not different when comparing HMEs with HHs with heated wire (RR = 0.379; 95% CI 0.140–1.384) (Fig. 3, upper panel). However, airway occlusion was higher with HMEs compared with HHs with non-heated wire (RR = 3.776; 95% CI 1.560–9.143) (Fig. 3, lower panel), but there were no differences between hydrophobic and hygroscopic HMEs compared with HHs (Additional file 2).

Fig. 2
figure 2

Artificial airway occlusion comparing the heat and moisture exchanger (HME) with the heated humidifier (HH). Weights: Kirkegaard 6.8%, Martin 7.1%, Misset 15.3%, Roustan 7.1%, Dreyfuss 5.9%, Branson 4.2%, Villafane 10.7%, Boots (2006) 4.2%, Hurni 5.9%, Kirton 6.6%, Kollef 4.1%, Lucchetti 6.9%, Lacherade 10.8%, boots (1997) 4.1%

Fig. 3
figure 3

Upper box artificial airway occlusion comparing the heat and moisture exchanger (HME) and the heated humidifier (HH) with heated wire. Weights: Branson 11%, Boots (1997) 11%, Kirton 19.2%, Kollef 11%, Lacherade 36.8% Boots (2006) 10.9%. Lower box artificial airway occlusion comparing HME and HH with non-heated wire. Weights: Kirkegaard 8.9%, Martin 9.7%, Misset 30%, Roustan 9.8%, Dreyfuss 7.7%, Villafane 16.8%, Hurni 7.7%, Lucchetti 9.3%

Secondary outcomes

The incidence of pneumonia was not different in patients treated with HMEs and HHs (Fig. 4) (RR = 932; 95% CI 0.730–1.190). Incidence of pneumonia was not different when comparing HMEs and HHs with heated wire (RR = 0.997; 95% CI 0.642–1.548), with significant inhomogeneity (I 2 = 54%; p = 0.042) (Fig. 5, upper panel), neither was it different with HHs with non-heated wire (RR = 0.756; 95% CI 0.479–1.193) (Fig. 5, lower panel).

Fig. 4
figure 4

Incidence of pneumonia comparing the heat and moisture exchanger (HME) with the heated humidifier (HH). Weights: Martin 2.5%, Roustan 4.7%, Dreyfuss 5%, Branson 2.3%, Boots (1997) 5%, Kirton 8.6%, Kollef 9.1%, Memish 9.4%, Diaz 5.5%, Lacherade 19.5%, Lorente 8.4%, Boots (2006) 14.3%, Oguz 5.1%

Fig. 5
figure 5

Upper panel: incidence of pneumonia comparing the heat and moisture exchanger (HME) and the heated humidifier (HH) with heated wire. Weights: Branson 5.7%, Boots (1997) 9.3%, Kirton 15%, Kollef 15.7%, Lachede 22%, Lorente 12.6%, Boots (2006) 19.7%. Lower panel incidence of pneumonia comparing HME and HH with non-heated wire. Weights: Martin 7.8%, Roustan 15.5%, Dreyfuss 16.7%, Memish 37.9%, Diaz 11.8%, Oguz 10.2%

Mortality was not different in patients treated with HMEs and HHs (Fig. 6, upper panel) (RR = 1.023; 95% CI 0.878–1.192). Mortality was comparable in patients treated with HMEs and HHs with heated wire (RR = 0.947; 95% CI 0.723–1.241) (Fig. 6, middle panel). We did not find differences in mortality when comparing HMEs and HHs with non-heated wire (RR = 1.186; 95% CI 0.852–1.650) (Fig. 6, lower panel).

Fig. 6
figure 6

Upper panel mortality comparing the heat and moisture exchanger (HME) with the heated humidifier (HH). Weights: Kirkegaard1%, Martin 3.4%, Roustan 4.6%, Dreyfuss 5.4%, Boots (1997) 1.6%, Hurni 7.9%, Kollef 16.1%, Memish 14.5%, Diaz 2.6%, Lacherade 28.3%, Boots (2006) 14.4%. Middle panel mortality comparing HME and HH with heated wire. Weights: Boots (1997) 4%, Boots (2006) 28.9%, Kollef 27.9%, Lacherade 39.2%. Lower box mortality comparing HME and HH with non-heated wire. Weights: Diaz 6.2%, Dreyfuss 15.6%, Hurni 17.51%, Kirkegaard 3.1%, Martin 9.2%, Memish 34.9%, Roustan 13.5%

Meta-regression analysis

The effect size in the treatment group on artificial airway occlusion was influenced by the percentage of patients with pneumonia included in the study (β = -0.058; p = 0.027; favoring HMEs in studies with high prevalence of pneumonia), and a trend was observed for the duration of MV (β = -0.108; p = 0.054; favoring HMEs in studies with longer MV time) (Fig. 7). No other significant associations with the effect size on any outcome measure were observed for the other clinical variables (see Additional file 3).

