Skip to main content

Nomenclature for renal replacement therapy in acute kidney injury: basic principles

Abstract

This article reports the conclusions of a consensus expert conference on the basic principles and nomenclature of renal replacement therapy (RRT) currently utilized to manage acute kidney injury (AKI). This multidisciplinary consensus conference discusses common definitions, components, techniques, and operations of the machines and platforms used to deliver extracorporeal therapies, utilizing a “machine-centric” rather than a “patient-centric” approach. We provide a detailed description of the performance characteristics of membranes, filters, transmembrane transport of solutes and fluid, flows, and methods of measurement of delivered treatment, focusing on continuous renal replacement therapies (CRRT) which are utilized in the management of critically ill patients with AKI. This is a consensus report on nomenclature harmonization for principles of extracorporeal renal replacement therapies. Devices and operations are classified and defined in detail to serve as guidelines for future use of terminology in papers and research.

Background

The management of critically ill patients with acute kidney injury (AKI) requiring renal replacement therapy (RRT) demands a multidisciplinary approach. In spite of previous efforts at harmonization, the terminology used to describe the different aspects and modalities of RRT is often confusing. A consensus conference on RRT terminology was organized to develop common definitions for the components, techniques, and operation of the machines and platforms used for acute extracorporeal therapies.

In this article, we report the conclusions of the consensus group on the basic principles underlying RRT technologies and the application of those principles to patient care, using “machine-centric” rather than “patient-centric” terminology. We provide a detailed description of the performance characteristics of membranes and filters, solute and fluid transport mechanisms across membranes, flow rate parameters, and methods of treatment evaluation, focusing on the continuous RRT (CRRT) used in the treatment of critically ill patients.

Methodology

A conference was organized in Vicenza, Italy, to gather experts in CRRT and members of CRRT manufacturing companies to establish consensus on technical terminology and definitions relevant to basic principles of CRRT and related technology [1]. The conference provided the background for a modified Delphi consensus methodology as previously utilized for the Acute Disease Quality initiative consensus sessions [2]. Prior to the conference, participants screened the literature of the last 25 years and previous taxonomy efforts [35]. Keywords included “continuous renal replacement therapy”, “dialysis”, “hemofiltration”, “convection”, “diffusion”, “ultrafiltration”, “dose”, “blood purification”, “renal support”, “multiorgan dysfunction”, together with the relative MeSH terms. Abstracts of 707 articles were screened and more than 300 papers were read in full and analyzed. Based on this literature search, a series of definitions and terms were proposed and consensus was achieved from the majority of experts who participated in the conference. Where consensus was lacking, different statements were created after two-thirds of the audience expressed a positive vote. We present the results of this effort of terminology harmonization called NSI (Nomenclature Standardization Initiative).

Characteristics of the membrane and filter

Geometric characteristics

The main one-dimensional geometric characteristics of hollow fiber membranes are length (L), mean inner radius (r i), wall thickness (t), and number of pores (Np). The membrane surface area depends on the number of fibers (Nf). Using these parameters, multidimensional characteristics [6] can be expressed as listed in Table 1.

Table 1 Multidimensional characteristics of the membranes

Performance characteristics

The performance characteristics define the potential applications of each membrane.

Membrane ultrafiltration coefficient and filter ultrafiltration coefficient

The membrane ultrafiltration coefficient (KUF) represents the water permeability of the filter membrane per unit of pressure and surface. It depends on both the dimensions of the membrane and the number of pores and is measured as:

$$ {K}_{UF} = \frac{Q_{UF}}{TMP}\cdot \frac{1}{A} $$

where QUF is the ultrafiltration flow rate, TMP is the transmembrane pressure, and A is the membrane surface area. The unit of measurement is ml/h/mmHg/m2. Treatment parameters that enhance or reduce pore blockage induce changes in the KUF.

The filter ultrafiltration coefficient (DKUF) is defined as the product of the KUF and membrane surface area (A):

$$ D{K}_{UF} = {K}_{UF}\cdot A $$

The unit of measurement is ml/h/mmHg. Membrane manufacturers measure DKUF as the ratio of the QUF per unit of applied TMP.

The KUF is used to define “high-flux” or “low-flux” membranes. Although there is no definitive consensus in the literature about the KUF cut-off value [7], it is generally assumed that a KUF <10 ml/h/mmHg/m2 identifies a low-flux membrane, a KUF of 10–25 ml/h/mmHg/m2 identifies middle-flux membranes, and a KUF >25 ml/h/mmHg/m2 identifies high-flux membranes.

The term high-flux has been generally used to define a membrane with an ultrafiltration coefficient >25 ml/h/mmHg/m2. This mainly describes the hydraulic permeability of the membrane (permeability to water). However, hydraulic permeability does not necessarily correspond to the permeability to solutes, which instead depends on the density of pores, the mean size of pores, and the distribution of pores. For this reason the terms high-flux and highly permeable membrane are not interchangeable.

Mass transfer area coefficient

The mass transfer area coefficient (K0A) represents the overall capacity of the membrane to provide diffusive removal of solutes over the entire filter surface. It is defined as the product of the solute flux per unit of membrane area (K0) and the membrane surface area. The unit of measurement is ml/min.

The K0A value can change during dialysis as a result of changes in membrane permeability or a loss of membrane exchange surface area.

Membrane sieving coefficient/rejection coefficient

The sieving coefficient (SC) is the ratio of a specific solute concentration in the ultrafiltrate (removed only by a convective mechanism), divided by the mean plasma concentration in the filter:

$$ SC=\frac{C_{UF}}{\left({C}_{Pi} + {C}_{Po}\right)/2} $$

where CUF is the solute concentration in the ultrafiltrate, and CPi and CPo the plasma solute concentrations at the inlet and outlet of the filter, respectively. A true calculation would require measurement of the solute concentration in plasma water rather than plasma to avoid interference of proteins. Nevertheless, for practical purposes, plasma concentration is normally accepted.

