Open Access

Implementing a bedside assessment of respiratory mechanics in patients with acute respiratory distress syndrome

  • Lu Chen1, 2,
  • Guang-Qiang Chen1, 2, 3,
  • Kevin Shore1,
  • Orest Shklar4,
  • Concetta Martins4,
  • Brian Devenyi4,
  • Paul Lindsay4,
  • Heather McPhail4,
  • Ashley Lanys2,
  • Ibrahim Soliman1,
  • Mazin Tuma1,
  • Michael Kim1,
  • Kerri Porretta4,
  • Pamela Greco4,
  • Hilary Every4,
  • Chris Hayes1, 2,
  • Andrew Baker1, 2,
  • Jan O. Friedrich1, 2 and
  • Laurent Brochard1, 2Email author
Critical Care201721:84

DOI: 10.1186/s13054-017-1671-8

Received: 19 August 2016

Accepted: 17 March 2017

Published: 4 April 2017

Abstract

Background

Despite their potential interest for clinical management, measurements of respiratory mechanics in patients with acute respiratory distress syndrome (ARDS) are seldom performed in routine practice. We introduced a systematic assessment of respiratory mechanics in our clinical practice. After the first year of clinical use, we retrospectively assessed whether these measurements had any influence on clinical management and physiological parameters associated with clinical outcomes by comparing their value before and after performing the test.

Methods

The respiratory mechanics assessment constituted a set of bedside measurements to determine passive lung and chest wall mechanics, response to positive end-expiratory pressure, and alveolar derecruitment. It was obtained early after ARDS diagnosis. The results were provided to the clinical team to be used at their own discretion. We compared ventilator settings and physiological variables before and after the test. The physiological endpoints were oxygenation index, dead space, and plateau and driving pressures.

Results

Sixty-one consecutive patients with ARDS were enrolled. Esophageal pressure was measured in 53 patients (86.9%). In 41 patients (67.2%), ventilator settings were changed after the measurements, often by reducing positive end-expiratory pressure or by switching pressure-targeted mode to volume-targeted mode. Following changes, the oxygenation index, airway plateau, and driving pressures were significantly improved, whereas the dead-space fraction remained unchanged. The oxygenation index continued to improve in the next 48 h.

Conclusions

Implementing a systematic respiratory mechanics test leads to frequent individual adaptations of ventilator settings and allows improvement in oxygenation indexes and reduction of the risk of overdistention at the same time.

Trial registration

The present study involves data from our ongoing registry for respiratory mechanics (ClinicalTrials.gov identifier: NCT02623192. Registered 30 July 2015).

Keywords

Pulmonary function test Respiratory physiology Esophageal pressure Mechanical ventilation Quality improvement

Background

Patients with acute respiratory distress syndrome (ARDS) present various degrees of impairment in respiratory mechanics and different physiological responses to a given level of positive end-expiratory pressure (PEEP). Applying the same ventilator regimen to every patient would be inadequate and at times can be potentially harmful. For example, the potential benefits of high PEEP in terms of oxygenation improvement and alveolar recruitment should be balanced against the risks induced by high pressures, such as hemodynamic impairment and overdistention. In other words, one needs to individualize the PEEP level by evaluating both its safety and its effectiveness for a specific patient [1]. This requires the assessment of gas exchange, respiratory mechanics, and hemodynamic variables. Additionally, partitioning lung and chest wall mechanics can also help the individualization of ventilator settings.

Despite their potential interest for clinical management, neither airway pressure (Paw)-based respiratory mechanics nor esophageal pressure (Pes)-based lung and chest wall mechanics are systematically assessed in routine practice. This discrepancy can be explained by technical issues [2] in obtaining accurate transpulmonary pressure (PL), by a lack of standardized procedures, and by the challenges of integrating the results of these measurements into ventilatory management. Even worse, the most recent large observational studies on patients with ARDS showed that simple parameters such as plateau pressure (Pplat) were not measured in the majority of the patients [3].

To improve the integration of respiratory mechanics measurement in our clinical practice, a group of physicians and respiratory therapists (RTs) at our institution introduced a respiratory mechanics test to systematically assess respiratory mechanics (i.e., performing a pulmonary function test) for patients with ARDS and be implemented as a quality improvement (QI) program. The goal was to provide clinicians with relevant physiological assessment that could be helpful for clinical practice. Because the needs for adjusting ventilator settings can be very different, this program did not include clinical recommendations or specific guidelines associated with these measurements.

Having implemented this systematic test in our clinical practice for 1 year, we retrospectively tried to assess if it had any impact. We looked for whether any changes in ventilatory settings were performed. We also assessed whether the observed changes modified physiological variables known to be associated with mortality, and we tried to understand whether the observed changes were consistent with the measurements.

Methods

Design and settings

This is a retrospective study of the impact of a 1-year program (see below) with an aim of systematically evaluating respiratory mechanics in patients with ARDS by comparing the ventilator settings and relevant physiological variables before and after performing the measurements. The program was decided by the critical care department at a teaching hospital (St. Michael’s Hospital, Toronto, ON, Canada) and implemented in both the medical-surgical and the trauma-neurosurgical intensive care units (ICUs). Of note, the measurements are entered into a registry for future studies.

Implementing the respiratory mechanics test in clinical practice

The procedure of the test was determined through discussion among ICU physicians, fellows, and RTs. A team of 22 users (3 physicians, 2 fellows, and 17 RTs) was recruited voluntarily to facilitate implementing the test in four aspects:
  1. a.

    To increase awareness and understanding of monitoring respiratory mechanics, education sessions consisting of lectures, bench and bedside hands-on sessions, and feedback rounds were provided to ICU clinicians. The education sessions were focused on explaining the importance of measuring respiratory mechanics, the technical approaches for measurements, the physiological and clinical meanings of the measured variables based on scientific evidence, and the limitations of those variables. We did not propose to use one single parameter on which to base changes of the ventilator settings; we proposed to incorporate multiple variables (e.g., airway pressure, Pplat, driving pressure [Pdriv], chest wall component, recruitability, oxygenation, and hemodynamic response to PEEP) into the global history of the patient and let the clinical team decide what was best for the patient.