Fig. 7
figure 7

Meta-regression on artificial airway occlusion including duration of mechanical ventilation (β = 2.637; p = 0.054) and pneumonia (β = 1.794; p = 0.012) as covariate. HME heat and moisture exchanger, HH heated humidifier

Discussion

In this systematic review and meta-analysis, we found: (1) no significant difference in artificial airway occlusion, pneumonia or mortality between HMEs and HHs, (2) no effect of HHs with and without heated wire compared to HMEs; however HHs with non-heated wire had the lower RR for artificial airway occlusion compared with HME, and (3) independently from the HH type, an advantage of HMEs in airway occlusion incidence was observed in studies with high incidence of pneumonia, and a trend toward favoring HMEs was observed for studies with prolonged MV. To our knowledge, this is the first systematic review performed (1) by dividing RCTs according to HHs with heated and non-heated wire and (2) including a meta-regression analysis on the potential effects of clinical variables on the efficacy of the two devices.

According to the American Association for Respiratory Care (AARC) guidelines, HHs should provide an absolute humidity level of between 33 and 44 mgH2O/L, whereas HMEs should provide a minimum humidity level of 30 mgH2O/L [1]. HHs may produce insufficient heat and humidification when the temperature is improperly selected or pre-set at a non-adjustable level rather than at the clinical setting [1]. However, insufficient heat and humidification may occur with HMEs too [1]. Only 37% of HMEs have been found to meet the standard criteria advocated by the AARC guidelines [1]. Insufficient airway humidification may lead to an increase in tracheal tube occlusion, a serious adverse event that may occur in mechanically ventilated patients and requires timely intervention. In an RCT comparing HMEs and HHs with increasing minute ventilation, the authors found that after 72 hours the inner diameter of the endotracheal tube decreased by 2.5–6.5 mm when gas conditioning was performed using HMEs and by 1.5 mm with HHs [72]. A systematic review showed that in patients ventilated more than 48 hours, there is no difference in tracheal tube occlusion when comparing HMEs and HHs [84]. In this meta-analysis, we found no differences in the incidence of artificial airway occlusion, but stratifying the comparison according to the type of HHs we found less risk for airway occlusion in HHs without heated wire compared with HME. However, these data were not confirmed by the sub-analysis comparing hydrophobic and hygroscopic HMEs with HHs. Probably, the main determinants of artificial airway occlusion are the duration of mechanical ventilation and pneumonia, rather than humidifier type per se, even if a prolonged use of HME (<72 hours) may increase this risk. Long-term invasive MV and the presence of ventilator-associated pneumonia (VAP) increased the risk of artificial airway occlusion threefold in one study [85] and twofold in another [86]. This is the first meta-analysis reporting a meta-regression of included studies in this field. Our meta-regression showed that in studies with high incidence of pneumonia and prolonged MV, the HMEs had a slight advantage in terms of the artificial airway occlusion.

Earlier models of HME were associated with an increased incidence of airway occlusion, which led to the exclusion of patients at high risk from the studies [86]. In contrast, trials using HMEs with enhanced intrinsic humidifying performance showed no difference in the incidence of airway occlusion [86]. A Cochrane review states that hydrophobic HMEs may reduce the risk of pneumonia and the use of an HME may increase artificial airway occlusion in certain subgroups of patients [4]. Our analysis includes more recently published studies. According to our meta-regression, the HME may reduce the risk of airway occlusion in selected patients affected by pneumonia.

Hess et al. concluded in their clinical practice guidelines that HMEs are associated with lower incidence of pneumonia compared with HHs [85]. However, there are concerns about the increased airway resistance and care of HME filters [85]. Kola et al. found a significant reduction in pneumonia using HMEs during MV, particularly when patients are ventilated for 7 days or more [86]. Hess et al. included studies published between 1990 and 1998 in their analysis of pneumonia [85]. Kola et al. reported the same results as Hess et al. but they only included one more study in their meta-analysis. Accordingly, the underlying mechanism of reduction in pneumonia may be due to the dryness of the ventilator circuit when using HMEs [85, 86]. Therefore, HMEs minimized the need for septic manipulations or aspirations of the airway/circuit and the circuit condensate [85, 86]. Furthermore, the inclusion of more recent studies may have changed the results. Indeed, Siempos et al. did not find any superiority of HMEs compared to HHs in reducing pneumonia, mortality or morbidity [87]. The results of Siempos et al. were groundbreaking and in line with the RCTs published at that time. The inclusion of three further RCTs [80,81,82] with 870 patients dramatically changed the previous results. The Cochrane review by Kelly et al. included adult and pediatric patients treated with HMEs and HHs [4]. There was no overall effect on artificial airway occlusion, mortality, pneumonia or respiratory complications; however, the arterial partial pressure of carbon dioxide (PaCO2) and minute ventilation were increased while body temperature was lower when HMEs were compared to HHs [4] in a meta-analysis including 18 RCTs and 2442 adult critically ill patients. In line with the available literature, we did not find any difference in artificial airway occlusion, pneumonia or mortality between HMEs and HHs, even if according to their nature, they have different characteristics. However, HMEs were found to increase the PaCO2 and work of breathing probably due to higher dead space, and to reduce the inner diameter of the endotracheal tube during prolonged MV [4]. Indeed, HME may negatively impact on ventilator function while increasing the dead space [11]. In spontaneously and assisted breathing patients, this requires increased minute ventilation and then the work of breathing, to maintain constant alveolar ventilation and PaCO2. In controlled MV, the additional dead space of HMEs may reduce alveolar ventilation and increase PaCO2 [11]. This effect of HME dead space may be further exacerbated by protective ventilation at low tidal volume (VT) and by using HMEs with a larger dead space [11].