SC is correctly measurable only in the absence of a gradient for diffusion (no concentration gradient through the membrane). Measurement of the SC varies during treatment because the characteristics of the membrane change. SC is specific for each solute and for every membrane (Fig. 1). The formula is commonly simplified to the ratio between the concentration in the ultrafiltrate and the concentration in pre-filter plasma.

Fig. 1
figure 1

Schematic diagram of sieving coefficient profiles for low-flux (blue), high-flux (red) and high cut-off membranes (green)

The rejection coefficient (RC) is defined as:

$$ RC=1-SC $$

Cut-Off

For a specific membrane, the cut-off represents the molecular weight of the smallest solutes retained by the membrane. Taking into account the normal distribution of membrane pore size, the statistical cut-off value is identified as the molecular weight of a solute with a SC of 0.1. For a specific membrane, the retention onset (cut-off 90 % or 0.9) represents the molecular weight of a molecule with a SC of 0.9. For a complete understanding of the performance characteristics of a membrane, the cut-off value and the retention onset both need to be taken into account, allowing evaluation of the profile of the SC curve for each membrane (Fig. 1) [8].

Clinically, the expression “high cut-off membrane” describes membranes with a cut-off value that approximates the molecular weight of albumin (before exposure to blood or plasma).

Mechanisms of solute and fluid transport

Solute transport occurs mainly by two phenomena: convection and diffusion. Fluid transport across semipermeable membranes is driven by ultrafiltration. Adsorption influences removal of hydrophobic (lipid-soluble) compounds by attachment of solute to the membrane. When solute removal rate (mass/time) is normalized by the concentration of blood/plasma entering the filter (mass/volume), the correct term to be used is “solute clearance” which is expressed in ml/min and describes the volume of blood completely purified by the solute in the unit of time.

Ultrafiltration and convection

Ultrafiltration describes the transport of plasma water (solvent, free of cells and colloids) through a semipermeable membrane, driven by a pressure gradient between blood and dialysate/ultrafiltrate compartments. It is influenced by the intrinsic properties of the filter, such as the DKUF, and the operating parameters (e.g., TMP) [9]. Quantitatively, ultrafiltration is defined by the ultrafiltration rate (QUF):

$$ {Q}_{UF} = D{K}_{UF} \cdot \kern0.5em TMP $$

The term ultrafiltration requires some specifications depending on the context in which it is utilized. When ultrafiltration is applied to a circuit or a CRRT treatment, specifications should be made using terms such as total ultrafiltration (UF = overall ultrafiltration volume produced during treatment) and net ultrafiltration (UFNET = net ultrafiltrate volume removed from the patient by the machine). In the first case, the overall volume can be completely replaced, partially replaced, or not replaced at all. UFNET is the difference between UF and the volume replaced in the circuit (Table 2).

Table 2 Fluids and flows in continuous renal replacement therapy

When techniques are discussed, ultrafiltration may be isolated (no other mechanism is utilized in the treatment and only volume control is achieved), be used as part of hemofiltration (the ultrafiltrate is partially or completely replaced achieving volume and solute control), or combined with diffusion in treatments such as hemodialysis (HD) or hemodiafiltration (HDF). Different membranes are utilized for different techniques.

Convection is the process whereby solutes pass through membrane pores, dragged by fluid movement (ultrafiltration) caused by a hydrostatic and/or osmotic transmembrane pressure gradient.

The convective flux (Jc) of a solute depends on the QUF, the membrane surface area (A), the solute concentration in plasma (CPi) and the solute SC:

$$ {J}_c = \frac{Q_{UF}}{A}\cdotp {C}_{Pi}\cdotp SC $$

Compared to diffusive transport, convective transport permits the removal of higher molecular weight solutes at a higher rate [10].

Transmembrane pressure

In hollow fiber filters, the TMP is the pressure gradient across the membrane. The terms that define this gradient are the hydrostatic pressure in the blood compartment (PB), the hydrostatic pressure in the dialysate/ultrafiltrate compartment (PD) and the blood oncotic pressure (πB). The TMP value varies with length (l) along the whole filter length (L):

$$ TMP(l)={P}_B(l)-{P}_D(l)-{\pi}_B(l) $$

Generally, TMP is expressed using a simplified formula:

$$ TM{P}^{*}=\frac{P_{Bi}+{P}_{Bo}}{2}-\frac{P_{Di}+{P}_{Do}}{2}-\frac{\pi_{Bi}+{\pi}_{Bo}}{2} $$

where PBi is the blood inlet pressure, PBo the blood outlet pressure, PDi the dialysate/ultrafiltrate inlet pressure, PDo the dialysate/ultrafiltrate outlet pressure, πBi the oncotic pressure of the inlet blood, and πBo the oncotic pressure of the outlet blood. It must be stressed that the TMP* is a positive, averaged value along the length of filter, and does not reflect the true local pressure profile in the filter. In other words, a positive TMP* does not imply a positive TMP (l) at each point in the filter.

