Feasibility and safety of low-flow extracorporeal carbon dioxide removal to facilitate ultra-protective ventilation in patients with moderate acute respiratory distress syndrome
Critical Care volume 20, Article number: 36 (2016)
Abstract
Background
Mechanical ventilation with a tidal volume (VT) of 6 mL/kg/predicted body weight (PBW), to maintain plateau pressure (Pplat) lower than 30 cmH2O, does not completely avoid the risk of ventilator induced lung injury (VILI). The aim of this study was to evaluate safety and feasibility of a ventilation strategy consisting of very low VT combined with extracorporeal carbon dioxide removal (ECCO2R).
Methods
In fifteen patients with moderate ARDS, VT was reduced from baseline to 4 mL/kg PBW while PEEP was increased to target a plateau pressure – (Pplat) between 23 and 25 cmH2O. Low-flow ECCO2R was initiated when respiratory acidosis developed (pH < 7.25, PaCO2 > 60 mmHg). Ventilation parameters (VT, respiratory rate, PEEP), respiratory compliance (CRS), driving pressure (DeltaP = VT/CRS), arterial blood gases, and ECCO2R system operational characteristics were collected during the period of ultra-protective ventilation. Patients were weaned from ECCO2R when PaO2/FiO2 was higher than 200 and could tolerate conventional ventilation settings. Complications, mortality at day 28, need for prone positioning and extracorporeal membrane oxygenation, and data on weaning from both MV and ECCO2R were also collected.
Results
During the 2 h run in phase, VT reduction from baseline (6.2 mL/kg PBW) to approximately 4 mL/kg PBW caused respiratory acidosis (pH < 7.25) in all fifteen patients. At steady state, ECCO2R with an average blood flow of 435 mL/min and sweep gas flow of 10 L/min was effective at correcting pH and PaCO2 to within 10 % of baseline values. PEEP values tended to increase at VT of 4 mL/kg from 12.2 to 14.5 cmH2O, but this change was not statistically significant. Driving pressure was significantly reduced during the first two days compared to baseline (from 13.9 to 11.6 cmH2O; p < 0.05) and there were no significant differences in the values of respiratory system compliance. Rescue therapies for life threatening hypoxemia such as prone position and ECMO were necessary in four and two patients, respectively. Only two study-related adverse events were observed (intravascular hemolysis and femoral catheter kinking).
Conclusions
The low-flow ECCO2R system safely facilitates a low volume, low pressure ultra-protective mechanical ventilation strategy in patients with moderate ARDS.
Background
Over-distention of the normally aerated lung and/or opening and closing of collapsed alveoli may worsen pulmonary damage in patients with acute respiratory distress syndrome (ARDS). Current guidelines for ARDS recommend a protective ventilation strategy based on limitation of tidal volume (VT) to 6 mL/kg predicted body weight and plateau pressure (Pplat) to 30 cmH2O, an approach that has been shown in a randomized clinical trial to reduce mortality by 9 % [1]. However, recent studies have shown that ARDS patients who are ventilated according to the ARDS Network (ARDSnet) protective ventilatory strategy may still be exposed to forces that can induce lung injury [2–5], thus challenging current recommendations on how to minimize the risk of ventilator-induced lung injury (VILI) [3]. Moreover, Hager and coworkers [6] showed that mortality decreases as Pplat is reduced. However, as this relationship appears to be linear [6], several authors have postulated that an ultra-protective ventilation strategy based on further reduction in VT from 6–4 mL/kg and Pplat from 30–25 cmH2O may improve outcomes [3]. Such tidal volumes reduce alveolar ventilation resulting in respiratory acidosis, which can be mitigated through the application of extracorporeal carbon dioxide removal (ECCO2R) [7–9].
The feasibility and safety of ultra-protective ventilation strategies facilitated by ECCO2R has been tested in several studies using a pump-less arteriovenous device operating at a blood flow rate of 1.0–1.5 L/min [10–12]. Information on feasibility and safety of ultra-protective ventilation strategies facilitated by low-flow venous-venous ECCO2R are limited to a single-center study [8].