Humidification is mandatory during MV. Nowadays, the airway humidification is appropriate in the absence of any contraindications listed by the AARC guidelines, such as altered body temperature, airway thermal injury, under hydrated secretions, increased work of breathing, hypoventilation, condensation and airway dehydration [1]. Clear advantages in terms of clinical outcomes for different humidification devices are far from being demonstrated. The present meta-analysis reported no superiority of HMEs over HHs in term of clinical outcomes, with similar results even when stratifying the studies according to the type of HH, while some advantage of HMEs might be possible in patients with pneumonia or those with a long MV time. The choice of humidifiers should be made according to the clinical context trying to avoid possible complications and reaching the appropriate performance at lower costs. However, to help clinicians make the correct choice between HHs and HMEs, further high-quality RCTs are needed to evaluate the incidence of respiratory complications other than pneumonia, gas exchange and work of breathing when comparing the HH and HME devices.

This systematic review and meta-analysis has several limitations that must be addressed. First, the quality of the included RCTs was relatively low and our conclusions may be limited by this point. Second, the diagnosis of pneumonia was differently defined across the studies and often mixed with VAP. Third, the definition of mortality varied between the studies: three studies reported the ICU mortality, two studies reported overall mortality, two studies reported hospital mortality, one studies reported mortality during MV and in four studies mortality was not reported. Fourth, we performed the meta-analyses of outcomes if reported by more than three RCTs. Fifth, as we found only one additional RCT published between 2006 and 2013, the present results may depend on the studies published before 2006. However, in contrast to previous reports we included a meta-regression analyzing and interpreting data from a new point of view. Sixth, we found few studies reporting the effective tracheal tube lumen and most of them with provided a poor definition of pneumonia diagnosed at ICU admission. Seventh, we are not able to stratify the meta-analysis according to the baseline respiratory condition or inclusion criteria. This did not allow us to suggest the use of HMEs or HHs in different respiratory diseases.

New, prospective RCTs are needed in terms of assessing the effect of HHs vs HMEs in patients with respiratory failure due to pneumonia, on airway diameter, the amount of secretion and the occurrence of artificial airway obstruction and VAP.

Conclusions

In this meta-analysis including 18 RCTs and 2442 adult critically ill patients, we found no superiority of HMEs or HHs, in terms of artificial airway occlusion, pneumonia and mortality. These results were also confirmed in the sub-analysis dividing HHs into heated and non-heated wire devices. However, HHs with non-heated wire had the lower RR for artificial airway occlusion compared with HMEs. A trend favoring HMEs was observed in studies including a high percentage of patients with pneumonia diagnosis at admission and those with prolonged MV. However, the choice of humidifiers should be made according to the clinical context, trying to avoid possible complications and reaching the appropriate performance at lower costs.

Abbreviations

AARC:

American Association for Respiratory Care

APACHE II:

Acute physiology and chronic health evaluation II

CI:

Confidence interval

HHs:

Heated humidifiers

HMEs:

Heat and moisture exchangers

ICU:

Intensive care unit

MV:

Mechanical ventilation

PaCO2 :

Arterial partial pressure of carbon dioxide

RCTs:

Randomized controlled trials

RR:

Relative risk

SAPS:

Simplified acute physiology score

VAP:

Ventilator-associated pneumonia

References

  1. American Association for Respiratory Care, Restrepo RD, Walsh BK. Humidification during invasive and noninvasive mechanical ventilation: 2012. Respir Care. 2012;57(5):782–8.

    Article  Google Scholar 

  2. Branson RD, Chatburn RL. Humidification of inspired gases during mechanical ventilation. Respir Care. 1993;38:461–8.

    Google Scholar 

  3. Lellouche F, Qader S, Taille S, Lyazidi A, Brochard L. Influence of ambient temperature and moisture ventilation on passive and active heat and moisture exchange. Respir Care. 2014;59(5):637–43.

    Article  PubMed  Google Scholar 

  4. Kelly M, Gillies D, Todd DA, Lockwood C. Heated humidification versus heat and moisture exchangers for ventilated adults and children. Cochrane Database Syst Rev. 2010;4:CD004711. doi:10.1002/14651858.CD004711.pub2.

  5. Dixon-Woods M, Shaw RL, Agarwal S, Smith JA. The problem of appraising qualitative research. Qual Saf Health Care. 2004;13:223–5.