Furthermore, CRRT machines do not usually directly measure the PDi or the oncotic pressure, so the TMP is estimated using an even simpler formula:

$$ TM{P}^{*}=\frac{P_{PRE}+{P}_{OUT}}{2}-{P}_{EFF} $$

where PPRE is the pre-filter pressure, POUT the post-filter pressure, and PEFF the pressure measured in the effluent line (all three measured by the machine). In the most common configuration, as blood flows down the filter, plasma water is removed and eliminated with the spent dialysate (if present), which flows in a counter-current direction. This ultrafiltration, called direct (or internal) filtration, identifies the one-directional movement of plasma water from the blood side to the dialysate/ultrafiltrate compartment of the filter due to a local positive TMP(l):

$$ {P}_B(l)>{P}_D(l)+{\pi}_B(l) $$

At a critical point on the filter, where PB (l) = PD (l) + πB (l), equilibrium is achieved. After this point, the TMP (l) may become negative (even if TMP* is positive) allowing dialysate fluid to flow back into the blood compartment, resulting in so-called back filtration [11]. Back filtration describes the movement of fluid from the dialysate compartment to the blood compartment.

Diffusion

Diffusion is a process whereby molecules move randomly across a semipermeable membrane. Solute movement occurs from a more concentrated to a less concentrated area, until an equilibrium is reached between the two compartments. The concentration gradient (C1 – C2 = dc) is the driving force. The unidirectional solute diffusive flux (Jd) through a semipermeable membrane follows Fick’s law of diffusion, being directly proportional to the diffusion coefficient (D) of the solute and inversely proportional to the distance between the compartments (dx) [10]:

$$ {J}_d\kern0.5em = - D\kern0.5em \left(\frac{dc}{dx}\right) $$

The diffusivity coefficient D can be approximated using the Stokes-Einstein equation:

$$ D=\frac{k_BT}{6\pi \mu R} $$

where kB is the Boltzmann constant, T the absolute temperature, μ the viscosity of the medium, and R the effective radius of the molecules. Assuming that most molecules are globular and their effective radius is proportional to the cube root of their molecular weight, D is higher for smaller molecular weight solutes [12].

Adsorption

Adsorption is an extracorporeal process in which molecules dissolved in plasma or blood (in particular peptides and proteins) bind to the membrane structure or to other adsorbing substances such as charcoal, resins, or gels. The characteristics that influence molecule-membrane interaction are typical for each molecule (i.e., dimension, charge, and structure) and for each particular membrane (i.e., porosity, composition, hydrophobicity, surface potential). Adsorption cartridges should be evaluated in terms of their device adsorption capability (DAC) and their selectivity. DAC represents the total quantity of a specific molecule that the device is able to adsorb, and should be of the same order of magnitude as the blood concentration of that molecule multiplied by the blood volume. Selectivity is a safety parameter: it defines what the device does not adsorb.

Modalities of extracorporeal RRT

Hemodialysis

The main mechanism of solute removal in hemodialysis is diffusion, which is chiefly effective in the removal of small solutes. Hemodialysis involves the use of a hemodialyzer, where blood and dialysate solution circulate counter-current or co-current. A counter-current configuration is preferred because the average concentration gradient is kept higher along the whole length of the dialyzer. Conversely, a co-current configuration guarantees better stability and control of hydrodynamic conditions, and better air removal during the priming phase [13]. High-flux filters permit achievement of significant convective transport: this modality is called high-flux hemodialysis [14].

Hemofiltration

Hemofiltration is an exclusively ultrafiltration/convection treatment in which high-flux membranes are utilized in the absence of dialysis fluid. Infusion of a sterile solution into the blood circuit reconstitutes the reduced plasma volume and reduces solute concentration. Infusion of a sterile solution (replacement fluid) can replace totally or partially the filtered volume. Replacement fluid can be infused pre-filter (pre-dilution) or post-filter (post-dilution). In terms of solute clearance, post-dilution is more efficient than pre-dilution, but can lead more easily to membrane fouling due to hemoconcentration [9].

Hemodiafiltration

Hemodiafiltration combines hemodialysis and hemofiltration, whereby the mechanisms involved in solute removal are both diffusive and convective. Since this modality utilizes high-flux membranes, adequate amounts of sterile solution must be infused to replace the removed volume (pre-filter or post-filter) [15].

Isolated ultrafiltration

The main goal of ultrafiltration is to remove fluid using semipermeable membranes without volume replacement, thus achieving volume but not solute control in the patient [16].

Plasmapheresis

Membrane plasmapheresis filters the plasma through plasma filters and replaces it with plasma-derived products, such as fresh frozen plasma, albumin, or other fluids. Alternatively, plasma can be extracted gravimetrically from whole blood using a centrifuge pump. Plasmapheresis is used to remove hydrophilic and lipophilic high molecular weight pathogenic substances [17].

Hemoperfusion/plasmaperfusion

In hemoperfusion or plasmaperfusion, blood or plasma circulates through a column containing specific sorbents, with adsorption as the only removal mechanism. Usually combined with other modalities, hemoperfusion and plasmaperfusion are used to remove specific hydrophobic (lipid-soluble) substances, toxins, or poisons [18].

Fluids, volumes and flows

Solute transport during extracorporeal treatments strictly depends on the operating conditions including blood flow rate, dialysate, net ultrafiltration, and replacement flow rates, designed to achieve the desired clearance performance. These typical CRRT parameters (fluids and flows) are listed in Table 2.

Filtration fraction and concentration ratio

The filtration fraction (FF) is defined as the ratio between the ultrafiltration flow rate (QUF) and the plasma flow rate (QP):

$$ FF=\frac{Q_{UF}}{Q_P} $$

Filtration fraction can also be measured by the following equation:

$$ FF = \frac{1- Pro{t}_{IN}}{Pro{t}_{OUT}} $$

where ProtIN is the protein concentration in plasma entering the filter and ProtOUT is the protein concentration in plasma exiting the filter.

A directly measured FF can be expressed as a fraction:

$$ FF = \frac{Q_{UF}}{Q_P}=\frac{Q_{UF}}{Q_B\left(1-HCT\right)+{Q}_R^{PRE}} $$

where QR PRE is the pre-replacement flow rate and QB the blood flow rate.