The aim of the current study was to assess in a multicenter trial the feasibility and safety of an ultra-protective ventilation strategy facilitated by low-flow veno-venous ECCO2R in patients with moderate ARDS. We used an ECCO2R system (Hemolung Respiratory Assist System, ALung Technologies), which is specifically designed to provide clinically significant CO2 removal at low blood flow rates (350–550 mL/min).
Methods
Patients were enrolled in four European intensive care units of academic hospitals. Local ethics committees approved the study protocol. Informed consent was obtained from the patients. In the case of incompetent patients, consent was obtained in accordance with local ethics committee procedures [13].
Patients
The study included fifteen adult patients with moderate ARDS according to the Berlin definition (PaO2/FiO2 (P/F) 100–200 mmHg, with positive end-expiratory pressure (PEEP) >5 cmH2O) [14], who were mechanically ventilated with an expected duration of ventilation longer than 24 h. Exclusion criteria were age <18 years, pregnancy, decompensated heart insufficiency or acute coronary syndrome, severe chronic obstructive pulmonary disease (COPD), respiratory acidosis with arterial PCO2 (PaCO2) >80 mmHg, acute brain injury, severe liver insufficiency (Child-Pugh score >7) or fulminant hepatic failure, heparin-induced thrombocytopenia, contraindication for systemic anticoagulation, patient moribund, decision to limit therapeutic interventions, catheter access to femoral vein or jugular vein impossible, pneumothorax, or platelet count <50 × 103/mL.
ECCO2R System
Low-flow ECCO2R was provided with the Hemolung Respiratory Assist System (RAS) (ALung Technologies, Inc, Pittsburgh, PA, USA) [15]. Briefly, venous blood is circulated through a 15.5-Fr dual lumen venous catheter (jugular or femoral) by a magnetically driven centrifugal pump at a flow rate of 350–550 mL/min. The pump is integrated within a cylindrical bundle of hollow fiber membranes, creating a flow pattern, which improves CO2 transfer efficiency relative to passive oxygenators. Sweep gas (air or 100 % O2) is drawn through the hollow fibers under negative pressure by a vacuum pump, creating a gradient for CO2 diffusion. Maintaining the sweep gas under negative pressure mitigates the risk of air embolism across the membrane, and also allows for automatic removal of plasmatic water condensation from the fiber lumens in order to preserve gas exchange efficiency. Level of blood flow, pump speed (RPM) and extracorporeal CO2 removal rate (vCO2) are displayed on a controller.
Study protocol
Patients were sedated, paralyzed and ventilated in accordance with the EXPRESS trial protocol [16]: VT of 6 mL/kg (ideal body weight); PEEP set to achieve Pplat of 28–30 cmH2O; respiratory rate (RR) set to 20–35 to maintain approximately the same minute ventilation as before study initiation. Percutaneous venous femoral or jugular cannulation was performed through insertion of a single dual-lumen catheter (15.5 Fr) after administration of a heparin bolus (80 IU/kg). The device was activated at a blood flow rate of 350–550 mL/min and a sweep gas of 0 L/min such that no CO2 removal was initially performed.
Following a 2-h run-in time, VT was gradually reduced from 6 to a minimum value of 4 mL/kg by 0.5 mL/kg every 30 minutes and PEEP was increased to target a Pplat between 23 and 25 cmH2O . If arterial pH was <7.25 with PaCO2 >60 mmHg, despite an increase in RR up to 35/min, sweep gas through the ECCO2R device was switched on with 100 % oxygen at 10 L/min to obtain an arterial pH ≥7.25 with a PaCO2 ≤60 mmHg and RR ≤35/min. If PaCO2 was >75 mmHg and/or pH <7.20, despite a respiratory rate of 35/min and optimized ECCO2R, sodium bicarbonate could be infused. If undesirable hypercapnia/acidosis persisted, V t was increased at the discretion of the treating physician. Refractory hypoxemia and/or hypercapnia could be managed at the discretion of the attending physician using veno-venous extracorporeal membrane oxygenation (ECMO), prone positioning, or nitric oxide (NO) inhalation. If PaCO2 was constantly <35 mmHg and/or pH was >7.50 under the aforementioned ECCO2R settings, the respiratory rate was decreased to 18–22/min and sweep gas flow was decreased to 2–5 L/min.