    Article  CAS  PubMed  PubMed Central  Google Scholar 

  6. Denzin N, Riessman C. Narrative analysis. J Commun. 1995;45:177–84.

    Article  Google Scholar 

  7. PRISMA guideline. http://www.prisma-statement.org/

  8. Peters JL, Sutton AJ, Jones DR, Abrams KR, Rushton L. Comparison of two methods to detect publication bias in meta-analysis. JAMA. 2006;295(6):676–80.

    Article  CAS  PubMed  Google Scholar 

  9. Bissonnette B. Passive or active inspired gas humidification increases thermal steady-state temperatures in anesthetized infants. Anesth Analg. 1989;69(6):783–7.

    Article  CAS  PubMed  Google Scholar 

  10. Bissonette B, Sessler DI, La Flamme P. Passive and active inspired gas humidification in infants and children. Anesthesiology. 1989;71(3):350–4.

    Article  Google Scholar 

  11. Campbell RS, Davis Jr K, Johannigman JA, Branson RD. The effects of passive humidifier dead space on respiratory variables in paralyzed and spontaneously breathing patients. Respir Care. 2000;45(3):306–12.

    CAS  PubMed  Google Scholar 

  12. Deriaz H, Fiez N, Lienhart A. Comparative effects of a hygrophobic filter and a heated humidifier on intraoperative hypothermia. Ann Fr Anesth Reanim. 1992;11(2):145–9.

    Article  CAS  PubMed  Google Scholar 

  13. Girault C, Breton L, Richard J, Tamion F, Vandelet P, Aboab J, et al. Mechanical effects of airway humidification devices in difficult to wean patients. Crit Care Med. 2003;31(5):1306–11.

    Article  PubMed  Google Scholar 

  14. Goldberg ME, Epstein R, Rosenblum F, Larijani GE, Marr A, Lessin J, et al. Do heated humidifiers and heat and moisture exchangers prevent temperature drop during lower abdominal surgery? J Clin Anesth. 1992;4(1):16–20.

    Article  CAS  PubMed  Google Scholar 

  15. Iotti GA, Olivei MC, Palo A, Galbusera C, Veronesi R, Comelli A, et al. Unfavorable mechanical effects of heat and moisture exchangers in ventilated patients. Intensive Care Med. 1997;23(4):399–405.

    Article  CAS  PubMed  Google Scholar 

  16. Le Bourdelles G, Mier L, Fiquet B, Djedaini K, Saumon G, Coste F, et al. Comparison of the effects of heat and moisture exchangers and heated humidifiers on ventilation and gas exchange during weaning trials from mechanical ventilation. Chest. 1996;110(5):1294–8.

    Article  PubMed  Google Scholar 

  17. Linko K, Honkavaara P, Nieminen MT. Heated humidification in major abdominal surgery. Eur J Anaesthesiol. 1984;1(3):285–91.

    CAS  PubMed  Google Scholar 

  18. MacIntyre NR, Anderson HR, Silver RM, Schuler FR, Coleman RE. Pulmonary function in mechanically ventilated patients during 24-hour use of a hygroscopic condensor humidifier. Chest. 1983;84(5):560–4.

    Article  CAS  PubMed  Google Scholar 

  19. Martin C, Papazian L, Perrin G, Saux P, Gouin F. Preservation of humidity and heat of respiratory gases in patients with a minute ventilation greater than 10 L/min. Crit Care Med. 1994;22(11):1871–6.

    Article  CAS  PubMed  Google Scholar 

  20. Pelosi P, Solca M, Ravagnan I, Tubiolo D, Ferrario L, Gattinoni L. Effects of heat and moisture exchangers on minute ventilation, ventilatory drive, and work of breathing during pressure-support ventilation in acute respiratory failure. Crit Care Med. 1996;24(7):1184–8.

    Article  CAS  PubMed  Google Scholar 

  21. Ricard JD, Markowicz P, Djedaini K, Mier L, Coste F, Dreyfuss D. Bedside evaluation of efficient airway humidification during mechanical ventilation of the critically ill. Chest. 1999;115(6):1646–52.

    Article  CAS  PubMed  Google Scholar 

  22. Roustan JP, Kienlen J, Aubas S, Du Cailar J. Evaluation of an exchange filter on heat and humidity in long duration mechanical ventilation. Comparison with heated humidification. Ann Fr Anesth Reanim. 1989;8 Suppl:R275.

  23. Thomachot L, Viviand X, Lagier P, Dejode JM, Albanese J, Martin C. Measurement of tracheal temperature is not a reliable index of total respiratory heat loss in mechanically ventilated patients. Crit Care. 2001;5(1):24–30.

    Article  CAS  PubMed  Google Scholar 

  24. Alagar R. Heated humidification reduces saline instillations, nebulized therapy, and cost in long term ventilated patients (abstract). American Journal of Respiratory and Critical Care Medicine. American Thoracic Society 2000 International Conference; May 5-10; Toronto, Canada. 2000; Vol. 161; 3 Suppl: A552.