For practical clinical purposes (as often used in CRRT machines) it is useful to define the concentration ratio (CR), which quantifies the magnitude of hemoconcentration inside the filter:

$$ CR=\frac{Q_{UF}}{Q_B + \kern0.5em {Q}_R^{PRE}}=\frac{Q_R^{POST} + {Q}_{UF}^{NET} + {Q}_R^{PRE}}{Q_B + \kern0.5em {Q}_R^{PRE}} $$

where QR POST is the post-replacement flow rate, QR PRE is the pre-replacement flow rate, and QUF NET the net ultrafiltration flow rate (all of which sum to Q UF ). Clinically, while the filtration fraction should be kept ideally below 30 %, the CR should be kept below 20–25 % [19], depending on initial hematocrit, to reduce hemoconcentration and mitigate protein-membrane interactions.

Treatment evaluation methods: the “dose” of RRT

Although the most appropriate dose has not been established for specific patients, large studies have demonstrated in the general population a direct relationship between dose and survival for both intermittent and CRRT modalities [2026]. Today, a growing body of evidence suggests the use of precision CRRT, which is characterized by the need to pay great attention to the balance between demand (of blood purification) and capacity (of the native kidney). In these circumstances, personalized prescription and monitoring of treatment dose is highly recommended [2730]. Although treatment adequacy should be considered more appropriately as a composite of different elements rather than an index based solely on urea kinetics, in CRRT a treatment efficiency equal or higher than 25 ml/kg/h is commonly considered adequate. This will approximately result in a daily standardized Kt/V = 1 which describes the efficacy of treatment for a specific patient.

Dose identifies the volume of blood cleared of waste products and toxins by the extracorporeal circuit per unit of time. In practice, it is measured as the rate of removal of a representative solute. Urea is the solute most commonly used to quantify dose [31] because it is an indicator of protein catabolism and is retained in kidney failure [12]. Originally, this solute-based approach was developed to measure the dose of dialysis prescribed to patients with end-stage renal disease. In these patients, application of this approach is relatively simple and correlates well with patient outcomes [20]. However, when using CRRT to treat critically ill patients, other measures of adequacy and dose should also be considered. One potentially easier and more reproducible means of estimating dose is incorporating the measurement of flow rates provided by the dialysis machine [32].

Multiple different definitions and formulas to calculate RRT “dose” have been proposed [33, 34]. In this section, we try to clarify the concept. During RRT, the definition of dose must include: target (patient), target (machine), current, average, projected, current effective delivered doses, and average effective delivered doses. Starting from these definitions, therapies should be identified by their efficiency, intensity, and efficacy.

Target dose (prescribed)

The target dose (prescribed) is the clearance prescribed for a specific patient in his/her specific clinical condition and represents the clearance the prescribing clinician wants to achieve in that patient.

Target machine dose (set)

The target machine dose is the clearance that the prescribing clinician wants to achieve from the machine. It is usually set as a target machine efficiency or by specifying the flow rate settings and RRT modality. The target machine dose can be modified during the treatment to reduce the mismatch between the target dose (prescribed) and the average effective delivered dose (measured).

Current dose (estimated from treatment parameters)

The current dose (estimated from treatment parameters) is the clearance at the present time estimated from the flow rates in the extracorporeal circuit. During downtime, when the machine treatment is stopped, the current dose is zero. Interruptions during the treatment can occur because of machine alarms, circuit clotting, vascular access malfunctions, or interruptions when the patient must leave the intensive care unit (ICU), such as for surgery or radiological investigations.

Average dose (measured/calculated)

The average dose is the clearance calculated for the current dose applied over the total treatment time. The total time of treatment is defined as the sum of the effective time of treatment and downtime. The effective time of treatment is the cumulative time during which the effluent pump is working. The average dose is usually an overestimate of the average effective delivered dose.

Projected dose (calculated/estimated)

The projected dose is the weighted-mean clearance that will theoretically be obtained at the end of the treatment. If the target machine dose is kept constant during treatment, the projected dose and the average dose will align. If the target machine dose is modified, the projected dose will depend on the average dose obtained until that moment and the new set target machine dose. The projected dose is usually an overestimate of the average effective delivered dose.

Current effective delivered dose (measured)

The current effective delivered dose (measured) is the clearance observed at every moment during the treatment. Unlike the current dose (estimated from treatment parameters), it is based on blood concentrations. The current effective delivered dose depends mainly on the specific RRT modality, treatment settings, and other technical and clinical issues that qualitatively and quantitatively affect clearance. The major determinants are differences between the displayed and real blood or effluent flow rates, inadequate vascular access, incorrect priming procedure, loss of surface area (clotting, air), loss of permeability (clotting of the membrane, protein cake deposition on the inner surface of membranes, concentration polarization), high blood viscosity and hematocrit, and excessive FF.

Average effective delivered dose (measured)

The average effective delivered dose (measured) or real dose is the clinically relevant (measured) clearance delivered to the patient. It is calculated on the basis of the weighted-mean of the current effective delivered dose, over the total time of treatment until that specific moment. The average effective delivered dose is the average of the current effective delivered dose during the time of treatment, and not of the current dose, because the latter is plagued by errors during times in which flow may be occurring with no solutes clearance, (e.g., bag changes, recirculation procedures). The largest discrepancies between the target dose and the average effective delivered dose are found in predominantly diffusion-based CRRT (i.e., continuous veno-venous hemodialysis and continuous veno-venous hemodiafiltration) [33].

In an ideal treatment, during which downtime and technical and/or clinical hindrances do not influence clearance, the target, target machine, current, average, projected, current effective delivered dose, and average effective delivered doses will be equal.