The ECCO2R-facilitated ultra-protective ventilation strategy was continued for at least 24 h. The potential for weaning from ultra-protective ventilation and ECCO2R was assessed daily if PaO2/FiO2 (P/F) was >200 by setting mechanical ventilation according to conventional ARDSnet settings (VT = 6 mL/kg, PEEP 5–10 cmH2O, RR 20–30/min, inspired O2 fraction (FiO2) = 40 %) and switching off sweep gas through the ECCO2R device. Under these conditions, if the patient remained stable for at least 12 h with Pplat <25 cmH2O and PaCO2 <50 mmHg (allowing for RR up to 30–35/min), ECCO2R was discontinued and the venous catheter removed.
ECCO2R parameters (blood flow, sweep gas flow, and CO2 removal rate), ventilator settings (VT, PEEP, RR, Pplat, mean airway pressure, minute ventilation, inspiratory-to-expiratory ratio, inspired fraction of oxygen), hemodynamics (mean arterial pressure, heart rate, dose of vasopressor) and arterial blood gas values (pH, PaO2, PaCO2, HCO3 –, lactate), heparin dose and activated partial thromboplastin time ratio (aPTTr) were collected at baseline, after run-in time, 30 minutes after every VT reduction and at least twice a day (08:00 am ± 2 h and 08:00 pm ± 2 h) in the subsequent days on ECCO2R. Blood chemistry data were collected daily. Respiratory system compliance [17] and driving pressure [18] were calculated according to the standard formula.
Serious adverse events (SAE) were prospectively defined as: (a) any event that is fatal or immediately life threatening, permanently disabling, severely incapacitating or requires prolonged hospitalization; or (b) any event that may jeopardize the patient and requires medical or surgical intervention to prevent one of the outcomes listed above; and (c) which the attending physician perceives might be directly related to enrollment in the clinical trial. Adverse events were considered to be study-related if the event followed a reasonable sequence from a study procedure and could readily have been produced by the study procedure. Adverse events were considered non study-related if they were related primarily to the underlying disease or to ARDS and its sequelae. Other adverse events not fulfilling the above definition were recorded in the patients’ case report forms (CRFs). Following discontinuation of ECCO2R, subjects were monitored for adverse events until hospital discharge or day 8 after enrollment, whichever occurred first.
Data are expressed as mean ± SD. Statistical analysis was performed by one-way analysis of variance for repeated measures, followed by Bonferroni post-hoc test for comparison between different time points (Stata Corp, College Station, TX, USA). P <0.05 was considered statistically significant.
Results
Fifteen patients with moderate ARDS were included in the period April to November 2014. Baseline characteristics of patients enrolled in the study are shown in Table 1.
Ventilation settings during the VT reduction phase are shown in Table 2. At baseline, all patients had PaO2/FiO2 ≤200, they were ventilated with a conventional protective ventilation strategy according to the EXPRESS trial protocol, and were paralyzed for a median time of 1 day kk. The initial stepwise reduction in VT, without ECCO2R, resulted in significant respiratory acidosis (pH <7.25) in all 15 patients at a mean VT of 3.96 ± 0.1 mL/kg. After initiation of ECCO2R, a VT of 4.29 ± 0.5 mL/kg was achieved and respiratory acidosis was significantly corrected, with pH and PaCO2 returning to within 10 % of baseline values obtained at VT = 6 mL/kg. The median number of days on ECCO2R was 3 (2–4). The reduction in VT was associated with a significant reduction in Pplat from 27.7 ± 1.6 to 23.9 ± 1 cmH2O (p <0.05) at day 1 and this difference remained significant throughout the study period (Table 3). PEEP tended to increase from 12 ± 3 to 14 ± 2 cmH2O at day 1, however, this difference was not statistically significant over time (Table 3). Driving pressure (Pplat – PEEP) was significantly reduced during the first 2 days compared to baseline (p <0.05); there were no significant differences in the values of respiratory system compliance (Fig. 1). At day 1, the ECCO2R device provided CO2 removal of 81 ± 9 mL/min at a blood flow rate of 435 ± 60 mL/min and sweep gas flow rate of 10 ± 0.3 L/min. The efficiency of the ECCO2R system was stable throughout the study period (Table 3). Infusion of bicarbonate and renal replacement therapies for acute kidney injury were never used in this cohort of patients.