  25. Branson RD, Davis Jr K, Campbell RS, Johnson DJ, Porembka DT. Humidification in the intensive care unit: prospective study of a new protocol utilizing heated humidification and a hydroscopic condenser humidifier. Chest. 1993;104(6):1800–05.

    Article  CAS  PubMed  Google Scholar 

  26. Branson RD, Campbell RS, Johannigman JA, Ottaway M, Davis Jr K, Luchette FA, et al. Comparison of conventional heated humidification with a new active hygroscopic heat and moisture exchanger in mechanically ventilated patients. Respir Care. 1999;44(8):912–17.

    Google Scholar 

  27. Christiansen S, Renzing K, Hirche H, Reidemeister JC. Measurements of inspired air humidity as provided by different humidifiers. Anasthesiol Intensivmed Notfallmed Schmerzther. 1998;33(5):300–5.

    Article  CAS  PubMed  Google Scholar 

  28. Cohen IL, Weinberg PF, Fein IA, Rowinski GS. Endotracheal tube occlusion associated with the use of heat and moisture exchangers in the intensive care unit. Crit Care Med. 1988;16(3):277–9.

    Article  CAS  PubMed  Google Scholar 

  29. Conti G, De Blasi RA, Rocco M, Pelaia P, Antonelli M, Bufi M, et al. Effects of the heat-moisture exchangers on dynamic hyperinflation of mechanically ventilated COPD patients. Intensive Care Med. 1990;16(7):441–3.

    Article  CAS  PubMed  Google Scholar 

  30. Dias MD, Pellacine EN, Zechineli CA. The new system of gaseous mixture humidification and heating of respirators: comparative study. Rev Assoc Med Bras. 1993;39(4):207–12.

    CAS  PubMed  Google Scholar 

  31. Dias MD, Pellacine EN, Zechineli CA. Bacterial aerosol generated by mechanical ventilators: comparative study. Rev Assoc Med Bras. 1997;43(1):15–20.

    Article  CAS  PubMed  Google Scholar 

  32. Fujita Y, Imanaka H, Fujino Y, Takeuchi M, Tomita T, Mashimo T, Nishimura M. Effect of humidifying devices on the measurement of tidal volume by mechanical ventilators. J Anesth. 2006;20(3):166–72.

    Article  PubMed  Google Scholar 

  33. Jaber S, Pigeot J, Fodil R, Maggiore S, Harf A, Isabey D, et al. Long-term effects of different humidification systems on endotracheal tube patency: evaluation by the acoustic reflection method. Anesthesiology. 2004;100(4):782–8.

    Article  PubMed  Google Scholar 

  34. Johnson PA, Raper RF, Fisher MDM. The impact of heat and moisture exchanging humidifiers on work of breathing. Anaesth Intensive Care. 1995;23(6):697–701.

    CAS  PubMed  Google Scholar 

  35. Kranabetter R, Leier M, Kammermeier D, Just HM, Heuser D. The effects of active and passive humidification on ventilation-associated nosocomial pneumonia. Anaesthesist. 2004;53(1):29–35.

    Article  CAS  PubMed  Google Scholar 

  36. Lellouche F, Qader S, Taille S, Lyazidi A, Brochard L. Under-humidification and over-humidification during moderate induced hypothermia with usual devices. Intensive Care Med. 2006;32(7):1014–21.

    Article  PubMed  Google Scholar 

  37. Luchetti M, Pigna A, Gentili A, Marraro G. Evaluation of the efficiency of heat and moisture exchangers during paediatric anaesthesia. Paediatr Anaesth. 1999;9(1):39–45.

    Article  CAS  PubMed  Google Scholar 

  38. MacLeod R, Bucknall T. Mechanical ventilation with heated humidifiers or with heat and moisture exchangers with microbiological filters did not reduce ventilator associated pneumonia in adults. Evid Based Nursing. 2006;9(3):82.

    Article  Google Scholar 

  39. Martin C, Papazian L, Perrin G, Bantz P, Gouin F. Performance evaluation of three vaporizing humidifiers and two heat and moisture exchangers in patients with minute ventilation > 10 L/min. Chest. 1992;102(5):1347–50.

    Article  CAS  PubMed  Google Scholar 

  40. Martin C, Thomachot L, Quinio B, Viviand X, Albanese J. Comparing two heat and moisture exchangers with one vaporizing humidifier in patients with minute ventilation greater than 10 L/min. Chest. 1995;107(5):1411–5.

    Article  CAS  PubMed  Google Scholar 

  41. McEvoy MT, Carey TJ. Shivering and rewarming after cardiac surgery: comparison of ventilator circuits with humidifier and heated wires to heat and moisture exchangers. Am J Crit Care. 1995;4(4):293–9.

    CAS  PubMed  Google Scholar 

  42. Nakagawa NK, Macchione M, Petrolino HM, Guimaraes ET, King M, Saldiva PH, et al. Effects of a heat and moisture exchanger and a heated humidifier on respiratory mucus in patients undergoing mechanical ventilation. Crit Care Med. 2000;28(2):312–7.