Efficiency, intensity and efficacy

Identified as a clearance (K), the efficiency represents the volume of blood cleared of a solute over a given period of time. It can be expressed as the ratio of blood volume over time (ml/min, ml/h, l/h, l/24 h, etc.) and is generally normalized to ideal patient weight (ml/kg/h). Efficiency depends on the reference molecules chosen (molecular size), removal mechanisms (diffusion, convection or both), and circuit operational characteristics (i.e., flow rates and type of filter). Efficiency can be used to compare different RRT treatments applied with the same modality using different settings and operational characteristics. Efficiency can be further categorized and defined as target efficiency, target machine efficiency, current efficiency, average efficiency, projected efficiency, current effective delivered efficiency, and average effective delivered efficiency. In Fig. 2, the different categories of efficiency during CRRT are illustrated with examples.

Fig. 2
figure 2

Practical example showing the different trends in efficiency (ml/kg/h, y axis) vs treatment time (h, x axis) during treatment with continuous renal replacement therapy (CRRT). Target efficiency (prescribed): “It is the amount of clearance prescribed for the specific patient in his/her specific clinical condition, and represents the amount of clearance that the doctor wants to achieve in that patient. Example: according to literature, the doctor decides that a dose of 35 ml/kg/h is the most adequate for his patient”. Target machine efficiency (set): “It is the amount of clearance that the physician wants to achieve in the machine. It is the only value that can be set in the machine. Example: taking into account the average downtime, the doctor sets the target machine dose to reach the target dose (prescribed). For example, to obtain a target dose (prescribed) of 35 ml/kg/h, the doctor sets flow rates and modalities to achieve a target machine dose of 40 ml/kg/h”. Current dose (estimated from treatment parameters): “It is the clearance at the present time, estimated considering the set flows in the extracorporeal circuit. During downtime, the current dose is zero. Example: based only on the instantaneous flow rates, the machine calculates the current dose at every moment of the treatment. A current dose of zero allows the user to recognize downtime”. Average dose (measured/calculated): “It is the clearance calculated for the current dose applied over the total time of treatment. Example: based on the total time of treatment and the current dose calculated at every moment, the machine displays the average dose. At a particular moment of the treatment, if the average dose equals 35 ml/kg/h (the target dose prescribed), the physician can assume that the patient is undertreated”. Projected dose (calculated/estimated): “It is the weighted-mean clearance that will theoretically be obtained at the end of the treatment. Example: based on the average dose obtained until a specific moment and the set target machine dose, the machine estimates the dose that theoretically will be obtained at the end of treatment session (24 h). At a particular moment during the treatment, if the projected dose is less than 35 ml/kg/h (target prescribed dose), the physician can assume that the patient will be undertreated at the end of the treatment”. Current effective delivered dose (measured): “It is the amount of clearance observed at every moment during treatment time. Unlike the current dose, it is based on blood concentrations. Example: the doctor now calculates actual blood clearance based on concentrations of solute markers. He often finds differences with the current dose (estimated from treatment parameters) because technical issues in the measurement of flow rates limit the accuracy of the estimation”. Average effective delivered dose (measured): “It is the clinically relevant amount of (measured) clearance delivered to the patient. It is calculated on the basis of the weighted-mean of the current effective delivered dose, over the total time of treatment until that specific moment”

Intensity can be defined by the product “efficiency × time”. In practice, intensity represents the blood volume cleared of a solute after a certain period of time; it can be expressed as ml or l. When comparing RRT modalities with different duration times, the use of intensity is more appropriate than the use of efficiency. For example, despite its low efficiency, use of CRRT for a long period of time results in increased treatment intensity.

Renal failure patients frequently require more than a single treatment; therefore, frequency of treatment should be considered when assessing dose. Specifically, the product of intensity times frequency (measured as treatment days/week) is useful to obtain information beyond a single treatment. Although intensity allows comparison between different treatments, it does not take into account the volume of the solute pool.

Efficacy measures the removal of a specific solute achieved by a given treatment in a given patient. It can be identified as the ratio of the entire volume cleared during the treatment to the volume of distribution of that solute. In practice, efficacy is a dimensionless number and can be numerically defined as the ratio between intensity and the volume of distribution of a specific solute.

Definitions of efficiency, intensity, and efficacy, together with the related formulas and abbreviations, are given in Table 3.

Table 3 Definitions and formulas for efficiencies, intensities and efficacies

Conclusions

Understanding the basic mechanisms underlying the process of RRT is essential to be able to make appropriate treatment choices for individual patients. Although apparently simple, those choices are in reality complex, and specific to each clinical situation.

The aim of this consensus is to standardize the nomenclature used by all parties involved in planning and delivering RRT at any level. We hope that the industry will also adopt this standard terminology in the future.

Abbreviations

r i :

Mean inner radius of the fibers

r p :

Mean inner radius of the membrane pores

ρ :

Membrane porosity

π B :

Oncotic pressure in the blood

π Bi :

Oncotic pressure of blood inlet

π Bo :

Oncotic pressure of blood outlet

A:

Membrane surface area

AKI:

Acute kidney injury

B.W.:

Ideal body weight

CBi :

Pre-filter blood concentration of the reference solute

CBo :

Post-filter blood concentration of the reference solute

CPi :

Pre-filter plasma concentration of the reference solute immediately before the filter

CPo :

Post-filter plasma concentration of the reference solute immediately after the filter

CR:

Concentration ratio

CRRT:

Continuous renal replacement therapy

CUF :

Concentration of the reference solute in the ultrafiltrate

D:

Diffusion coefficient

DAC:

Device adsorption capability

dc:

Concentration gradient

DKUF :

Filter ultrafiltration coefficient

dx:

