Skip to main content

Can calculation of energy expenditure based on CO2 measurements replace indirect calorimetry?

An Erratum to this article was published on 12 April 2017

Abstract

Background

Methods to calculate energy expenditure (EE) based on CO2 measurements (EEVCO2) have been proposed as a surrogate to indirect calorimetry. This study aimed at evaluating whether EEVCO2 could be considered as an alternative to EE measured by indirect calorimetry.

Methods

Indirect calorimetry measurements conducted for clinical purposes on 278 mechanically ventilated ICU patients were retrospectively analyzed. EEVCO2 was calculated by a converted Weir’s equation using CO2 consumption (VCO2) measured by indirect calorimetry and assumed respiratory quotients (RQ): 0.85 (EEVCO2_0.85) and food quotient (FQ; EEVCO2_FQ). Mean calculated EEVCO2 and measured EE were compared by paired t test. Accuracy of EEVCO2 was evaluated according to the clinically relevant standard of 5% accuracy rate to the measured EE, and the more general standard of 10% accuracy rate. The effects of the timing of measurement (before or after the 7th ICU day) and energy provision rates (<90 or ≥90% of EE) on 5% accuracy rates were also analyzed (chi-square tests).

Results

Mean biases for EEVCO2_0.85 and EEVCO2_FQ were -21 and -48 kcal/d (p = 0.04 and 0.00, respectively), and 10% accuracy rates were 77.7 and 77.3%, respectively. However, 5% accuracy rates were 46.0 and 46.4%, respectively. Accuracy rates were not affected by the timing of the measurement, or the energy provision rates at the time of measurements.

Conclusions

Calculated EE based on CO2 measurement was not sufficiently accurate to consider the results as an alternative to measured EE by indirect calorimetry. Therefore, EE measured by indirect calorimetry remains as the gold standard to guide nutrition therapy.

Background

Indirect calorimetry is the gold standard method to determine energy expenditure (EE) and guide nutrition therapy in critically ill patients in order to avoid deleterious under- or overfeeding [13]. Indirect calorimeters analyze respiratory gases of the patients to measure their oxygen consumption (VO2) and carbon dioxide production (VCO2) and derive EE by the Weir’s equation [4]. The ratio of VCO2 to VO2 (VCO2/VO2), called the respiratory quotient (RQ), is considered as an indicator of measurement adequacy (i.e. <0.67 and >1.3 require careful interpretation) [5] and of substrates oxidation in stable state subjects [6]. However, indirect calorimetry is not conducted routinely in most intensive care units (ICU) mainly due to the lack of a reliable indirect calorimeter, and manpower with appropriate expertise to conduct and analyze the results [7]. Recent studies comparing currently available calorimeters have demonstrated that the measured EE is variable from one calorimeter to another [8, 9], although they are still more consistent than the EE calculated from predictive equations [10].

Methods to calculate EE from CO2 measurements (EEVCO2) derived from mechanical ventilators have been proposed as a surrogate to indirect calorimetry [11, 12]. This simplified method measures only the VCO2 derived from measurements of exhaled gas volume and CO2 concentrations. The approach assumes that the RQ value is either a fixed value (e.g. 0.85) [12] or equal to the food quotient (FQ) [11], i.e. the estimated RQ resulting from the oxidation of energy substrates. Recent studies have demonstrated that EEVCO2 estimates EE of critically ill patients more accurately than predictive equations [11, 12]. However, the validity of this method as an alternative to indirect calorimetry is questionable, since the variability of RQ demonstrated in previous literature is likely to influence the accuracy of the EEVCO2 calculation [5].

This study aims at determining whether EEVCO2 can be considered an alternative to EE measured by indirect calorimetry in critically ill patients on mechanical ventilation.

Methods

Study design

This retrospective observational study includes the measurements of indirect calorimetry performed in the mixed medical-surgical adult ICU at the Geneva University Hospital, Switzerland.