At day one, the heparin dosage was 341 ± 138 IU/kg/day with an aPTT ratio of 1.77 ± 0.7, and remained stable over time. The average baseline value of hemoglobin (Hb) was 10.4 ± 2 gr/dL, and the median Hb threshold for transfusion was 6.9 gr/dL (6.9–7.0). On day 1, four patients received 2.25 ± 0.5 red blood cell (RBC) units and 1.25 ± 2.5 pools of platelets. On day 2, only two patients received 1.5 ± 0.7 units of RBCs. On day 3, only one patient received two units of RBCs.
Two study-related adverse events were observed. In one patient, intravascular hemolysis (plasma-free hemoglobin 401.4 mg/dL) was observed resulting in a discontinuation of ECCO2R after 2 days. Kinking of the ECCO2R catheter caused a reduction in circuit blood flow in another patient. Individual data for patients who needed rescue therapies for worsening hypoxemia are given in Table 4. The overall mortality at day 28 was 47 %. Among the eight survivors, six were successfully weaned from both ECCO2R and mechanical ventilation; while two patients were still dependent on ventilator support at 28 days.
Discussion
The main finding of the current study is that the low-flow ECCO2R system effectively prevents respiratory acidosis consequent to the reduction of tidal volume to 4 mL/kg in patients with moderate ARDS. Severe hypoxemia occurred in about one third of the patients and was managed by prone positioning in conjunction with ECCO2R; conversion to VV-ECMO was required in two patients. Side effects related to ECCO2R (intravascular hemolysis and pump malfunction by femoral catheter kinking) were observed in two patients.
The landmark study by the ARDSnet demonstrated a 9 % mortality reduction in patients with ARDS by limiting plateau pressure to <30 cmH2O [1]. However, recent studies have shown that patients with ARDS may still be at risk of VILI despite values of Pplat ≤30 cmH2O [2, 4, 5]. Terragni [2] and Bellani [4] showed that in some patients, tidal hyperinflation may still occur despite limiting VT to 6 mL/kg and Pplat to 30 cmH2O, and that this is associated with biological signs of VILI, such as higher levels of inflammatory mediators and increased metabolic activity of the lungs. Grasso and coworkers showed that hyperinflation might be attenuated by reducing the levels of PEEP based on the shape of the airway pressure time curve and that this is associated with lower levels of pulmonary inflammatory cytokine release [5]. Moreover, Hager and coworkers found that ARDS patients may benefit from VT reduction even if they already have Pplat <30 cmH2O [6].
ECCO2R has been proposed to partially clear CO2 and consequently reduce the need of minute ventilation as delivered by conventional mechanical ventilation [9]. The first evidence that ECCO2R might be a safe adjunctive therapy to conventional mechanical ventilation for ARDS patients dates back to the early 1980s [19, 20]. At that time, a modified veno-venous ECMO circuit with a blood flow not >1 L/min allowed for dramatic reductions of minute ventilation and ventilator-applied pressure [19]. More recently, Terragni and colleagues demonstrated that a modified renal replacement therapy circuit with an oxygenator in series with the hemofilter could facilitate an ultra-protective ventilation strategy, which may mitigate VILI [8]. Our results extend these findings by performing a prospective multicenter study in 15 ARDS patients across four European ICUs.