     
  2. b.

    To standardize the procedures, we developed written protocols to guide esophageal catheter placement and the associated systematic measurements.

     
  3. c.

    To simplify the calculations, we developed a custom-programmed Portable Document Format form (PDF; Adobe Systems, San Jose, CA, USA) to automatically calculate physiological parameters and generate a clinical report (see Additional file 1: Appendixes S1 and S2).

     
  4. d.

    This clinical report was delivered to the caregivers in charge of the patients.

     

Patient enrollment process for the test

All patients admitted to the ICUs meeting the Berlin definition of ARDS [4] and receiving invasive mechanical ventilation were eligible. A daily screening was done, mostly on the weekdays. It was left at the discretion of the clinical team to decide to perform measurements, place esophageal catheters, and accept possible transient changes in sedation or paralysis (Additional file 1: Figure S1). Esophageal catheter insertion was recommended when the ratio of partial pressure of arterial oxygen and fraction of inspired oxygen (PaO2/FiO2) was ≤200 mmHg. In the group of patients with mild ARDS (i.e., PaO2/FiO2 > 200 mmHg), catheters were placed at the discretion of the clinical team. In the following cases, the clinical team discussed the benefits of doing the measurements on a case-by-case basis: (a) severe hemodynamic instability (i.e., >30% increase in the dose of vasopressors in the last 6 h or need for >0.5 μg/kg/minute of norepinephrine); or (b) a known esophageal problem, active upper gastrointestinal bleeding, or any other contraindication to the insertion of a gastric tube.

Measurements

Each patient enrolled underwent measurements of respiratory mechanics performed by one or two trained RTs and/or fellows (depending upon availability of clinicians) following a standardized protocol. More than 20 clinicians were considered as trained users. The patients were measured at the early stage of ARDS, and all of them were already deeply sedated and often paralyzed. Additional sedation with or without paralysis could be transiently necessary to suppress or minimize spontaneous breathing. This approach was accepted as part of our clinical practice to get reliable measurements of passive respiratory mechanics. Nevertheless, the decisions of deepening sedation and/or using paralysis for an individual patient were made at the discretion of the clinical team. The absence of spontaneous effort was confirmed by the absence of a negative Paw swing during a 3-second end-expiratory occlusion and by the presence of positive Pes swings during tidal breathing. Volume-controlled ventilation (VCV) was used during the measurements with a standardized tidal volume (VT) of 6 ml/kg of predicted body weight (PBW), a constant inspiratory flow of 50–60 L/minute, and 0.3-second pause at the end of inspiration. Respiratory rate (RR) was set to maintain a minute ventilation (VE) similar to premeasurement level. PEEP and the fraction of inspired oxygen (FiO2) were maintained at the clinically chosen level. Paw, airway flow, and airway volume were directly taken from the ventilator monitoring system. Pes was measured using a catheter with an air-filled balloon (CooperSurgical, Trumbull, CT, USA) with a pressure transducer connected to a bedside monitor (similarly to measuring central venous pressure). Measuring Pes was recommended as a component of the test for patients with a PaO2/FiO2 ratio ≤200 mmHg, but the decision of placing a catheter was left at the discretion of the clinicians. The validity of the Pes was confirmed using an occlusion test during spontaneous breathing (premeasurement) or a positive pressure occlusion test by manually compressing the thorax during passive breathing [2, 5]. Elastance, resistance, and other derived variables were automatically calculated using the programmed PDF form. The process of conducting the measurements was as follows:
  1. a.

    Paw-based respiratory mechanics were measured by using end-expiratory and end-inspiratory occlusions for 1–2 seconds. The absence of leakage during an end-inspiratory occlusion was confirmed by the equivalence of expiratory VT between the breaths with occlusion to the one without occlusion. Total positive end-expiratory pressure (PEEPtot), airway peak pressure (Ppeak), and airway Pplat were recorded. Intrinsic PEEP, Pdriv (Pplat − PEEPtot), respiratory system compliance, and resistance were then calculated automatically.

     
  2. b.

    Pes-based lung and chest wall mechanics [2] were measured simultaneously using end-expiratory and end-inspiratory occlusions. Transpulmonary pressure at end expiration (PL,end-exp) and transpulmonary pressure at end inspiration (PL,end-insp), lung compliance, chest wall compliance, and the ratio of lung elastance to respiratory system elastance were calculated automatically. PL, unless specifically indicated such as elastance-derived transpulmonary plateau pressure, was calculated using direct measurement of Pes.

     
  3. c.

    Oxygenation and hemodynamic responses to PEEP were assessed by increasing PEEP by 3–5 cmH2O (preferably 5 cmH2O) from the clinical PEEP level if the Pplat was <35 cmH2O (in the vast majority of the cases). PEEP was reduced by 3–5 cmH2O if the Pplat had reached 35 cmH2O or in the presence of poor hemodynamic tolerance. We report this procedure as an incremental PEEP trial. FiO2 was kept constant for comparing the change in PaO2/FiO2.

     
  4. d.

    Alveolar derecruitment was estimated using a single-breath simplified decremental PEEP maneuver (Fig. 1) performed within 10–15 seconds from the high PEEP used in the preceding step. A prolonged expiration (6–9 seconds) maneuver was performed while abruptly decreasing PEEP by 10 cmH2O from a high to a low level for one breath. Because inspiratory VT was unchanged, the difference in expiratory VT values between the expired VT displayed immediately after decreasing PEEP and the breath before changing PEEP was referred to as the total change in lung volume from high to low PEEP. In parallel, the predicted change in lung volume was estimated by the product of respiratory system compliance at low PEEP (or zero PEEP) and the change in pressure (i.e., 10 cmH2O of change in PEEP). When the total change in lung volume was greater than this predicted value, the difference was taken as an estimate of derecruited lung volume (Vder). A high Vder (e.g., ≥150 ml) due to reducing PEEP suggested that the PEEP was effective in recruiting the lung (or in maintaining the lung recruited). The rationale for this approach was reported previously, although the current method was simplified to make its use feasible rapidly at the bedside [6]. Of note, we reduced PEEP to estimate derecruitment instead of raising PEEP to estimate recruitment because we speculated that derecruitment may occur faster than recruitment and is easier to detect. These technical simplifications have not been fully validated, however, and this was made clear to the clinicians.