    Article  CAS  PubMed  Google Scholar 

  43. Prat G, Renault A, Tonnelier JM, Goetghebeur D, Oger E, Boles JM, et al. Influence of the humidification device during acute respiratory distress syndrome. Intensive Care Med. 2003;29(12):2211–5.

    Article  PubMed  Google Scholar 

  44. Prin S, Chergui K, Augarde R, Page B, Jardin F, Vieillard- Baron A. Ability and safety of a heated humidifier to control hypercapnic acidosis in severe ARDS. Intensive Care Med. 2002;28(12):1756–60.

    Article  PubMed  Google Scholar 

  45. Rathgeber J, Henze D, Zuchner K. Air conditioning with a high-performance HME (heat and moisture exchanger) − an effective and economical alternative to active humidifiers in ventilated patients. A prospective and randomized clinical study. Anaesthesist. 1996;45(6):518–25.

    Article  CAS  PubMed  Google Scholar 

  46. Rathgeber J, Betker T, Zuchner K. Measurement of water vapour pressure in the airways of mechanically ventilated patient using different types of humidifier. Anasthesiol Intensivmed Notfallmed Schmerzther. 2001;36(9):560–5.

    Article  CAS  PubMed  Google Scholar 

  47. Schiffmann H, Rethgeber J, Singer D, Harms K, Bolli A, Zuchner K. Airway humidification in mechanically ventilated neonates and infants: A comparative study of a heat and moisture exchanger vs. a heated humidifier using a new fast-response capacitive humidity sensor. Crit Care Med. 1997;25(10):1755–60.

    Article  CAS  PubMed  Google Scholar 

  48. Takumi Y, Aochi O. Effect of emphysematous expansion and humidification on the intrapulmonary gas exchange. Anaesthesist. 1970;19(10):373–83.

    CAS  PubMed  Google Scholar 

  49. Thomachot L, Viviand X, Arnaud S, Vialet R, Albanese J, Martin C. Preservation of humidity and heat of respiratory gases in spontaneously breathing, tracheostomized patients. Acta Anaesthesiol Scand. 1998;42(7):841–4.

    Article  CAS  PubMed  Google Scholar 

  50. Wilmshurst JM, Rahman MA, Shah V, Elton P, Long D, Martin N. The heat moisture exchange device (HME) in neonatal ventilation. Am J Perinatol. 1999;16(1):13–6.

    Article  CAS  PubMed  Google Scholar 

  51. Boyer A, Vargas F, Hilbert G, Gruson D, Mousset-Hovaere M, Castaing Y, Dreyfuss D, Ricard JD. Small dead space heat and moisture exchangers do not impede gas exchange during noninvasive ventilation: a comparison with a heated humidifier. Intensive Care Med. 2010;36(8):1348–54.

    Article  PubMed  Google Scholar 

  52. Arieli R, Daskalovic Y, Ertracht O, Arieli Y, Adir Y, Abramovich A, Halpern P. Flow resistance, work of breathing of humidifiers, and endotracheal tubes in the hyperbaric chamber. Am J Emerg Med. 2011;29(7):725–30.

    Article  PubMed  Google Scholar 

  53. Thille AW, Bertholon JF, Becquemin MH, Roy M, Lyazidi A, Lellouche F, Pertusini E, Boussignac G, Maître B, Brochard L. Aerosol delivery and humidification with the Boussignac continuous positive airway pressure device. Respir Care. 2011;56(10):1526–32.

    Article  PubMed  Google Scholar 

  54. Retamal J, Castillo J, Bugedo G, Bruhn A. Airway humidification practices in Chilean intensive care units. Rev Med Chil. 2012;140(11):1425–30.

    Article  PubMed  Google Scholar 

  55. Tonnelier A, Lellouche F, Bouchard PA, L’Her E. Impact of humidification and nebulization during expiratory limb protection: an experimental bench study. Respir Care. 2013;58(8):1315–22.

    Article  PubMed  Google Scholar 

  56. Schena E, Saccomandi P, Cappelli S, Silvestri S. Mechanical ventilation with heated humidifiers: measurements of condensed water mass within the breathing circuit according to ventilatory settings. Physiol Meas. 2013;34(7):813–21.

    Article  CAS  PubMed  Google Scholar 

  57. Ehrmann S, Roche-Campo F, Sferrazza Papa GF, Isabey D, Brochard L, Apiou-Sbirlea G. REVA research network. Aerosol therapy during mechanical ventilation: an international survey. Intensive Care Med. 2013;39(6):1048–56.

    Article  CAS  PubMed  Google Scholar 

  58. Alonso-Iñigo JM, Almela A, Albert A, Carratalá JM, Fas MJ. Active humidification with Boussignac CPAP: in vitro study of a new method. Respir Care. 2013;58(4):647–54.

    PubMed  Google Scholar 

  59. Schena E, Quaranta A, Saccomandi P, Silvestri S. Performances of heated humidifiers in mechanical ventilation: a preliminary intra-breath analysis. Conf Proc IEEE Eng Med Biol Soc. 2013;2013:934–7.