Monitoring of respiratory mechanics may help clinicians in assessing the risk of VILI [17]. In the presence of normal chest wall elastance, values of Pplat around 30 cmH2O may increase the risk of alveolar hyperinflation, which is the main determinant of VILI [2]; analysis of a large dataset including ARDS patients previously enrolled in clinical trials, shows that driving pressure (i.e., the ratio of tidal volume to respiratory system compliance) is an independent risk factor for hospital mortality, because regardless of the changes in VT and PEEP, only changes in driving pressure affected the outcome [18]. Our data show that an ultra-protective ventilation strategy facilitated by low-flow veno-venous ECCO2R resulted in values of respiratory mechanics associated with enhanced protection from VILI as: (1) values of Pplat around 25 cmH2O, as achieved in this study, have been associated with lower serum levels of the pro-inflammatory cytokine interleukin-6 and a smaller percentage of lung hyperinflation, as demonstrated by computed tomography [2, 7, 8]; and (2) our ultra-protective ventilation strategy resulted in a significant reduction in driving pressure. These findings allow us to speculate that in our patients, global interaction between moderate-to-high levels of PEEP and very low tidal volume, facilitated by ECCO2R, might be beneficial to minimize the risk of VILI.
In recent years, advances in technology have generated renewed interest in all extracorporeal support techniques. While high flow veno-venous ECMO has been increasingly used to treat life-threatening hypoxemia [21, 22], ECCO2R systems are used to provide partial-to-full CO2 removal with minimal effects on oxygenation [9]. Appropriate strategies to manage worsening hypoxemia during ECCO2R treatment is a compelling issue. In fact, depending on the severity of lung injury, patients with ARDS may experience worsening of arterial oxygenation that might be life threatening [23]. Moreover, the use of very low tidal volumes may expose patients to the risk of de-recruitment in the event that insufficient PEEP is applied. Consequently, prone positioning while the patient is still on ECCO2R, or shift to high-flow veno-venous ECMO, might be necessary. In this regard, prone positioning has been demonstrated to be effective not only in improving oxygenation but also in decreasing early and late mortality [24]. In the current study, indication for rescue therapies for profound hypoxemia was not established by protocol. Within the first week of enrollment 27 % of patients in our cohort required prone positioning when their PaO2/FiO2 dropped to a median value of 137. Notably, patients underwent prone position without any interruption of ECCO2R and had improvement in arterial oxygenation. Only two patients required escalation from ECCO2R to ECMO because of life-threatening hypoxemia. Mortality at 28 days was 47 %, which was expected in a cohort of patients with moderate and severe ARDS.
Previous-generation ECCO2R systems have carried a high rate of mechanical complications such as pump malfunction, membrane clotting, and catheter displacement [8]. In the current study, target blood flow rates were not reached in one case only due to a kinked catheter; otherwise, the treatment was consistent over time. Compared with arterio-venous systems in which limb ischemia, compartment syndrome, and intracranial hemorrhage have been described [7], only one case of intravascular hemolysis requiring transfusion was reported in our series.
Some limitations of this study should be addressed. First, inferences from this study are limited by its small sample size. Second, although we speculate that the strategy tested in this multicenter study was lung-protective, we did not measure pro-inflammatory mediators associated with VILI and we did not evaluate lung volume/densities distribution with computed tomography of the chest. The precise impact of worsening oxygenation during ECCO2R treatment is also not clear. Consequently, this approach will be systematically tested in an upcoming randomized clinical trial, such as that under the auspices of European Society of Intensive Care Medicine, which will test the feasibility, safety and efficacy of several ECCO2R systems to facilitate ultra-protective ventilation - VT of 4 mL/kg predicted body weight (PBW) and Pplat <25 cmH2O - in patients with moderate and severe ARDS (SUPERNOVA: a strategy of ultraprotective lung ventilation with extracorporeal CO2 removal for new-onset moderate to severe ARDS) [25].
Conclusions
In conclusion, low-flow ECCO2R is feasible, safe and efficient. It facilitates an ultra-protective mechanical ventilation strategy for reducing tidal volume to 4 mL/kg to maintain Pplat <25 cmH2O. This approach allows the delivery of mechanical ventilation with relatively low driving airway pressures while preserving sufficient gas exchange and preventing hypercapnia and respiratory acidosis due to reduced ventilation. The current study provides a clinical and physiological rationale for testing whether ultra-protective mechanical ventilation improves clinical outcomes of patients with moderate and severe ARDS in randomized control trials.