Distance between compartments

EAed :

Average effective delivered efficacy

EAm :

Average efficacy

ECd :

Current effective delivered efficacy

ECr :

Current efficacy

EPr :

Projected efficacy

ET :

Target efficacy

ETm :

Target machine efficacy

FF:

Filtration fraction

HCT:

Hematocrit

IAed :

Average effective delivered intensity

IAm :

Average intensity

ICd :

Current effective delivered intensity

ICr :

Current intensity

ICU:

Intensive care unit

IPr :

Projected intensity

IT :

Target intensity

ITm :

Target machine intensity

Jc :

Convective flux

Jd :

Diffusive flux

K:

Clearance

K0 :

Solute flux per unit of membrane area

K0A:

Mass transfer area coefficient

KAed :

Average effective delivered efficiency

KAm :

Average efficiency

kB :

Boltzmann constant

KCd :

Current effective delivered efficiency

KCr :

Current efficiency

KPr :

Projected efficiency

KT :

Target efficiency (prescribed)

K'Tm :

New target machine efficiency set at time t1

KTm :

Target machine efficiency

KUF :

Membrane ultrafiltration coefficient

l:

Infinitesimal part of the membrane’s length

L:

Length of the membrane

Nf :

Number of fibers in the filter

Np :

Number of pores in the membrane of the filter

PB :

Hydrostatic pressure in the blood compartment

PBi :

Hydrostatic pressure in the inlet part of the blood compartment

PBo :

Hydrostatic pressure in the outlet part of the blood compartment

PD :

Hydrostatic pressure in the dialysate/ultrafiltrate compartment

PDi :

Hydrostatic pressure in the inlet part of the dialysate compartment

PDo :

Hydrostatic pressure in the outlet part of the dialysate compartment

PEFF :

Pressure in the effluent line

POUT :

Pressure in the out-flow line

PPRE :

Blood pre-filter pressure measured by the machine

ProtOUT :

Protein concentration in plasma exiting the filter

ProtIN :

Protein concentration in plasma entering the filter

QB :

Blood flow rate

QD :

Dialysate flow rate

QEFF :

Effluent flow rate

QP :

Plasma flow rate

QP-R :

Replacement plasma flow rate

QP-UF :

Plasma ultrafiltration flow rate

QR :

Total replacement flow rate

QR POST :

Replacement flow rate post-filter

QR PRE :

Replacement flow rate pre-filter

QUF :

Ultrafiltration flow rate

QUF NET :

Net ultrafiltration flow rate

R:

Radius of the molecules

RC:

Rejection coefficient

RRT:

Renal replacement therapy

SC:

Sieving coefficient

T:

Absolute temperature

t:

Thickness of the fibers

t1 :

Treatment elapsed time (0 < t1 < ttot)

TMP*:

Approximated cumulative pressure gradient across the entire membrane

TMP:

Transmembrane pressure

ttot :

Total time of treatment

UF:

Overall ultrafiltration volume produced during treatment

UFNET :

Net ultrafiltrate volume removed from the patient by the machine

V:

Volume of distribution of the reference solute

Vb F :

Filter priming volume

Vb TOT :

Total priming volume

μ:

Viscosity

References

  1. Ronco C. The Charta of Vicenza. Blood Purif. 2015;40(1):I–V.

    Article  Google Scholar 

  2. Kellum JA, Bellomo R, Ronco C. Acute Dialysis Quality Initiative (ADQI): methodology. Int J Artif Organs. 2008;31(2):90–3.

    CAS  PubMed  Google Scholar 

  3. Ronco C, Bellomo R. Continuous renal replacement therapies: the need for a standard nomenclature. Contrib Nephrol. 1995;116:28–33.

    Article  CAS  PubMed  Google Scholar 

  4. Ronco C, Bellomo R. Continuous renal replacement therapy: evolution in technology and current nomenclature. Kidney Int Suppl. 1998;66:S160–4.

    CAS  PubMed  Google Scholar 

  5. Cerda J, Ronco C. Modalities of continuous renal replacement therapy: technical and clinical considerations. Semin Dial. 2009;22(2):114–22.

    Article  PubMed  Google Scholar 

  6. Clark WR, Hamburger RJ, Lysaght MJ. Effect of membrane composition and structure on solute removal and biocompatibility in hemodialysis. Kidney Int. 1999;56(6):2005–15.

    Article  CAS  PubMed  Google Scholar 

  7. Uhlenbusch-Korwer I, Bonnie-Schorn E, Grassman A, Vienken J. Performance parameters. In: Vienken J, editors. Understanding Membranes and Dialyzers, vol 5. Pabst. 2004. p. 103–55.

  8. Boschetti-de-Fierro A, Voigt M, Storr M, Krause B. Extended characterization of a new class of membranes for blood purification: the high cut-off membranes. Int J Artif Organs. 2013;36(7):455–63.

    Article  PubMed  Google Scholar 

  9. Ficheux A, Ronco C, Brunet P, Argiles A. The ultrafiltration coefficient: this old 'grand inconnu' in dialysis. Nephrol Dial Transplant. 2015;30(2):204-8.

  10. Ronco C, Ghezzi PM, Brendolan A, Crepaldi C, La Greca G. The haemodialysis system: basic mechanisms of water and solute transport in extracorporeal renal replacement therapies. Nephrol Dial Transplant. 1998;13 Suppl 6:3–9.

    Article  CAS  PubMed  Google Scholar 

  11. Rangel AV, Kim JC, Kaushik M, Garzotto F, Neri M, Cruz DN, Ronco C. Backfiltration: past, present and future. Contrib Nephrol. 2011;175:35–45.