Data collection

We included 278 mechanically ventilated ICU patients from the institutional database of indirect calorimetry, conducted by the Clinical Nutrition Department. We collected data regarding physical characteristics (age, gender, height, anamnestic weight, body mass index), diagnosis and severity at ICU admission (primary diagnosis, Acute Physiology and Chronic Health Evaluation (APACHE) II score [13], Simplified Acute Physiology Score (SAPS) II [14]), and patient observations at the time of the measurement (body temperature, Glasgow coma scale, Sedation-Agitation Scale [15]). We also reported the characteristics of energy provision at the time of the indirect calorimetry, consisting of the route of administration (enteral nutrition (EN), parenteral nutrition (PN), combined (EN + PN), or non-nutritional energy sources), the total administered energy (kcal/d) along with the composition of the energy substrates (carbohydrate, protein, lipids). Energy from glucose-containing solutions and propofol was included in the calculation of the total administered energy.

Indirect calorimetry

All measurements were conducted using the Deltatrac Metabolic Monitor® (Datex, Helsinki, Finland) which was calibrated before each measurement according to the manufacturer’s procedure. Calorimetry was not conducted in patients with ventilator settings exceeding the general limits (fraction of inhaled O2 (FiO2) < 60%, positive end expiratory pressure (PEEP) < 9cmH2O), or with contraindicated treatments (e.g. chest drains for pneumothorax, inhaled nitric oxide). Patients were in resting state for at least 1 hour prior to the measurement, i.e. free of physical activity, mobilization, and transport. Measurements were conducted to obtain valid recordings for at least 20 minutes in a single measurement session for each patient after reaching steady state; recordings from the first 5 minutes of measurements and from non-stable states were excluded. As the Deltatrac® was a stand-alone device, minute-by-minute readings of VO2 (ml/min STPD) and VCO2 (ml/min STPD) were obtained as printouts or handwritten recordings. The results were reported in the clinical database as the means of valid recordings for the duration of the measurement. Mean VO2 and mean VCO2 of each patient was used for the individual calculation the RQ (=VCO2/VO2) and the determination of the measured EE by the modified Weir’s equation:

$$ \mathrm{E}\mathrm{E}\ \left(\mathrm{kcal}/\mathrm{d}\right) = 1.44 \times \left[3.941 \times {\mathrm{VO}}_2\left(\mathrm{mL}/ \min \right) + 1.11 \times {\mathrm{VCO}}_2\left(\mathrm{mL}/ \min \right)\right] $$

Fraction of inspired O2 (FiO2, %), and minute volume of ventilation (L/min BTPS) at the time of the indirect calorimetry were also collected. The day of the measurement was recorded as the number of days after ICU admission.

EEVCO2 calculation

The EEVCO2 was calculated using VCO2 values measured by indirect calorimetry, and two different RQ values: fixed value of 0.85 (EEVCO2_0.85) and FQ (EEVCO2_FQ).

Calculation of EEVCO2_0.85

By assuming an RQ = 0.85 [12], the Weir’s equation can be rewritten as:

$$ {\mathrm{EEVCO}}_2\_0.85\left(\mathrm{kcal}/\mathrm{d}\right)=1.44\times \left[3.941\times {\mathrm{VCO}}_2\left(\mathrm{mL}/ \min \right)/0.85+1.11\times {\mathrm{VCO}}_2\left(\mathrm{mL}/ \min \right)\right] $$

Calculation of EEVCO2_FQ

The FQ calculation was based on the O2 utilization and CO2 production during energy substrate oxidation [5]. Carbohydrates are oxidized for energy as:

C6H12O6+ 6O2 → 6CO2+ 6H2O + 679 kcal.

The ratio of produced CO2 to consumed O2, or FQ, is therefore equal to 1.0 (6 CO2/6 O2 = 1).