     
Fig. 1

Illustration of simplified decremental positive end-expiratory pressure (PEEP) maneuver for estimating derecruited lung volume. In this example, respiratory frequency was transiently reduced to 10 breaths per minute to allow a prolonged expiration. Afterward, PEEP was reduced from 15 to 5 cmH2O. The difference in expiratory tidal volumes (i.e., integral of flow) between the breath while decreasing PEEP (blue area) and the one before changing PEEP (red area) was referred to as the total change in lung volume. Derecruited volume was the difference between the total changes in measured vs. predicted lung volumes (see text for details). Paw Airway pressure

A clinical report (Additional file 1: Appendix S2) was then generated automatically with reference thresholds from the literature for diagnostic purposes. At the end of the measurements, initial ventilator settings were resumed, and the clinical report was given to the clinical team in charge of the patient. The clinicians then decided whether to change the ventilator settings if deemed necessary. There were no therapeutic recommendations attached to the report, and the clinical team was free to use and integrate the data into a more cohesive clinical decision-making process. The intention of this procedure was to provide intensivists with reliable information on the pulmonary function of patients with ARDS and possibly to allow clinicians to individualize ventilator settings.

Patient enrollment for the present analysis

Patients who had been enrolled in the program during its first year of implementation were eligible for the present study. Patients whose results of the respiratory mechanics were not provided to the clinical team after the measurements owing to technical problems (i.e., no clinical report was generated for reference) were excluded from the study.

Endpoints for the analysis

We designed a statistical plan to decide a priori which variables we wanted to test. The first question was to determine the proportion of patients in whom ventilatory settings were changed after the measurements. We also wanted to determine if the changes were consistent with the measurements. The relevant endpoints were the effects of these changes on physiological parameters known to be associated with mortality, namely the oxygenation index (OI = mean Paw × FiO2 × 100/PaO2) [7, 8], the estimated physiological dead-space fraction [9], Pplat, and Pdriv [10]. The OI integrates the intensity of ventilator assistance in terms of delivered mean Paw and its arterial oxygenation output. A low OI means that a relatively small intensity of assistance is needed for each millimeter of mercury of PaO2, whereas a high OI means that high delivered pressure and/or FiO2  is needed for each millimeter of mercury of PaO2. The dead-space fraction was calculated using the Enghoff modification of the Bohr equation with an estimation of resting energy expenditure using the Harris-Benedict equation and an assumption that the respiratory quotient equals 0.8. It is noteworthy that these two parameters, OI and dead-space fraction, were not directly given to clinicians. The Pdriv, calculated as the difference in Paw between a 1- to 2-second end-inspiratory occlusion (i.e., Pplat) and an end-expiratory occlusion (i.e., PEEPtot), was referred to as Pdriv in this study. For the purpose of our retrospective study, however, we found that the PEEPtot was not always assessed and documented by clinicians before or after the measurements and that the Pplat was sometimes estimated from the Ppeak in patients receiving pressure-controlled ventilation (PCV). The estimated Pdriv before and after the measurements could therefore slightly differ from the Pdriv during the measurements, and it was calculated as the difference between estimated Pplat and PEEP (not PEEPtot) and referred to as ∆P [10].

We also calculated corrected VE (i.e., VE for a PaCO2 of 40 mmHg) as a surrogate of dead space [4]. Inspired by the OI, we proposed a new index—the “Oxygenation/Stretch Index” (O/SI). Comparing it with the OI, we replaced the mean Paw with the Pdriv and switched the denominator with the nominator for easier calculation. Using Pdriv instead of mean Paw makes the index less dependent on the type of ventilator mode used (VCV vs. PCV). It was calculated using the following equation:
$$ \mathrm{O}/\mathrm{S}\mathrm{I} = \frac{\mathrm{Pa}{\mathrm{O}}_2/\mathrm{Fi}{\mathrm{O}}_2}{\Delta \mathrm{P}}\kern0.5em =\kern0.5em \frac{\mathrm{Pa}{\mathrm{O}}_2}{\mathrm{Fi}{\mathrm{O}}_2 \times \Delta \mathrm{P}} $$

The denominator can be referred to as the “cost” of mechanical ventilation, whereas the numerator is the “benefit.” Therefore, a low O/SI suggests a low “benefit-to-cost ratio” of mechanical ventilation; that is, the benefit was achieved by paying a relatively high cost, whereas a high O/SI suggests a high benefit-to-cost ratio. Both the corrected VE and the O/SI before and after the measurements were compared. The O/SI was not provided to clinicians but was calculated for the study.

Data collection and statistical analysis

We collected information by reviewing the clinical charts for patients’ demographic, physiological, and radiographic characteristics; ARDS risk factors; coexisting conditions; ventilator settings; arterial blood gas (ABG) analysis; and documented respiratory parameters (e.g., VE and Pplat) before and after measurements (i.e., premeasurement and postmeasurement). We used the ABGs that were closest to the time of measurements but at least 1 h away from the measurements, as well as the corresponding ventilator settings and respiratory parameters. (Also refer to the discussion.) Detailed physiological variables obtained during the measurements were also collected.

The details of statistical analysis are reported in the additional files. Statistical methods are also described in the notes of the tables. Notably, because we decided the variables to be tested a priori in the statistical plan, we decided against using a Bonferroni adjustment, which would have highly increased the risk of type II errors [11].