    CAS  PubMed  Google Scholar 

  60. Lellouche F, Pignataro C, Maggiore SM, Girou E, Deye N, Taillé S, Fischler M, Brochard L. Short-term effects of humidification devices on respiratory pattern and arterial blood gases during noninvasive ventilation. Respir Care. 2012;57(11):1879–86.

    Article  PubMed  Google Scholar 

  61. Schena E, Saccomandi P, Ramandi C, Silvestri S. A novel control strategy to improve the performances of heated wire humidifiers inartificial neonatal ventilation. Physiol Meas. 2012;33(7):1199–211.

    Article  CAS  PubMed  Google Scholar 

  62. Morán I, Cabello B, Manero E, Mancebo J. Comparison of the effects of two humidifier systems on endotracheal tube resistance. Intensive Care Med. 2011;37(11):1773–9.

    Article  PubMed  Google Scholar 

  63. Schena E, De Paolis E, Silvestri S. Gas pre-warming for improving performances of heated humidifiers in neonatal ventilation. Conf Proc IEEE Eng Med Biol Soc. 2011;2011:1205–8.

    CAS  PubMed  Google Scholar 

  64. Lorente L, Blot S, Rello J. New issues and controversies in the prevention of ventilator-associated pneumonia. Am J Respir Crit Care Med. 2010;182(7):870–6.

    Article  PubMed  Google Scholar 

  65. Verta A, Schena E, Silvestri S. Mathematical model and minimal measurement system for optimal control of heated humidifiers in neonatal ventilation. Med Eng Phys. 2010;32(5):475–81.

    Article  PubMed  Google Scholar 

  66. Kirkegaard L, Andersen BN, Jensen S. Moistening of inspired air during respirator treatment. Comparison between the water-bath evaporator and hygroscopic moisture heat exchanger. Ugeskr Laeger. 1987;149(3):152–5.

    CAS  PubMed  Google Scholar 

  67. Martin C, Perrin G, Gevaudan MJ, Saux P, Gouin F. Heat and moisture exchangers and vaporizing humidifiers in the intensive care unit. Chest. 1990;97(1):144–9.

    Article  CAS  PubMed  Google Scholar 

  68. Misset B, Escudier B, Rivara D, Leclercq B, Nitenberg G. Heat and moisture exchanger vs heated humidifier during long-term mechanical ventilation. A prospective randomized study. Chest. 1991;100(1):160–3.

    Article  CAS  PubMed  Google Scholar 

  69. Roustan JP, Kienlen J, Aubas P, Aubas S, du Cailar J. Comparison of hydrophobic heat and moisture exchangers with heated humidifier during prolonged mechanical ventilation. Intensive Care Med. 1992;18(2):97–100.

    Article  CAS  PubMed  Google Scholar 

  70. Dreyfuss D, Djedaini K, Gros I, Mier L, LeBourdelles G, Cohen Y, et al. Mechanical ventilation with heated humidifiers or heat and moisture exchangers: effects on patient colonization and incidence of nosocomial pneumonia. Am J Respir Crit Care Med. 1995;151(4):986–92.

    CAS  PubMed  Google Scholar 

  71. Branson RD, Davis Jr K, Brown R, Rashkin M. Comparison of three humidification techniques during mechanical ventilation: patient selection, cost and infections considerations. Respir Care. 1996;41(9):809–16.

    Google Scholar 

  72. Villafane MC, Cinnella G, Lofaso F, Isabey D, Harf A, Lemaire F, et al. Gradual reduction of endotracheal tube diameter during mechanical ventilation via different humidification devices. Anesthesiology. 1996;85(6):1341–9.

    Article  CAS  PubMed  Google Scholar 

  73. Boots R, Howe S, George N, Harris F, Faoagali J. Clinical utility of hygroscopic heat and moisture exchangers in intensive care patients. Crit Care Med. 1997;25(10):1707–12.

    Article  CAS  PubMed  Google Scholar 

  74. Hurni J, Feihl F, Lazor R, Leuenberger P, Perret C. Safety of combined heat and moisture exchanger filters in long term mechanical ventilation. Chest. 1997;111(3):686–91.

    Article  CAS  PubMed  Google Scholar 

  75. Kirton O, De Haven B, Morgan J, Morejon O, Civetta J. A prospective, randomised comparison of an in-line heat moisture exchange filter and heated wire humidifiers: rates of ventilator-associated early-onset (community-acquired) or late-onset (hospital-acquired) pneumonia and incidence of endotracheal tube occlusion. Chest. 1997;112(4):1055–9.

    Article  CAS  PubMed  Google Scholar 

  76. Kollef MH, Shapiro SD, Boyd V, Silver P, Von Harz B, Trovillion E, et al. A randomized clinical trial comparing an extended-use hygroscopic condenser humidifier with heated-water humidification in mechanically ventilated patients. Chest. 1998;113(3):759–67.

    Article  CAS  PubMed  Google Scholar 

  77. Luchetti M, Stuani A, Castelli G, Marraro G. Comparison of three different humidification systems during prolonged mechanical ventilation. Minerva Anestesiol. 1998;64(3):75–81.