Fat oxidation is described by the following the reaction (e.g. palmitic acid):

CH3(CH2)14COOH + 23O2 → 16CO2+ 16H2O + 2,395 kcal

The results depend on the type of fatty acid, but on average the FQ will be 0.7 [5]. Protein oxidation is not as simple to express in a formula, as they are not always completely oxidized in the body. The FQ for protein is 0.8 [5]. By assuming that RQ is equal to FQ, the Weir’s equation can be rewritten as following:

$$ {\mathrm{EEVCO}}_2\_\mathrm{F}\mathrm{Q}\left(\mathrm{kcal}/\mathrm{d}\right)=1.44\times \left[3.941\times {\mathrm{VCO}}_2\left(\mathrm{mL}/ \min \right)/\mathrm{F}\mathrm{Q}+1.11\times {\mathrm{VCO}}_2\left(\mathrm{mL}/ \min \right)\right] $$

VCO2 was obtained from indirect calorimetry measurement, and FQ was calculated from the composition of the energy sources actually administered at the time of the measurement, including energy sources related to drug administration [11].

Data analysis and statistics

Patient characteristics are presented as mean ± standard deviation (SD), median (interquartile range), or number (percentage) as appropriate. Normality of distribution was confirmed based on the skewness and kurtosis analyses prior to parametric analyses. We used paired t tests to compare calculated EEVCO2 and EE measured by indirect calorimetry, and chi-squared tests for comparisons of proportions. The agreement between the measured EE and EEVCO2 were analyzed using Bland-Altman plots. The individual accuracy of EEVCO2 in comparison to measured EE was expressed as 5% accuracy rates defined as the proportion of calculated EEVCO2 values within the clinically relevant limits, i.e. 5% of the measured EE. The results were also evaluated by 10% accuracy rates, a more general standard of used in previous literature [11].

EEVCO2 accuracy rates were further tested according to the timing of measurement, i.e. within or after the 7th ICU day, and the rate of energy provision, i.e. <90 or ≥90% of measured EE. These factors were selected to test the hypothesis that patients treated for longer duration and fed closer to the energy targets are metabolically more stable and make better candidates for EEVCO2 [11].

All statistical analyzes were conducted on IBM SPSS Statistics® software version 22 (IBM, Corp., Armonk, NY, USA). Statistical significance was defined as p values <0.05.

Results

Patient characteristics

Patients’ characteristics are shown in (Table 1). Patients with a wide variety of primary diagnoses were analyzed, reflecting patient recruitment in a polyvalent ICU. Ninety-nine percent of the patients were fed with nutrition formulas at the time of the measurement, mainly through the enteral route (90%, including EN + PN). The remaining patients received energy only from non-nutrition energy source, i.e. drug-related glucose or lipid administration.

Table 1 Patient characteristics

Accuracy of EEVCO2 vs. measured EE

Mean biases (lower, upper 95% confidence intervals) for EEVCO2_0.85 and EEVCO2_FQ were -21 kcal/d (-41, 1) and -48 kcal/d (-67, -28), respectively (Table 2). The limits of agreement in Bland-Altman plots were (+314, -356) and (+272, -367), respectively (Fig. 1). The 5% accuracy rates compared to measured EE were 46.0 and 46.4%, while 10% accuracy rates were 77.7 and 77.3%, respectively (Table 2).

Table 2 Comparison of energy expenditure based on CO2 measurements (EEVCO2) to energy expenditure (EE) measured by indirect calorimetry in mean bias and accuracy rates
Fig 1
figure 1

Correlation and limits of agreement between calculated energy expenditure based on CO2 measurements (EEVCO2) and energy expenditure (EE) measured by indirect calorimetry. a EEVCO2_0.85 vs. measured EE. b EEVCO2_FQ vs. measured EE. Solid lines indicate the bias and fine dotted lines indicate the limits of agreement in the Bland-Altman plots. EEVCO 2 _0.85 EEVCO2 calculated from CO2 consumption (VCO2) and assumed respiratory quotient (RQ) of 0.85, EEVCO 2 _FQ EEVCO2 calculated from VCO2 and assuming RQ as food quotient (FQ)

Accuracy rates according to the timing of evaluation

There were 131 (47%) patients who were evaluated within 7 days after ICU admission, and 147 (53%) patients after the 7th day of ICU stay. There were no significant differences in the 5% accuracy rates of EEVCO2 based on the fixed RQ or the FQ when comparing the measurements performed before or after the 7th ICU day (Table 3).