Results

During the first year of implementation (August 2014 to August 2015), 62 patients were enrolled and had measurements performed (Additional file 1: Figure S1). One patient was excluded from the study because of obvious input errors in the measurements (data were not used by clinicians).

The study cohort consisted of 61 patients. Their main characteristics are described in Table 1. Fifty-four patients (88.5%) were measured within 48 h of ARDS identification, and the remainder were measured between 48 and 120 h after ARDS identification. The main reason for a delay between identifying ARDS and performing measurements was hemodynamic instability. Paw-based respiratory mechanics were measured in all patients, and an esophageal catheter was placed in 54 patients (88.5%). In one patient, the catheter failed to pass through the upper esophageal sphincter, and in another, the catheter was electively placed with the assistance of gastroscopy. Pes was measured in 53 (86.9%) of 61 patients. A positive occlusion test was performed in all patients, with a ratio of changes in Pes and Paw during end-expiratory occlusion (∆Pes/∆Paw) of 1.1 ± 0.2 (mean ± SD). Seven patients (11.5%) were measured without obtaining ABGs for assessment of oxygenation response, and three patients (5.0%) were measured without performing a decremental PEEP maneuver for estimation of alveolar derecruitment. During the measurements, 21 patients (34.4%) received additional sedatives, and 3 patients (5.0%) required additional neuromuscular blockers to suppress or minimize spontaneous breathing. The 30-day mortality, ICU mortality, and hospital mortality rates in the study cohort were 37.7%, 37.7%, and 41.0%, respectively. ICU and hospital lengths of stay were 13.5 [8.8–29.0] and 19.0 [10.0–40.0] days, respectively.
Table 1

Characteristics of the patients (N = 61)

Characteristic

Value

Male sex, n (%)

48 (78.7)

Age, years

56 [45–68]

Predicted body weighta, kg

68 ± 11

Body mass index, kg/m2

29 ± 7

APACHE II score at admissionb

28 ± 10

SOFA score at inclusionc

12 ± 4

Days of NIV prior to intubation, n

0 [0–1]

Days of IMV at inclusiond, n

2 [1–6]

Days of ARDS at inclusiond, n

0 [0–1]

Risk factors of ARDSe, n (%)

 

 Pneumonia

32 (52.5)

 Extrapulmonary sepsis

9 (14.8)

 Trauma

9 (14.8)

 Noncardiogenic shock

7 (11.5)

 Pancreatitis

5 (8.2)

 Aspiration

4 (6.6)

 Pulmonary contusion

3 (4.9)

 Pulmonary vasculitis

2 (3.3)

 Drug overdose

2 (3.3)

 Blood transfusion

1 (1.6)

 No risk factor

4 (6.6)

Severity of ARDS, n (%)

 

 Mild

9 (14.8)

 Moderate

39 (63.9)

 Severe

13 (21.3)

Patients treated with ECMO

3 (4.9)

Patients with tracheostomy

9 (14.8)

Days of IMV after inclusion, n

9 [4–20]

Duration of IMV, days

12 [7–24]

Abbreviations: APACHE Acute Physiology and Chronic Health Evaluation, ARDS Acute respiratory distress syndrome, ECMO Extracorporeal membrane oxygenation, IMV Invasive mechanical ventilation, NIV Noninvasive ventilation, SOFA Sepsis-related Organ Failure Assessment

Dichotomous or nominal categorical variables are described in number (percent); continuous variables are described as mean ± SD or median [IQR], as appropriate

aPredicted body weight of male patients was calculated as 50 + 0.91 × (centimeters of height − 152.4) and that of female patients as 45.5 + 0.91 × (centimeters of height − 152.4)

bAPACHE II score at intensive care unit admission

cSOFA score at day of patient assessment (i.e., the day the patient was measured)

dThe days receiving IMV before inclusion and the days of ARDS identification before inclusion, respectively

ePatients can have more than one risk factor. The sum of the percent is hence >100%

Ventilator settings

Comparing postmeasurement with premeasurement, the ventilator settings were found to be altered in 41 patients (67.2%) and unchanged in the others. Changes included switching pressure-targeted mode to volume-targeted mode in 19 patients (31.1%), reducing peak inspiratory pressure in 4 patients who were receiving pressure-targeted mode, decreasing PEEP in 22 patients, and increasing PEEP in 9 patients (Fig. 2). VT was slightly lower in postmeasurement than in premeasurement (Table 2).
Fig. 2

Panel a shows the clinical adjustments in positive end-expiratory pressure (PEEP) and effects on the oxygenation index (OI) (n = 59). b Patients were classified in three groups according to the change in their PEEP level. Bonferroni adjustment was not used. (Refer to main text for explanations.)

Table 2

Ventilator settings and physiological variables before and after measurements (N = 61)

 

Premeasurement

Postmeasurement

P value

Ventilator settings

 VT/PBW, ml/kg

6.5 [6.2–7.0]

6.4 [6.2–6.7]

0.006

 PEEP, cmH2O

12 [10–14]

12 [10–14]

0.077

 FiO2

0.60 [0.50–0.70]

0.60 [0.50–0.70]

0.325

Physiological variables

 PaCO2, mmHg

41 [38–50]

42 [38–50]

0.553

 PaO2/FiO2, mmHg

146 ± 60

162 ± 69

0.020

 Pplat a, cmH2O

30 ± 5

28 ± 5

0.004

 ∆Pa, cmH2O

18 [14–20]

15 [12–19]

0.023

 OIb, cmH2O/mmHg

15.2 ± 7.4

13.8 ± 8.3

0.021

 O/SIa, mmHg/cmH2O

7.5 [5.4–11.5]

8.2 [5.9–14.7]