Key messages
-
Low-flow extracorporeal CO2 removal safely and effectively facilitates an ultra-protective ventilation strategy in patients with moderate ARDS
-
The current study provides a clinical and physiological rationale to study whether ultra-protective mechanical ventilation improves clinical outcomes of patients with moderate and severe ARDS in randomized control trials
References
ARDSnet. Ventilation with lower tidal volumes as compared with traditional tidal volumes for acute lung injury and the acute respiratory distress syndrome. The Acute Respiratory Distress Syndrome Network. N Engl J Med. 2000;342(18):1301–8.
Terragni PP, Rosboch G, Tealdi A, Corno E, Menaldo E, Davini O, Gandini G, Herrmann P, Mascia L, Quintel M, et al. Tidal hyperinflation during low tidal volume ventilation in acute respiratory distress syndrome. Am J Respir Crit Care Med. 2007;175(2):160–6.
Slutsky AS, Ranieri VM. Ventilator-induced lung injury. N Engl J Med. 2013;369(22):2126–36.
Bellani G, Guerra L, Musch G, Zanella A, Patroniti N, Mauri T, Messa C, Pesenti A. Lung regional metabolic activity and gas volume changes induced by tidal ventilation in patients with acute lung injury. Am J Respir Crit Care Med. 2011;183(9):1193–9.
Grasso S, Stripoli T, De Michele M, Bruno F, Moschetta M, Angelelli G, Munno I, Ruggiero V, Anaclerio R, Cafarelli A, et al. ARDSnet ventilatory protocol and alveolar hyperinflation: role of positive end-expiratory pressure. Am J Respir Crit Care Med. 2007;176(8):761–7.
Hager DN, Krishnan JA, Hayden DL, Brower RG. Tidal volume reduction in patients with acute lung injury when plateau pressures are not high. Am J Respir Crit Care Med. 2005;172(10):1241–5.
Bein T, Weber-Carstens S, Goldmann A, Muller T, Staudinger T, Brederlau J, et al. Lower tidal volume strategy (approximately 3 ml/kg) combined with extracorporeal CO2 removal versus ‘conventional’ protective ventilation (6 ml/kg) in severe ARDS: the prospective randomized Xtravent-study. Intensive Care Med. 2013;39(5):847–56.
Terragni PP, Del Sorbo L, Mascia L, Urbino R, Martin EL, Birocco A, et al. Tidal volume lower than 6 ml/kg enhances lung protection: role of extracorporeal carbon dioxide removal. Anesthesiology. 2009;111(4):826–35.
Fanelli V, Costamagna A, Ranieri VM. Extracorporeal support for severe acute respiratory failure. Semin Respir Crit Care Med. 2014;35(4):19–527.
Bein T, Weber F, Philipp A, Prasser C, Pfeifer M, Schmid F-X, et al. A new pumpless extracorporeal interventional lung assist in critical hypoxemia/hypercapnia. Crit Care Med. 2006;34(5):1372–7.
Elliot SC, Paramasivam K, Oram J, Bodenham AR, Howell SJ, Mallick A. Pumpless extracorporeal carbon dioxide removal for life-threatening asthma. Crit Care Med. 2007;35(3):945–8.
Flörchinger B, Philipp A, Klose A, Hilker M, Kobuch R, Rupprecht L, et al. Pumpless Extracorporeal Lung Assist: A 10-Year Institutional Experience. Ann Thorac Surg. 2008;86(2):410–7.
Ranieri VM, Thompson BT, Barie PS, Dhainaut JF, Douglas IS, Finfer S, et al. Drotrecogin alfa (activated) in adults with septic shock. N Engl J Med. 2012;366(22):2055–64.
Ranieri VM, Rubenfeld GD, Thompson BT, Ferguson ND, Caldwell E, Fan E, et al. Acute respiratory distress syndrome: the Berlin Definition. JAMA. 2012;307(23):2526–33.
Burki NK, Mani RK, Herth FJ, Schmidt W, Teschler H, Bonin F, et al. A novel extracorporeal CO(2) removal system: results of a pilot study of hypercapnic respiratory failure in patients with COPD. Chest. 2013;143(3):678–86.
Mercat A, Richard JC, Vielle B, Jaber S, Osman D, Diehl JL, et al. Positive end-expiratory pressure setting in adults with acute lung injury and acute respiratory distress syndrome: a randomized controlled trial. JAMA. 2008;299(6):646–55.