Table 3 Accuracy of energy expenditure based on CO2 measurements (EEVCO2) versus energy expenditure (EE) measured by indirect calorimetry according to the timing of evaluation and energy provision rate at the time of evaluation

Accuracy rates according to the energy provision rates at the time of evaluation

There were 139 (50%) patients who were fed < 90% of the measured EE at the time of the evaluation, and 139 (50%) patients were fed ≥ 90% of EE. There were no significant differences in the 5% accuracy rates of EEVCO2 based on the fixed RQ or the FQ in patients fed < 90% of the measured EE compared to those fed ≥ 90% of EE (Table 3).

Discussion

The aim of this study was to evaluate whether EEVCO2 could be a considered as an alternative to measured EE in mechanically ventilated patients. Although the mean EEVCO2 were not considerably different from mean measured EE, 5% accuracy rates were only 46%, regardless of the RQ used for the calculation. The accuracy rates did not change for evaluations before or after the 7th ICU day, or for patients who were fed less or more than 90% of their measured EE. Therefore, we concluded that EEVCO2 should not be considered as an alternative to EE measured by indirect calorimetry, regardless of the time from ICU admission or the feeding status at the time of the evaluation.

The effect of RQ on EEVCO2

The overall accuracy of the EEVCO2 calculation method is dependent on the accuracy of the assumed RQ and the VCO2 measurements. In the present analysis, VCO2 was measured by the indirect calorimeter, thus leaving RQ as the only determinant of the difference between calculated EEVCO2 and measured EE. The RQ can be affected by metabolic consequences related to feeding and stress response secondary to critical illness, as well as non-nutrition factors such as acid-base status. In mechanically ventilated subjects, suboptimal ventilation could lead to unstable CO2 elimination, resulting in erroneous representation of the metabolic status by VCO2. In such cases, RQ may be more correctly described as respiratory exchange ratio, as RQ refers to the gas exchange ratio of the true metabolic consequence of energy oxidation. Our study demonstrates the practical implication of the RQ variability by investigating its effect on the accuracy of EEVCO2 calculation.

The measured RQ was indeed variable among patients as observed in the standard deviation (±0.09) of the measured RQ (Table 1). While optimizing the conditions for VCO2 measurements may decrease the variability of the RQ [11], this level of variability leads to a difference of > 8% of the calculated EE. By fixing the RQ at a certain value, this variability directly results in the level of inaccuracy of the calculated EE when compared with the measured EE. The use of FQ as suggested by Stapel et al. [11] did not help to improve the EEVCO2 in our analysis. The method is based on the hypothesis that FQ will more accurately predict RQ, which is not often the case in critically ill patients with variable metabolic and feeding status. The standard deviation of FQ was ±0.01, demonstrating the limited possibility of agreement with the measured RQ, which presented a much larger variation. Detailed investigation by McClave et al. concluded that RQ is neither a reliable indicator of the feeding status nor strongly associated with non-nutritional factors such as condition of ventilation and acid-base disturbance [5], suggesting the difficulty of predicting the RQ. Thus, measuring VO2 by indirect calorimetry remains as the only valid solution to determine the RQ accurately.

Another important factor is that VCO2 has less impact on the EE compared to VO2. This phenomenon can be observed in the multiples for VO2 and VCO2 in the Weir’s equation (3.941 and 1.11, respectively), giving 3.6 times more weight to the VO2 value. As a result, VO2 has a much larger influence on the EE. However, the complexity of O2 measurements in variable high O2 concentration ranges precludes the VCO2 analyses to be accurately conducted with ventilator-derived measurements. Indirect calorimeters are specially designed to solve this issue, by installing precision O2 analyzers and implementing appropriate calibration procedures.