0.029

 VD/VT, est

0.63 ± 0.10

0.62 ± 0.12

0.494

 VE, corr, L/minute

13.0 ± 3.2

12.8 ± 3.3

0.421

Abbreviations: Vt/PBW Tidal volume per predicted body weight, PEEP Positive end-expiratory pressure, FiO 2 Fraction of inspired oxygen, PaCO 2 Partial pressure of carbon dioxide, PaO 2 /FiO 2 Ratio of partial pressure of arterial oxygen to fraction of inspired oxygen, Pplat Plateau pressure, ∆P Driving pressure, OI Oxygenation index, O/SI Oxygenation/Stretch Index, indicating a benefit-to-cost ratio, V D /V T , esti Estimated physiological dead-space fraction, V E,corr Corrected expired volume per minute

Continuous variables were described as mean ± SD and compared using paired t tests or described as median [IQR] and compared using Wilcoxon matched-pairs signed-rank sum tests, as appropriate. Bonferroni adjustment was not used. (Refer to main text for explanations.)

a N = 45; 6 patients who were ventilated in volume control mode without documented Pplat were excluded. ∆P was calculated as the difference between Pplat and PEEP

b N = 59; 2 patients who were ventilated in pressure support mode either before or after the measurements were excluded from analysis

Physiological outcomes

Comparing postmeasurement with premeasurement values (Table 2), both OI and distending pressures (i.e., Pplat and ∆P) were significantly reduced. Measured at the same level of FiO2, the PaO2/FiO2 and the O/SI were significantly increased. In patients in whom PEEP was increased, the PaO2/FiO2 increased from 124 ± 78 mmHg to 173 ± 89 mmHg (P = 0.010), whereas the mean Paw increased from 18 ± 3 cmH2O to 21 ± 3 cmH2O (P < 0.0005). In patients whose PEEP was decreased, the PaO2/FiO2 remained stable (150 ± 64 mmHg vs. 153 ± 70 mmHg; P = 0.772), whereas the mean Paw decreased from 20 ± 4 cmH2O to 16 ± 4 cmH2O (P < 0.0005) (Fig. 2). On average, for the whole group, PaO2/FiO2 improved, whereas mean Paw decreased.

The estimated physiological dead-space fraction and corrected VE remained unchanged (Table 2). Arterial pH, bicarbonate level, heart rate, and mean arterial pressure remained unchanged (Additional file 1: Table S3).

Changes in ventilator settings associated with measurements

To determine whether the observed clinical adjustments in PEEP were consistent with the results of the respiratory mechanics tests, we separated patients into three groups according to whether the PEEP was decreased by the clinicians, unchanged, or increased postmeasurement compared with premeasurement. Comparisons of the physiological variables between groups are shown in Table 3. The three groups significantly differed on four parameters, with gradually increasing values suggesting that these values could be the reason for making the changes. The airway Pplat and the elastance-derived transpulmonary Pplat were higher in the PEEP-decreased group than in the group with unchanged PEEP. PL,end-exp was not different between groups. The changes in Pdriv and lung Pdriv during the incremental PEEP trial demonstrated a consistent trend among the three groups, and the lung Pdriv displayed a statistically significant difference. Vder was lower in the group with decreased PEEP than in the group with increased PEEP. A Spearman’s correlation analysis also confirmed that Pplat, elastance-derived transpulmonary Pplat, and the changes in lung Pdriv were correlated with the differences in PEEP before and after measurements, again suggesting that the amount and the direction of the PEEP change could be explained by values of respiratory mechanics (Additional file 1: Table S1).
Table 3

Relationship between measurements and subsequent clinical adjustments of PEEP

 

PEEP at postmeasurement vs. at premeasurement

 

Measured variables

Decreased (n = 20)

Unchanged (n = 27)

Increased (n = 8)

P value

Markers of overdistentiona

 Pplat, cmH2O

28 ± 5b

25 ± 4b

26 ± 2

0.013

 Elastance-derived PL,plat, cmH2O

21 [20–26]b

17 [16–20]b

18 [16–21]

0.034

 PL,end-insp, cmH2O

5 [3–18]

8 [4–9]

4 [0–10]

0.327

 Pdriv, cmH2O

15 ± 5

12 ± 4

13 ± 3

0.098

 PL,driv, cmH2O

12 ± 5

9 ± 4

9 ± 5

0.063

Risk of atelectasisa

 PL,end-exp, cmH2O

−2 ± 5

−2 ± 5

−5 ± 5

0.335

Response to the incremental PEEP trialc

 Changes in Pdriv, cmH2O

2.0 [0–3.5]

1.0 [0–1.0]

0.5 [0–12]

0.169

 Changes in PL,driv, cmH2O

1.9 ± 2.5

0.7 ± 1.8

−0.3 ± 1.2

0.042

 Changes in PaO2/FiO2, mmHg

−4 [−17 to 12]

0 [−18 to 14]

3 [−29 to 10]

0.226

 Changes in MAP, mmHg

−2 [−10 to 3]

−2 [2-9]

−3 [−8 to 0]

0.675

Recruitabilityd

 Vder, ml

105 ± 61e

142 ± 106

208 ± 124e

0.036

Abbreviations: P plat Airway plateau pressure, Elastance-derived P L,plat Elastance-derived transpulmonary plateau pressure, calculated using airway plateau pressure times the ratio of lung elastance to respiratory system elastance, P L,end-insp Transpulmonary pressure measured at end inspiration occlusion, P driv Driving pressure, P L,driv Lung driving pressure (i.e., difference between PL,end-insp and PL,end-exp, P driv driving pressure, measured by the difference between plateau pressure and total PEEP), P L,end-exp Transpulmonary pressure measured at the end-expiratory occlusion, MAP Mean arterial pressure, V der derecruited volume