    CAS  PubMed  Google Scholar 

  78. Memish ZA, Oni GA, Djazmati W, Cunningham G, Mah MW. A randomized clinical trial to compare the effects of a heat and moisture exchanger with a heated humidifying system on the occurrence rate of ventilator-associated pneumonia. Am J Infect Control. 2001;29(5):301–5.

    Article  CAS  PubMed  Google Scholar 

  79. Diaz RB, Barbosa DA, Bettencourt AR, Vianna LAC, Gir E, Guimaraes T. Evalution [sic] the use of hygroscopic humidifier filters to prevent nosocomial pneumonia. Acta Paulista de Enfermagem. 2002;15(4):32–44.

    Google Scholar 

  80. Lacherade JC, Auburtin M, Cerf C, Van de Louw A, Soufir L, Rebufat Y, et al. Impact of humidification systems on ventilator-associated pneumonia: a randomized multicenter trial. Am J Respir Crit Care Med. 2005;172(10):1276–82.

    Article  PubMed  Google Scholar 

  81. Lorente L, Lecuona M, Jimenez A, Mora ML, Sierra A. Ventilator-associated pneumonia using a heated humidifier or a heat and moisture exchanger: a randomized controlled trial. Crit Care. 2006;10(4):R116.

    Article  PubMed  PubMed Central  Google Scholar 

  82. Boots RJ, George N, Faoagali JL, Druery J, Dean K, Heller RF. Double-heater-wire circuits and heat-and-moisture exchangers and the risk of ventilator-associated pneumonia. Crit Care Med. 2006;34(3):687–93.

    Article  PubMed  Google Scholar 

  83. Oguz S, Deger I. Ventilator-associated pneumonia in patients using HME filters and heated humidifiers. Ir J Med Sci. 2013;182(4):651–5.

    Article  CAS  PubMed  Google Scholar 

  84. Bench S. Humifification in thr long-term ventilated patients; a systematic review. Intens Crit Care Nurs. 2033;19:75–84.

    Article  Google Scholar 

  85. Hess DR, Kallstrom TJ, Mottram CD, et al. American Association for Respiratory Care are of the ventilator circuit and its relation to ventilator associated pneumonia. Respir Care. 2003;48:869–79.

    PubMed  Google Scholar 

  86. Kola A, Eckmanns T, Gaistmeier P. Efficacy of heat and moisturing exchangers in preventing ventilator-associated pneumonia: meta-analysis of randomized controlled trials. Intensive Care Med. 2005;31:5–11.

    Article  PubMed  Google Scholar 

  87. Siempos II, Vardakas KZ, Kopterides P, Falagas ME. Impact of passive humidifcation on clinical outcomes of mechanically ventilated patients: a meta-analysis of randomized controlled trials. Crit Care Med. 2007;35:2843–51.

    Article  PubMed  Google Scholar 

Download references

Acknowledgements

Not applicable.

Funding

The authors received no funding for this study.

Availability of data and materials

Not applicable.

Authors’ contributions

MV, DC, YS, LB, AME, PP and GS contributed substantially to the study design, data analysis and interpretation, and the writing of the manuscript. All authors read and approved the final manuscript.

Competing interests

The authors declare that they have no competing interests.

Consent for publication

Not applicable.

Ethics approval and consent to participate

Not applicable.

Publisher’s Note

Springer Nature remains neutral with regard to jurisdictional claims in published maps and institutional affiliations.

Author information

Authors and Affiliations

Authors

Corresponding author

Correspondence to Maria Vargas.

Additional files

Additional file 1:

Quality assessment. Quality assessment of these studies included: 1) use of randomization sequence generation, 2) reporting and type of allocation concealment, 3) blinding, 4) reporting of incomplete outcome data, 5) comparability of the groups at baseline. H: high, L: low, U: unknown. (PDF 80 kb)

Additional file 2:

Meta-regression. Meta-regressions for artificial airway occlusion, pneumonia and mortality. (PDF 411 kb)

Additional file 3:

Meta-analyses for hydrophobic and hygroscopic HME vs HH for Artificial airway occlusion, mortality and Pneumonia. (PDF 231 kb)

Rights and permissions

Open Access This article is distributed under the terms of the Creative Commons Attribution 4.0 International License (http://creativecommons.org/licenses/by/4.0/), which permits unrestricted use, distribution, and reproduction in any medium, provided you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons license, and indicate if changes were made. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated.

Reprints and permissions

About this article

Check for updates. Verify currency and authenticity via CrossMark

Cite this article

Vargas, M., Chiumello, D., Sutherasan, Y. et al. Heat and moisture exchangers (HMEs) and heated humidifiers (HHs) in adult critically ill patients: a systematic review, meta-analysis and meta-regression of randomized controlled trials. Crit Care 21, 123 (2017). https://doi.org/10.1186/s13054-017-1710-5

Download citation

  • Received:

  • Accepted:

  • Published:

  • DOI: https://doi.org/10.1186/s13054-017-1710-5

Keywords