Inaccuracy of EEVCO2

Mean EEVCO2 were in better limits of agreement to measured EE than previously reported comparisons to EE calculated by predictive equations [16]. However, individualized analysis revealed the inaccuracy hidden behind the comparison of the means. Previous studies have evaluated the accuracy of the EEVCO2 method according to 10% accuracy rates to measured EE [11, 12]. In fact, the 10% accuracy rates of EEVCO2 in our patients (77–78%) were better than in the previous study by Stapel et al. (61%) [11, 12], perhaps due to the use of VCO2 values measured by indirect calorimetry. In the present study, we implemented 5% accuracy rates according to the clinical relevancy of the results. For example, 5% accuracy for the mean measured EE of the present study (1956 kcal/d) means allowing ±98 kcal/d difference in the calculated EE; while it will be ±196 kcal/d for 10% accuracy. The results can vary within the ranges of 1858–2054 kcal/d for 5% accuracy, and 1760–2152 kcal/d for 10% accuracy. It is irrelevant to consider a method that allows almost 400 kcal/d difference in the calculated results as an alternative to measuring EE by indirect calorimetry. In addition, EEVCO2 was calculated based on the VCO2 measured by the indirect calorimeter, meaning that the difference in the results could only arise from the calculation based on assumed RQ. For these reasons, we decided to evaluate the validity of calculated EEVCO2 as an alternative to measured EE according to 5% accuracy rates.

Timing of the measurement

We anticipated that the accuracy rates for EEVCO2 would be different when measured before or after the 7th day of ICU admission. This was not the case, and suggests that the metabolic state of critically ill patients is difficult to predict, even after the 7th day of ICU admission when clinicians expect that the initial stress of critical illness starts to resolve [17]. The shift between the acute and subacute (or later) phase of critical illness is generally characterized by the shift from the catabolic to anabolic condition, and complicates the metabolic pathways [18]. Prolonged immobilization and organ support therapies can also alter the metabolism, not to mention the effect of repeated stress due to secondary infections and organs failure [17, 19, 20]. Our data thus suggest that a similar variability and complicated metabolic pathways exists also during the acute phase of critical illness [21].

The effect of energy provision

The accuracy rates of the calculated EEVCO2 also remained unchanged when the patients were fed less or more than 90% of their measured EE. This observation can partly be explained by the relationship between feeding and RQ, as the accuracy of EEVCO2 relies on the accuracy of the assumed RQ to the measured RQ. Higher rate of feeding correlates with higher RQ [5], suggesting that RQ is likely to deviate higher from the generally accepted value (i.e. 0.85) with higher rate of energy provision. At the same time, RQ is highly variable and unpredictable in critically ill patients, limiting its validity as an indicator of energy substrate oxidation [5]. Thus, variable energy provision rates can lead to variable differences between the assumed and measured RQ, leading to the inaccuracy of EEVCO2 calculations. The inaccuracy of EEVCO2 can result in misleading energy provision targets, enhancing the risk of underfeeding and overfeeding, which are both associated with a worsening of the outcome [2224].

Strengths and weaknesses

A major strength of the study is the large number of patients studied. Our analysis is based on 278 mechanically ventilated patients with various diagnoses, enhancing the generalizability of the results to most ICU patients. However, it should be noted that their length of stay was longer than the mean (<4 days) at our ICU. The selection of mechanically ventilated patients also means that patients had at least one organ failure [17], while the assumption of RQ in the calculation of EEVCO2 may be better applicable once the patients are stabilized [11].

As this was a retrospective study, the timing of the indirect calorimetry measurements was not controlled. Stratifying patients before and after the 7th ICU day may not have reflected the characteristics of acute and late-phase metabolism of ICU patients. However, we believe that this limitation can also be seen as an advantage as indirect calorimetry is usually recommended when considerable changes are observed or suspected in the patients’ conditions. In this regards, the non-significance of the stratification according to the timing of the measurement (i.e. before or after 7th ICU day) strengthened the clinical relevance of our analysis.