Continuous variables were described as mean ± SD and compared using one-way analysis of variance with the Bonferroni post hoc test, or described as median [IQR] and compared using the Kruskal-Wallis test with Dunn’s post hoc test, as appropriate. Bonferroni adjustment was not used. (Refer to main text for explanations.) There is some variation in the number of measurements because of missing data. We report the number of measurements in detail for each variable in the additional files

aVariables reflecting the risks of overdistention and atelectasis were measured at clinical PEEP level

b P < 0.05 PEEP-decreased vs. PEEP-unchanged

cResponses to an increment of PEEP in 3–5 cmH2O, expressed per 1-cmH2O PEEP increase

dAssessed by estimating the alveolar derecruitment with decreasing PEEP by 10 cmH2O

e P < 0.05 PEEP-decreased vs. PEEP-increased

Discussion

To our knowledge, this may be the first demonstration that systematic respiratory mechanics testing can be implemented with adequate validity and timelines in ICUs as a monitoring tool used for ventilatory adjustments. This provided physiological parameters for clinicians and helped to define the ventilatory therapy in patients with ARDS, as indicated by the changes in ventilator settings and in physiological variables observed after the test. Indeed, oxygenation could be improved using lower airway and distending pressures, a result that was unexpected, as indicated by the changes in OI, O/SI, Pplat, and ∆P. Such modifications would not have been performed without a systematic assessment.

Validity of the measurements

Measurements of Pes in routine clinical practice out of the field of research have often been considered to be a challenge [2]. By implementing education sessions, standardized procedures, and an electronic form for automatic calculations, the measurements of Pes were conducted in our ICUs by a number of healthcare professionals (RTs, medical doctors) who were not experienced researchers. The ∆Pes/∆Paw ratio during the positive pressure occlusion test—a method used to validate appropriate positioning of the esophageal catheter and the Pes measurements—was consistent across our cohort and very close to unity, indicating that the obtained Pes was reliable and provided a valid measure of changes in pleural pressure. These ∆Pes/∆Paw ratios were also confirmed using occlusion tests against inspiratory effort [5] in six patients after resumption of spontaneous efforts.

Changes in mode

The ventilator mode was often switched from pressure-targeted mode to volume-targeted mode, which may have been a direct effect of the test. An overall statistically significant but not clinically significant reduction in VT was also observed after the measurements. Though there is no evidence showing that one mode is superior [13], VCV mode offers several advantageous features compared with PCV mode:
  1. a.

    Strictly controlling the VT at a target value was proven beneficial for survival [14].

     
  2. b.

    Volume control permits an easier approach to monitoring respiratory mechanics. By setting an inspiratory pause time of 0.3–0.5 seconds, one can monitor Pplat in real time for each breath as well as its trend, leading to easy estimation of essential parameters such as Pdriv, compliance, and resistance. Peak Paw set in PCV is often used as a surrogate of Pplat, which potentially provides a simple method to limit Pdriv. This approach can overestimate Pplat when inspiratory flow does not return to zero at the end of inspiration; however, it can underestimate Pplat in the presence of the patient’s inspiratory effort.

     
  3. c.

    By using an inspiratory pause time with a high and constant inspiratory flow, VCV can improve the elimination of carbon dioxide [15].

     

Changes in PEEP

A major change observed concerned the individualization of PEEP, as depicted in Fig. 2. Terragni et al. described how respiratory system mechanics may help in preventing lung-injury in patients with ARDS [12]. As shown in Table 3, the clinicians used the variables reflecting the risk of overdistention, the response to PEEP (see also the individual oxygenation response to PEEP in Fig. 3), and the recruitability to adjust PEEP, rather than the one reflecting the risk of atelectasis (PL,end-exp).
Fig. 3

Individual oxygenation responses to the incremental positive end-expiratory pressure trial. PaO 2 /FiO 2 ratio Ratio of partial pressure of arterial oxygen to fraction of inspired oxygen; PEEPtot Total positive end-expiratory pressure

Physiological effects

The changes in ventilator settings, especially a frequent reduction in PEEP level, eventually led to lower airway Pplat and Pdriv, and at the same time to improvements in oxygenation, OI, and O/SI (Table 2). This result was rather unexpected. We were surprised by the highly variable response to PEEP changes (Fig. 3). These results are different form the usual expectation of the effects of PEEP, but, in reality, they looked quite similar to recently published results of randomized clinical trials [16]. Oxygenation was improved but was not a primary physiological endpoint in our study for several reasons. With high Paw, improvements in oxygenation can be achieved by a reduction in cardiac output and shunt, which may worsen oxygenation delivery [17]. With high pressure, improvements in oxygenation have been associated with similar [18] or even worse mortality [14, 19]. OI takes into account both Paw and oxygenation, and it has been shown to be associated with survival as well as ∆P [8, 10, 20, 21]. We also found that the O/SI, as a new index to evaluate the benefit-to-cost ratio of mechanical ventilation, was significantly improved after the measurements. Although there are no data to support the possible association between O/SI and survival, we separated patients in our cohort by hospital outcome and found that O/SI (calculated using data from the measurements) in survivors was significantly higher than in nonsurvivors (14.6 [10.9-18.8] and 8.8 [6.0-12.6], respectively; P < 0.0005). We reasoned that the O/SI, similarly to OI, may be more meaningful than oxygenation itself regarding clinical outcome.

Time points for comparisons

We decided to focus our comparison of the physiological variables at 1 h after the measurements to better ensure that the adjustments of ventilator settings and the physiological impact were likely secondary to the measurements. Later, the ventilation mode can be switched to partial assist mode. Also, comparing mean Paw (required for calculating OI) or Pdriv in pressure-target mode becomes challenging once the patient recovers spontaneous breathing effort.

We, however, also reviewed data at 24 h and 48 h. The improvements in physiological variables were consistent, indicated by the progressively lower OI at 24 and 48 h (12.4 ± 7.1 and 10.2 ± 0.9, respectively; P = 0.016) than at 1 h after the measurements. Of note, 16 and 21 patients (26.2% and 34.4%, respectively) had spontaneous effort, defined as actual RR exceeding preset RR or receiving partial support mode, – at 24 and 48 h.