Grasso S, Fanelli V, Cafarelli A, Anaclerio R, Amabile M, Ancona G, et al. Effects of high versus low positive end-expiratory pressures in acute respiratory distress syndrome. Am J Respir Crit Care Med. 2005;171(9):1002–8.
Amato MB, Meade MO, Slutsky AS, Brochard L, Costa EL, Schoenfeld DA, et al. Driving pressure and survival in the acute respiratory distress syndrome. N Engl J Med. 2015;372(8):747–55.
Gattinoni L, Pesenti A, Mascheroni D, Marcolin R, Fumagalli R, Rossi F, et al. Low-frequency positive-pressure ventilation with extracorporeal CO2 removal in severe acute respiratory failure. JAMA. 1986;256(7):881–6.
Gattinoni L, Agostoni A, Pesenti A, Pelizzola A, Rossi GP, Langer M, et al. Treatment of acute respiratory failure with low-frequency positive-pressure ventilation and extracorporeal removal of CO2. Lancet. 1980;2(8189):292–4.
Peek GJ, Mugford M, Tiruvoipati R, Wilson A, Allen E, Thalanany MM, et al. Efficacy and economic assessment of conventional ventilatory support versus extracorporeal membrane oxygenation for severe adult respiratory failure (CESAR): a multicentre randomised controlled trial. Lancet. 2009;374(9698):1351–63.
Noah MA, Peek GJ, Finney SJ, Griffiths MJ, Harrison DA, Grieve R, et al. Referral to an extracorporeal membrane oxygenation center and mortality among patients with severe 2009 influenza A(H1N1). JAMA. 2011;306(15):1659–68.
Briel M, Meade M, Mercat A, Brower RG, Talmor D, Walter SD, et al. Higher vs lower positive end-expiratory pressure in patients with acute lung injury and acute respiratory distress syndrome: systematic review and meta-analysis. JAMA. 2010;303(9):865–73.
Guerin C, Reignier J, Richard JC, Beuret P, Gacouin A, Boulain T, et al. Prone positioning in severe acute respiratory distress syndrome. N Engl J Med. 2013;368(23):2159–68.
SUPERNOVA. SUPERNOVA: a strategy of ultraprotective lung ventilation with extracorporeal co2 removal for new-onset moderate to severe ARDS. http://www.esicm.org/research/trials-group/supernova.
Acknowledgements
Financial support for the study was provided by ALung Technologies, Inc.
Author information
Authors and Affiliations
Corresponding author
Additional information
Competing interests
V. Fanelli: speaker honoraria for educational programs from ALung Technologies. J. Mancebo: steering committee member for ALung Technologies Registry. O. Moerer: none. M. Quintel: none. S. Morley: employee of ALung Technologies receiving salary and stock compensation. I. Moran: none. F. Parrilla: none. A. Costamagna: none. M. Ranieri: Medical advisory board for ALung Technologies. A. Combes: Medical advisory board and speaker honoraria from ALung Technologies.
Authors’ contributions
Study conception and design: AC and VMR. Acquisition of data: VF, VMR, JM, OM, MQ, IM, FP, AC MG, and AC. Analysis and interpretation of data: VF, VMR, JM, OM, and AC. Drafting the manuscript: VF. Critical revision of the manuscript for important intellectual content: VF, VMR, JM, OM, AC, and SM. Supervision: VMR and AC. All authors read and approved the final manuscript.
Rights and permissions
Open Access This article is distributed under the terms of the Creative Commons Attribution 4.0 International License (http://creativecommons.org/licenses/by/4.0/), which permits unrestricted use, distribution, and reproduction in any medium, provided you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons license, and indicate if changes were made. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated.
About this article
Cite this article
Fanelli, V., Ranieri, M.V., Mancebo, J. et al. Feasibility and safety of low-flow extracorporeal carbon dioxide removal to facilitate ultra-protective ventilation in patients with moderate acute respiratory distress syndrome. Crit Care 20, 36 (2016). https://doi.org/10.1186/s13054-016-1211-y
Received:
Accepted:
Published:
DOI: https://doi.org/10.1186/s13054-016-1211-y