Another limitation arising from the retrospective nature of the study is that the analyzed measurements were conducted for clinical purpose, and not originally intended for research. Clinical conditions during the measurements such as sedation levels, ventilator types and settings, or duration of the measurements were not predefined. These factors could have affected the variability represented by the SD of the RQ, which was higher than in a previous report [11]. The VO2 and VCO2 values obtained from the clinical database were already the averaged results of the minute-by-minute readings by the Deltatrac® during the measurements, thus precluding the assessment of the stability of each measurement. We also had no comparison of 24-hour measurements, recently proposed as one of the benefits of the EEVCO2 method. However, it should be noted that our team members are trained to routinely conduct indirect calorimetry strictly according to our protocol, to ensure the quality of the clinical measurements.

The measurements of VCO2 in the present study were from indirect calorimeter, and not the measurements by mechanical ventilators as proposed in recent literature [11, 12]. We also did not conduct simultaneous measurements with EEVCO2 devices and indirect calorimeters. Our Deltatrac® has been used over the years, but has been regularly maintained by the Biomedical Department and has been calibrated before each measurement to assure optimal performance. The accuracy of VCO2 measurements of the Deltatrac® has been shown to be within 2–4% of expected values in in vitro validations [25], while the level of accuracy for VCO2 measurements in ventilators can vary as much as ±9%, according to the instructions manuals. Therefore, it is unlikely that the use of VCO2 measured by mechanical ventilators will significantly improve the accuracy of EEVCO2 and thus influence the conclusions of our study.

Conclusions

Calculated energy expenditure based on CO2 measurements (EEVCO2) was not sufficiently accurate to consider the results as an alternative to measured EE by indirect calorimetry according to this retrospective study in critically ill patients. Therefore, EE measured by indirect calorimetry remains as the gold standard to guide nutrition therapy. Prospective studies are warranted to further quantify the limitations of the EEVCO2 method.

Abbreviations

EE:

energy expenditure

EEVCO2 :

EE calculated from VCO2 and assumed RQ

FQ:

food quotient

RQ:

respiratory quotient

VCO2 :

volume of carbon dioxide production

VO2 :

volume of oxygen consumption

References

  1. McClave SA, Taylor BE, Martindale RG, Warren MM, Johnson DR, Braunschweig C, McCarthy MS, Davanos E, Rice TW, Cresci GA, et al. Guidelines for the Provision and Assessment of Nutrition Support Therapy in the Adult Critically Ill Patient: Society of Critical Care Medicine (SCCM) and American Society for Parenteral and Enteral Nutrition (A.S.P.E.N.). JPEN J Parenter Enteral Nutr. 2016;40(2):159–211.

    Article  CAS  PubMed  Google Scholar 

  2. Dhaliwal R, Cahill N, Lemieux M, Heyland DK. The Canadian critical care nutrition guidelines in 2013: an update on current recommendations and implementation strategies. Nutr Clin Pract. 2014;29(1):29–43.

    Article  PubMed  Google Scholar 

  3. Singer P, Berger MM, Van den Berghe G, Biolo G, Calder P, Forbes A, Griffiths R, Kreyman G, Leverve X, Pichard C, et al. ESPEN Guidelines on Parenteral Nutrition: intensive care. Clin Nutr. 2009;28(4):387–400.

    Article  PubMed  Google Scholar 

  4. Weir JB. New methods for calculating metabolic rate with special reference to protein metabolism. J Physiol. 1949;109(1-2):1–9.

    Article  PubMed  PubMed Central  Google Scholar 

  5. McClave SA, Lowen CC, Kleber MJ, McConnell JW, Jung LY, Goldsmith LJ. Clinical use of the respiratory quotient obtained from indirect calorimetry. JPEN J Parenter Enteral Nutr. 2003;27(1):21–6.

    Article  PubMed  Google Scholar 

  6. Guttormsen AB, Pichard C. Determining energy requirements in the ICU. Curr Opin Clin Nutr Metab Care. 2014;17(2):171–6.

    Article  PubMed  Google Scholar 

  7. De Waele E, Spapen H, Honore PM, Mattens S, Van Gorp V, Diltoer M, Huyghens L. Introducing a new generation indirect calorimeter for estimating energy requirements in adult intensive care unit patients: feasibility, practical considerations, and comparison with a mathematical equation. J Crit Care. 2013;28(5):884. e1–6.