Limitations

Our study has limitations. First, though we tried to ensure that measurements were likely based on reasonable physiological principles, what we established was an association between the measurements and the changes in settings, owing to the nature of an observational study. Second, with no control group, we are not able to describe any results on outcomes. Third, the results are limited to a single center. Fourth, there were some missing data in the assessment, such as the response to PEEP and the recruitability. Fifth, the simplified method for estimating derecruitment at the bedside requires further validation. Sixth, although we tried to minimize a direct influence of the initiators of the project (LC, LB) on the ventilator settings, interactions were frequent at the beginning of the project, and it is difficult to ascertain their exact influence. These interactions were meant to overcome the technical and knowledge barriers to measurement and interpretation of respiratory mechanics in clinical practice.

Conclusions

A respiratory mechanics test can be embedded in clinical practice and provides physiological parameters for clinicians. It leads to individualization of ventilator settings in patients with ARDS that improved physiological endpoints known to be associated with clinical outcomes, and it allowed reduction in the pressures needed to maintain oxygenation.

Abbreviations

ABG: 

Arterial blood gas

APACHE: 

Acute Physiology and Chronic Health Evaluation

ARDS: 

Acute respiratory distress syndrome

ECMO: 

Extracorporeal membrane oxygenation

FiO2

Fraction of inspired oxygen

ICU: 

Intensive care unit

IMV: 

Invasive mechanical ventilation

MAP: 

Mean arterial pressure

NIV: 

Noninvasive ventilation

OI: 

Oxygenation index

O/SI: 

Oxygenation/stretch index

PaCO2

Partial pressure of arterial carbon dioxide

PaO2/FiO2 ratio: 

Ratio of partial pressure of arterial oxygen to fraction of inspired oxygen

Paw: 

Airway pressure

PBW: 

Predicted body weight

PCV: 

Pressure-controlled ventilation

Pdriv: 

Driving pressure (Pplat − PEEPtot) or ∆P (Pplat − PEEP)

PEEP: 

Positive end-expiratory pressure

PEEPtot: 

Total positive end-expiratory pressure

Pes: 

Esophageal pressure

∆Pes/∆Paw: 

Ratio of changes in Pes and Paw during end-expiratory occlusion

PL

Transpulmonary pressure

PL

end-exp, Transpulmonary pressure at end expiration

PL

end-insp, Transpulmonary pressure at end inspiration

Ppeak: 

Airway peak pressure

Pplat: 

Airway plateau pressure

QI: 

Quality improvement

RR: 

Respiratory rate

RT: 

Respiratory therapist

SOFA: 

Sepsis-related Organ Failure Assessment

VCV: 

Volume-controlled ventilation

VD/VT,esti

Estimated physiological dead-space fraction

Vder

Derecruited lung volume

VE,corr

Corrected expired volume per minute

VE

Minute ventilation

VT

Tidal volume

Declarations

Acknowledgements

We greatly appreciate the continuous support from the nurses and physicians in the Department of Critical Care, as well as the Department of Respiratory Therapy, for their support and involvement with this project, with special thanks to leadership members (Karen Wannamaker, Margaret Oddi). We are particularly grateful to the respiratory therapists for their noteworthy contributions, namely Erica Berridge, Kay Ching, Karima Dewji, Joana Ferreira, Stacey Halliday, Noreen Kassam, Lesia Kolwzan, Julia Lee, Jennifer Lim, Priya Patel, Rickey Russell, Carolins Sierra, Nicole Wailoo, and Tracy Walsh. We thank the physicians for their collaboration, including Dr. Karen Burns, Dr. Claudia Dos Santos, Dr. David Hall, Dr. David Klein, Dr. Natalie Wong, Dr. Sara Gray, Dr. John Laffey, Dr. Warren Lee, Dr. Simon Abrahamson, Dr. John Marshall, and Dr. Antoine Pronovost. We also thank Carolyn Campbell and Orla Smith for their continuous help with the research organization. We are greatly thankful to Jianing Gu for her hard work on data collection. We are full of gratitude to Dr. Jordi Mancebo for his helpful comments on the manuscript.

Funding

LB holds the Keenan Chair in Critical Care Medicine and Acute Respiratory Failure at St. Michael’s Hospital. No other funding was available.

Availability of data and materials

The datasets used and/or analyzed during the present study are available from the corresponding author on reasonable request.

Authors’ contributions

LC, GQC, OS, CM, BD, PL, HM, IS, MT, MK, KP, PG, and HE made substantial contributions to data acquisition and helped to revise the manuscript. LC, GQC, KS, AL, and LB participated in the design of the study, performed the statistical analysis, and helped to revise the manuscript. LC, GQC, OS, CM, PG, HE, CH, AB, JOF, and LB conceived of the study and participated in its design and coordination. LC and LB drafted the manuscript. GQC, OS, CM, PG, HE, CH, AB, and JOF helped to revise the draft of the manuscript. All authors read and approved the final manuscript.

Authors’ information

Not applicable.

Competing interests

The authors declare that they have no competing interests.

Consent for publication

Not applicable.

Ethics approval and consent to participate

This study was approved by the research ethics board (REB#16-095) of St. Michael’s Hospital (Toronto, ON, Canada). Informed consent was waived on the basis of the study’s retrospective observational design preserving confidentiality of personal information. Please refer to the REB approval letter (see Additional file 2 ).

Publisher’s Note

Springer Nature remains neutral with regard to jurisdictional claims in published maps and institutional affiliations.

Open AccessThis article is distributed under the terms of the Creative Commons Attribution 4.0 International License (http://creativecommons.org/licenses/by/4.0/), which permits unrestricted use, distribution, and reproduction in any medium, provided you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons license, and indicate if changes were made. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated.