    Article  PubMed  Google Scholar 

  12. Leonard E. The bases of Mass Separation Processes. In: Ronco C, Bellomo R, Kellum J, editors. Critical Care Neprology. Saunders Elsevier; 2009. p. 1131-5.

  13. Kim JC, Cruz D, Garzotto F, Kaushik M, Teixeria C, Baldwin M, Baldwin I, Nalesso F, Kim JH, Kang E, et al. Effects of dialysate flow configurations in continuous renal replacement therapy on solute removal: computational modeling. Blood Purif. 2013;35(1-3):106–11.

    Article  CAS  PubMed  Google Scholar 

  14. Lee K, Jeong JH, Mun CH, Lee SR, Yoo KJ, Park YW, Won YS, Min BG. Convection-enhanced high-flux hemodialysis. Artif Organs. 2007;31(8):653–8.

    Article  PubMed  Google Scholar 

  15. Ronco C. Evolution of hemodiafiltration. Contrib Nephrol. 2007;158:9–19.

    Article  PubMed  Google Scholar 

  16. Costanzo MR, Ronco C. Isolated ultrafiltration in heart failure patients. Curr Cardiol Rep. 2012;14(3):254–64.

    Article  PubMed  Google Scholar 

  17. Nakanishi T, Suzuki N, Kuragano T, Nagasawa Y, Hasuike Y. Current topics in therapeutic plasmapheresis. Clin Exp Nephrol. 2014;18(1):41–9.

    Article  CAS  PubMed  Google Scholar 

  18. Winchester JF. Sorbent hemoperfusion in end-stage renal disease: an in-depth review. Adv Ren Replace Ther. 2002;9(1):19–25.

    Article  PubMed  Google Scholar 

  19. Ledebo I. Principles and practice of hemofiltration and hemodiafiltration. Artif Organs. 1998;22(1):20–5.

    Article  CAS  PubMed  Google Scholar 

  20. Eknoyan G, Beck GJ, Cheung AK, Daugirdas JT, Greene T, Kusek JW, Allon M, Bailey J, Delmez JA, Depner TA, et al. Effect of dialysis dose and membrane flux in maintenance hemodialysis. N Engl J Med. 2002;347(25):2010–9.

    Article  PubMed  Google Scholar 

  21. Ronco C, Bellomo R, Homel P, Brendolan A, Dan M, Piccinni P, La Greca G. Effects of different doses in continuous veno-venous haemofiltration on outcomes of acute renal failure: a prospective randomised trial. Lancet. 2000;356(9223):26–30.

    Article  CAS  PubMed  Google Scholar 

  22. Ricci Z, Ronco C. Renal replacement II: dialysis dose. Crit Care Clin. 2005;21(2):357–66.

    Article  PubMed  Google Scholar 

  23. Ronco C, Cruz D, Oudemans van Straaten H, Honore P, House A, Bin D, Gibney N. Dialysis dose in acute kidney injury: no time for therapeutic nihilism—a critical appraisal of the Acute Renal Failure Trial Network study. Crit Care. 2008;12(5):308.

    Article  PubMed  PubMed Central  Google Scholar 

  24. Bellomo R, Cass A, Cole L, Finfer S, Gallagher M, Lo S, McArthur C, McGuinness S, Myburgh J, Norton R, et al. Intensity of continuous renal-replacement therapy in critically ill patients. N Engl J Med. 2009;361(17):1627–38.

    Article  PubMed  Google Scholar 

  25. Palevsky PM, Zhang JH, O'Connor TZ, Chertow GM, Crowley ST, Choudhury D, Finkel K, Kellum JA, Paganini E, Schein RM, et al. Intensity of renal support in critically ill patients with acute kidney injury. N Engl J Med. 2008;359(1):7–20.

    Article  CAS  PubMed  Google Scholar 

  26. Schiffl H, Lang SM, Fischer R. Daily hemodialysis and the outcome of acute renal failure. N Engl J Med. 2002;346(5):305–10.

    Article  PubMed  Google Scholar 

  27. Ostermann M, Joannidis M, Pani A, Floris M, De Rosa S, Kellum JA, Ronco C. Patient Selection and Timing of Continuous Renal Replacement Therapy. Blood Purif. 2016;42(3):224–37.

    Article  PubMed  Google Scholar 

  28. Bagshaw SM, Chakravarthi MR, Ricci Z, Tolwani A, Neri M, De Rosa S, Kellum JA, Ronco C. Precision Continuous Renal Replacement Therapy and Solute Control. Blood Purif. 2016;42(3):238–47.

    Article  PubMed  Google Scholar 

  29. Cerda J, Baldwin I, Honore PM, Villa G, Kellum JA, Ronco C. Role of Technology for the Management of AKI in Critically Ill Patients: From Adoptive Technology to Precision Continuous Renal Replacement Therapy. Blood Purif. 2016;42(3):248–65.

    Article  CAS  PubMed  Google Scholar 

  30. Murugan R, Hoste E, Mehta RL, Samoni S, Ding X, Rosner MH, Kellum JA, Ronco C. Precision Fluid Management in Continuous Renal Replacement Therapy. Blood Purif. 2016;42(3):266–78.

    Article  CAS  PubMed  Google Scholar 

  31. Garred L, Leblanc M, Canaud B. Urea kinetic modeling for CRRT. Am J Kidney Dis. 1997;30(5 Suppl 4):S2–9.

    Article  CAS  PubMed  Google Scholar 

  32. Ricci Z, Bellomo R, Ronco C. Dose of dialysis in acute renal failure. Clin J Am Soc Nephrol. 2006;1(3):380–8.

    Article  PubMed  Google Scholar 

  33. Lyndon WD, Wille KM, Tolwani AJ. Solute clearance in CRRT: prescribed dose versus actual delivered dose. Nephrol Dial Transplant. 2012;27(3):952–6.