    Article  Google Scholar 

  8. Sundström M, Fiskaare E, Tjäder I, Norberg Å, Rooyackers O, Wernerman J. Measuring Energy Expenditure in the Intensive Care Unit: a comparison of indirect calorimetry by E-sCOVX and Quark RMR with Deltatrac II in mechanically ventilated critically ill patients. Crit Care. 2016. doi:10.1186/s13054-016-1232-6.

    Google Scholar 

  9. Graf S, Karsegard VL, Viatte V, Heidegger CP, Fleury Y, Pichard C, Genton L. Evaluation of three indirect calorimetry devices in mechanically ventilated patients: which device compares best with the Deltatrac II(®)? A prospective observational study. Clin Nutr. 2015;34(1):60–5.

    Article  PubMed  Google Scholar 

  10. De Waele E, Opsomer T, Honore PM, Diltoer M, Mattens S, Huyghens L, Spapen H: Measured versus calculated resting energy expenditure in critically ill adult patients. Do mathematics match the gold standard? Minerva Anestesiol. 2015;81(3):272–82.

    PubMed  Google Scholar 

  11. Stapel SN, de Grooth HJ, Alimohamad H, Elbers PW, Girbes AR, Weijs PJ, Oudemans-van Straaten HM. Ventilator-derived carbon dioxide production to assess energy expenditure in critically ill patients: proof of concept. Crit Care. 2015;19:370.

    Article  PubMed  PubMed Central  Google Scholar 

  12. Rousing ML, Hahn-Pedersen MH, Andreassen S, Pielmeier U, Preiser JC. Energy expenditure in critically ill patients estimated by population-based equations, indirect calorimetry and CO2-based indirect calorimetry. Ann Intensive Care. 2016;6(1):16.

    Article  PubMed  PubMed Central  Google Scholar 

  13. Knaus WA, Draper EA, Wagner DP, Zimmerman JE. APACHE II: a severity of disease classification system. Crit Care Med. 1985;13(10):818–29.

    Article  CAS  PubMed  Google Scholar 

  14. Le Gall JR, Lemeshow S, Saulnier F. A new Simplified Acute Physiology Score (SAPS II) based on a European/North American multicenter study. JAMA. 1993;270(24):2957–63.

    Article  PubMed  Google Scholar 

  15. Riker RR, Fraser GL, Simmons LE, Wilkins ML. Validating the Sedation-Agitation Scale with the Bispectral Index and Visual Analog Scale in adult ICU patients after cardiac surgery. Intensive Care Med. 2001;27(5):853–8.

    Article  CAS  PubMed  Google Scholar 

  16. Graf S, Pichard C, Genton L, Oshima T, Heidegger CP. Energy expenditure in mechanically ventilated patients: the weight of body weight! Clin Nutr. 2015. doi:10.1016/j.clnu.2015.11.007. (article in press).

  17. Puthucheary ZA, Rawal J, McPhail M, Connolly B, Ratnayake G, Chan P, Hopkinson NS, Phadke R, Dew T, Sidhu PS, et al. Acute skeletal muscle wasting in critical illness. JAMA. 2013;310(15):1591–600.

    Article  CAS  PubMed  Google Scholar 

  18. Oshima T, Hiesmayr M, Pichard C. Parenteral nutrition in the ICU setting: need for a shift in utilization. Curr Opin Clin Nutr Metab Care. 2016;19(2):144–50.

    Article  CAS  PubMed  Google Scholar 

  19. Hartl WH, Jauch KW. Metabolic self-destruction in critically ill patients: origins, mechanisms and therapeutic principles. Nutrition. 2014;30(3):261–7.

    Article  PubMed  Google Scholar 

  20. Hickmann CE, Roeseler J, Castanares-Zapatero D, Herrera EI, Mongodin A, Laterre PF. Energy expenditure in the critically ill performing early physical therapy. Intensive Care Med. 2014;40(4):548–55.