Authors’ Affiliations

(1)
Interdepartmental Division of Critical Care Medicine, University of Toronto
(2)
Keenan Research Centre and Li Ka Shing Institute, Department of Critical Care, St. Michael’s Hospital
(3)
Department of Critical Care Medicine, Beijing Tiantan Hospital, Capital Medical University
(4)
Department of Respiratory Therapy, St. Michael’s Hospital

References

  1. Chen L, Brochard L. Lung volume assessment in acute respiratory distress syndrome. Curr Opin Crit Care. 2015;21(3):259–64.View ArticlePubMedGoogle Scholar
  2. Akoumianaki E, Maggiore SM, Valenza F, Bellani G, Jubran A, Loring SH, et al. The application of esophageal pressure measurement in patients with respiratory failure. Am J Respir Crit Care Med. 2014;189(5):520–31.View ArticlePubMedGoogle Scholar
  3. Bellani G, Laffey JG, Pham T, Fan E, Brochard L, Esteban A, et al. Epidemiology, patterns of care, and mortality for patients with acute respiratory distress syndrome in intensive care units in 50 countries. JAMA. 2016;315(8):788–800.View ArticlePubMedGoogle Scholar
  4. The ARDS Definition Task Force. Acute respiratory distress syndrome: the Berlin Definition. JAMA. 2012;307(23):2526–33.Google Scholar
  5. Baydur A, Behrakis PK, Zin WA, Jaeger M, Milic-Emili J. A simple method for assessing the validity of the esophageal balloon technique. Am Rev Respir Dis. 1982;126(5):788–91.PubMedGoogle Scholar
  6. Dellamonica J, Lerolle N, Sargentini C, Beduneau G, Di Marco F, Mercat A, et al. PEEP-induced changes in lung volume in acute respiratory distress syndrome: two methods to estimate alveolar recruitment. Intensive Care Med. 2011;37(10):1595–604.View ArticlePubMedGoogle Scholar
  7. Marini JJ, Ravenscraft SA. Mean airway pressure: physiologic determinants and clinical importance—Part 1: physiologic determinants and measurements. Crit Care Med. 1992;20(10):1461–72.View ArticlePubMedGoogle Scholar
  8. Dechert RE, Park PK, Bartlett RH. Evaluation of the oxygenation index in adult respiratory failure. J Trauma Acute Care Surg. 2014;76(2):469–73.View ArticlePubMedGoogle Scholar
  9. Beitler JR, Thompson BT, Matthay MA, Talmor D, Liu KD, Zhuo H, et al. Estimating dead-space fraction for secondary analyses of acute respiratory distress syndrome clinical trials. Crit Care Med. 2015;43(5):1026–35.View ArticlePubMedPubMed CentralGoogle Scholar
  10. Amato MB, Meade MO, Slutsky AS, Brochard L, Costa EL, Schoenfeld DA, et al. Driving pressure and survival in the acute respiratory distress syndrome. N Engl J Med. 2015;372(8):747–55.View ArticlePubMedGoogle Scholar
  11. Perneger TV. What’s wrong with Bonferroni adjustments. BMJ. 1998;316(7139):1236–8.View ArticlePubMedPubMed CentralGoogle Scholar
  12. Terragni PP, Rosboch GL, Lisi A, Viale AG, Ranieri VM. How respiratory system mechanics may help in minimising ventilator-induced lung injury in ARDS patients. Eur Respir J. 2003;22(42 Suppl):15 s–21s.Google Scholar
  13. Rittayamai N, Katsios CM, Beloncle F, Friedrich JO, Mancebo J, Brochard L. Pressure-controlled vs volume-controlled ventilation in acute respiratory failure: a physiology-based narrative and systematic review. Chest. 2015;148(2):340–55.View ArticlePubMedGoogle Scholar
  14. The Acute Respiratory Distress Syndrome Network. Ventilation with lower tidal volumes as compared with traditional tidal volumes for acute lung injury and the acute respiratory distress syndrome. N Engl J Med. 2000;342(18):1301–8.View ArticleGoogle Scholar
  15. Devaquet J, Jonson B, Niklason L, Si Larbi AG, Uttman L, Aboab J, et al. Effects of inspiratory pause on CO2 elimination and arterial PCO2 in acute lung injury. J Appl Physiol (1985). 2008;105(6):1944–9.Google Scholar
  16. Goligher EC, Kavanagh BP, Rubenfeld GD, Adhikari NK, Pinto R, Fan E, et al. Oxygenation response to positive end-expiratory pressure predicts mortality in acute respiratory distress syndrome: a secondary analysis of the LOVS and ExPress trials. Am J Respir Crit Care Med. 2014;190(1):70–6.View ArticlePubMedGoogle Scholar
  17. Suter PM, Fairley B, Isenberg MD. Optimum end-expiratory airway pressure in patients with acute pulmonary failure. N Engl J Med. 1975;292(6):284–9.View ArticlePubMedGoogle Scholar
  18. Mercat A, Richard JCM, Vielle B, Jaber S, Osman D, Diehl JL, et al. Positive end-expiratory pressure setting in adults with acute lung injury and acute respiratory distress syndrome: a randomized controlled trial. JAMA. 2008;299(6):646–55.View ArticlePubMedGoogle Scholar
  19. Meade MO, Cook DJ, Guyatt GH, Slutsky AS, Arabi YM, Cooper DJ, et al. Ventilation strategy using low tidal volumes, recruitment maneuvers, and high positive end-expiratory pressure for acute lung injury and acute respiratory distress syndrome: a randomized controlled trial. JAMA. 2008;299(6):637–45.View ArticlePubMedGoogle Scholar
  20. Monchi M, Bellenfant F, Cariou A, Joly LM, Thebert D, Laurent I, et al. Early predictive factors of survival in the acute respiratory distress syndrome: a multivariate analysis. Am J Respir Crit Care Med. 1998;158(4):1076–81.View ArticlePubMedGoogle Scholar
  21. Seeley E, McAuley DF, Eisner M, Miletin M, Matthay MA, Kallet RH. Predictors of mortality in acute lung injury during the era of lung protective ventilation. Thorax. 2008;63(11):994–8.View ArticlePubMedPubMed CentralGoogle Scholar

Copyright

© The Author(s). 2017

Advertisement