    Article  CAS  PubMed  Google Scholar 

  34. Clark WR, Turk JE, Kraus MA, Gao D. Dose determinants in continuous renal replacement therapy. Artif Organs. 2003;27(9):815–20.

    Article  PubMed  Google Scholar 

Download references

Acknowledgments

Members of the Nomenclature Standardization Initiative alliance and Signing Members of the Charta of Vicenza:

Bagshaw S. M., Critical Care Medicine, University of Alberta, Edmonton, Canada

Balducci A., Estor Spa, Pero (MI), Italy

Baldwin I., Austin Health, Heidelberg, Australia

Barbarigo F., Ospedale S. Bortolo, Vicenza, Italy

Bellomo R., The University of Melbourne, Melbourne, Australia

Braganò P., Estor Spa, Pero (MI), Italy

Braunsky J., Nx Stage Medical Inc., Lawrence, MA, USA

Calletti F., SINED srl, Cadriano di Granarolo, Bologna, Italy

Cerda J., Albany Medical College, Albany, NY, USA

Chawla L., George Washington University, Washington, DC, USA

De Rosa S., St. Bortolo Hospital and IRRIV, Vicenza, Italy

Di Somma S., Università La Sapienza e Ospedale Sant'Andrea, Roma, Italy

Doi K., Critical Care Medicine, The University of Tokyo, Tokyo, Japan

DosilRosende P., Asahi Kasei Medical GmbH, Frankfurt, Germany

Emmett M., Baylor University Medical Center, Dallas, TX, USA

Fecondini L., Medica SpA, Medolla (MO), Italy

Galavotti D., Rand Srl, Medolla (MO), Italy

Garzotto F., IRRIV, St Bortolo Hospital, Vicenza, Italy

Gibney N., University of Alberta, Edmonton, Canada

Goldstein S. L., Cincinnati Children's Hospital, Cincinnati, OH, USA

Guadagni G., Spectral Medical Inc., Toronto, Ontario, Canada

Honoré P., ICU, UZ Brussel, VUB University, Brussels, Belgium

Hoste E., Ghent University Hospital, Gent, Belgium

Inadome S., Asahi Kasei Medical GmbH, Frankfurt, Germany

Kashani K., Mayo Clinic, Rochester, MN, USA

Katz N., Johns Hopkins University, Baltimore, MD, USA

Kellum J., Center for Critical Care Nephrology, Dep. of Critical Care Medicine, University of Pittsburgh, Pittsburgh, USA

Kenley R., Aethlon Medical Inc., Libertyville, IL, USA

Kobayashi Y., Asahi Kasei Medical GmbH, Frankfurt, Germany

Lannoy J., Nikkiso, Langenhagen, Germany

Lewington A., St. James's University Hospital, Leeds, West Yorkshire, UK

Lorenzin A., IRRIV, St Bortolo Hospital, Vicenza, Italy

Mariano F., Divisione Nefrologia Dialisi e Trapianto, Città della Salute e della Scienza, Torino, Italy

McCullough P. A., Baylor University Medical Center, Dallas, TX, USA

Mehta R. L., University of California, San Diego, CA, USA

Menneguerre J., B. Braun Medical AG, Crissier, Switzerland

Mettifogo M., Dep. Nephrology Dialysis &Transplantation, St Bortolo Hospital, Vicenza, Italy

Neri M., IRRIV, St Bortolo Hospital, Vicenza, Italy

Ostermann M., Department of Critical Care & Nephrology, Guy's & St. Thomas' Hospital, London, UK

Pani A., Nefrologia e Dialisi, Ospedale G. Brotzu, Cagliari, Italy

Pirazzoli P., Bellco, Mirandola (MO), Italy

Pohlmeier R., Fresenius Medical Care, Bad Homburg v.d.H., Germany

Pouchoulin D., Gambro Industires, Meyzieu, France

Ricci Z., Ospedale Pediatrico Bambino Gesù, Roma, Italy

Ronco C., IRRIV, St Bortolo Hospital, Vicenza, Italy

Rosner M., University of Virginia, Charlottesville, VA, USA

Seamann M., Internal Medicine III, Medical University of Vienna, Vienna, Austria

Shaw A., Vanderbilt University Medical Center, Nashville, TN, USA

Tolwani A., University of Alabama at Birmingham, Birmingham, AL, USA

Villa G., University of Florence, Firenze, Italy

Vincent J.L., Erasme Hospital, Université libre de Bruxelles, Brussels, Belgium

Wendon J., King's College Hospital, London, UK

Funding

No sources of funding are declared for this manuscript.

Authors contributions

All authors attended the consensus expert conference contributed to drafting, reviewing, and editing the manuscript, and approved the final version.

Competing interests

The authors declare that they have no competing interests.

Author information

Authors and Affiliations

Authors

Consortia

Corresponding author

Correspondence to Claudio Ronco.

Rights and permissions

Open Access This article is distributed under the terms of the Creative Commons Attribution 4.0 International License (http://creativecommons.org/licenses/by/4.0/), which permits unrestricted use, distribution, and reproduction in any medium, provided you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons license, and indicate if changes were made. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated.

Reprints and permissions

About this article

Check for updates. Verify currency and authenticity via CrossMark

Cite this article

Neri, M., Villa, G., Garzotto, F. et al. Nomenclature for renal replacement therapy in acute kidney injury: basic principles. Crit Care 20, 318 (2016). https://doi.org/10.1186/s13054-016-1489-9

Download citation

  • Received:

  • Accepted:

  • Published:

  • DOI: https://doi.org/10.1186/s13054-016-1489-9

Keywords