    Article  PubMed  Google Scholar 

  21. Andreis DT, Singer M. Catecholamines for inflammatory shock: a Jekyll-and-Hyde conundrum. Intensive Care Med. 2016;42(9):1387–97. doi:10.1007/s00134-016-4249-z.

    Article  CAS  PubMed  Google Scholar 

  22. Ekpe K, Novara A, Mainardi JL, Fagon JY, Faisy C. Methicillin-resistant Staphylococcus aureus bloodstream infections are associated with a higher energy deficit than other ICU-acquired bacteremia. Intensive Care Med. 2014;40(12):1878–87.

    Article  CAS  PubMed  Google Scholar 

  23. Petros S, Horbach M, Seidel F, Weidhase L. Hypocaloric vs normocaloric nutrition in critically ill patients: a prospective randomized pilot trial. JPEN J Parenter Enteral Nutr. 2014;40(2):242–9.

    Article  PubMed  Google Scholar 

  24. Weijs P, Looijaard W, Beishuizen A, Girbes A, Oudemans-van Straaten HM. Early high protein intake is associated with low mortality and energy overfeeding with high mortality in non-septic mechanically ventilated critically ill patients. Crit Care. 2014;18(6):701.

    Article  PubMed  PubMed Central  Google Scholar 

  25. Takala J, Keinanen O, Vaisanen P, Kari A. Measurement of gas exchange in intensive care: laboratory and clinical validation of a new device. Crit Care Med. 1989;17(10):1041–7.

    Article  CAS  PubMed  Google Scholar 

Download references

Funding

The study was funded by the European Society for Intensive Care Medicine (ESICM), the European Society for Clinical Nutrition and Metabolism European (ESPEN), and the public Foundation Nutrition 2000Plus, Switzerland.

Availability of data and material

The datasets generated and analyzed during the current study are available from the corresponding author on reasonable request.

Authors’ contributions

TO designed the study, collected, analyzed, and interpreted the patient data, and drafted the manuscript. SG collected the patient data and participated in the analysis and interpretation of the data, and drafted the manuscript. CPH, LG, and JP participated in the analysis and interpretation of the patient data, and drafted the manuscript. CP participated in the study design, analysis and interpretation of the patient data, and drafted the manuscript. All authors have critically revised the manuscript and gave approval for the final manuscript.

Competing interests

TO received financial support as an unrestricted academic research grant from public institutions (Geneva University Hospital) and the Foundation Nutrition 2000 Plus. CP received financial support as research grants and an unrestricted academic research grant, as well as an unrestricted research grant and consulting fees, from Abbott, Baxter, B. Braun, Cosmed, Fresenius-Kabi, Nestle Medical Nutrition, Novartis, Nutricia-Numico, Pfizer, and Solvay, outside the submitted work. The other authors declare that they have no competing interests related to the current work.

Ethics approval and consent to participate

Data collection and analyses were approved by the Institutional Ethics Committee at Geneva University Hospital. Individual patient consent was waived due to the retrospective and observational nature of the study.

Author information

Authors and Affiliations

Authors

Corresponding author

Correspondence to Taku Oshima.

Additional information

An erratum to this article is available at http://dx.doi.org/10.1186/s13054-017-1650-0.

Rights and permissions

Open Access This article is distributed under the terms of the Creative Commons Attribution 4.0 International License (http://creativecommons.org/licenses/by/4.0/), which permits unrestricted use, distribution, and reproduction in any medium, provided you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons license, and indicate if changes were made. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated.

Reprints and permissions

About this article

Check for updates. Verify currency and authenticity via CrossMark

Cite this article

Oshima, T., Graf, S., Heidegger, CP. et al. Can calculation of energy expenditure based on CO2 measurements replace indirect calorimetry?. Crit Care 21, 13 (2017). https://doi.org/10.1186/s13054-016-1595-8

Download citation

  • Received:

  • Accepted:

  • Published:

  • DOI: https://doi.org/10.1186/s13054-016-1595-8

Keywords