Open Access

Functional hemodynamic monitoring

Critical Care20059:566

DOI: 10.1186/cc3927

Published: 22 November 2005


Hemodynamic monitoring is a central component of intensive care. Patterns of hemodynamic variables often suggest cardiogenic, hypovolemic, obstructive, or distributive (septic) etiologies to cardiovascular insufficiency, thus defining the specific treatments required. Monitoring increases in invasiveness, as required, as the risk for cardiovascular instability-induced morbidity increases because of the need to define more accurately the diagnosis and monitor the response to therapy. Monitoring is also context specific: requirements during cardiac surgery will be different from those in the intensive care unit or emergency department. Solitary hemodynamic values are useful as threshold monitors (e.g. hypotension is always pathological, central venous pressure is only elevated in disease). Some hemodynamic values can only be interpreted relative to metabolic demand, whereas others have multiple meanings. Functional hemodynamic monitoring implies a therapeutic application, independent of diagnosis such as a therapeutic trial of fluid challenge to assess preload responsiveness. Newer methods for assessing preload responsiveness include monitoring changes in central venous pressure during spontaneous inspiration, and variations in arterial pulse pressure, systolic pressure, and aortic flow variation in response to vena caval collapse during positive pressure ventilation or passive leg raising. Defining preload responsiveness using these functional measures, coupled to treatment protocols, can improve outcome from critical illness. Potentially, as these and newer, less invasive hemodynamic measures are validated, they could be incorporated into such protocolized care in a cost-effective manner.


Hemodynamic monitoring is a cornerstone of care for the hemodynamically unstable patient, but it requires a manifold approach and its use is both context and disease specific. One of the primary goals of hemodynamic monitoring is to alert the health care team to impending cardiovascular crisis before organ injury ensues; it is routinely used in this manner in the operating room during high-risk surgery. Another goal of hemodynamic monitoring is to obtain information specific to the disease processes, which may facilitate diagnosis and treatment and allow one to monitor the response to therapy.

The effectiveness of hemodynamic monitoring depends both on available technology and on our ability to diagnose and effectively treat the disease processes for which it is used. The utility of hemodynamic monitoring has evolved as it has merged with information technology and as our understanding of disease pathophysiology has improved. Within this context, hemodynamic monitoring represents a functional tool that may be used to derive estimates of performance and physiological reserve that may in turn direct treatment. However, no monitoring device can improve patient-centered outcomes useless it is coupled to a treatment that improves outcome. Thus, hemodynamic monitoring must be considered within the context of proven medical therapies, success of which is dependent on the clinical condition, pathophysiological state and ability to reverse the identified disease process.

Rationale for hemodynamic monitoring

A progression of arguments supporting the use of specific monitoring techniques can be proposed. At the basic level, monitoring can be defended on the basis of historical controls. In this regard, prior experience with similar monitoring techniques indicates that they can identify known complications that are undetectable with less invasive means. Clearly, the mechanism by which the benefit is achieved need not be understood or even postulated.

Further support for hemodynamic monitoring comes from an understanding of the pathophysiology of the process being treated, such as heart failure or hypovolemic shock. Weil and Shubin [1] defined circulatory shock as decreased ability of blood flow to meet the metabolic demands of the body. Using their classic approach, four basic groups of circulatory shock can be defined: hypovolemic, cardiogenic, obstructive, and distributive. Certain combinations of hemodynamic findings allow the etiology of circulatory shock to be defined using this nosology. Tissue hypoperfusion is common in all forms of shock (with the possible exception of hyperdynamic septic shock). Because specific types of circulatory shock require different therapies and target end-points of resuscitation, defining the cardiovascular state is important in determining both treatment options and their goals. Much of the rationale for hemodynamic monitoring resides at this level. The implied assumption here is that knowledge of how a disease process creates its effect will allow one to prevent the process from altering measured bodily functions, thus preventing disease progression and promoting recovery. This argument may not be valid, primarily because knowledge of the specific process in individual patients is often inadequate. Furthermore, measures of global blood flow and systemic arterial pressure, and changes in them in response to shock and its treatment poorly reflect regional and microcirculatory blood flow [24].

The most important support for hemodynamic monitoring is that, by altering therapy in otherwise unexpected ways, it can improve outcome in terms of survival and quality of life. Few therapies can claim such a benefit, although the trial by Rivers and coworkers [5] represents a notable exception in this regard; those investigators reported that measures of blood flow sufficiency and resuscitation to sustain blood flow improved outcome from septic shock.

Hemodynamic monitoring must also be considered within the context of the patient, pathophysiology, time point in the disease process, and position within the health care delivery system at which it is used. The site where monitoring takes place has a major impact on type of monitoring, and its risks, utility and efficacy. Monitoring outside the hospital and emergency department may be less invasive than in the operating room or intensive care unit. The time point during the course of disease when monitoring is applied will also have profound effects on outcome. For example, preoperative optimization of cardiovascular status [6] and emergency department early goal-directed therapy in septic shock [5] reduces morbidity, whereas the same monitoring and treatment applied after injury in unstable patients with existing shock-induced organ injury does not improve outcome [79].

Static hemodynamic monitoring variables

Specific hemodynamic variables are commonly measured and displayed at the bedside, and their values are often used in clinical decision making. However, the utility of each variable as a single absolute value is questionable. Some individual hemodynamic values are useful primarily as threshold monitors. For example, because a primary determinate of organ perfusion is perfusion pressure, systemic hypotension to below a certain threshold is clinically relevant. Furthermore, elevation in central venous pressure (CVP; i.e. >10 mmHg) reflects right ventricular (RV) pressure overload, although this gives no information on the precise etiology involved. Other hemodynamic values can only be interpreted relative to metabolic demand. For example, because blood flow varies to match metabolic requirements, which in turn can vary considerably, there is no one specific value of cardiac output or oxygen delivery that can be defined as 'normal'. These characteristics of blood flow reflect either an ability or an inability to meet the metabolic demands of the body. Finally, other measures are of questionable value in evaluating one parameter but are important in monitoring another. For example, pulmonary arterial occlusion pressure (Ppao) is a poor measure of left ventricular (LV) preload but is a good measure of the back-pressure to pulmonary blood flow and the hydrostatic forces producing pulmonary edema.

Some specific uses for hemodynamic values measured at a single point in time are described in Table 1. Although grouping hemodynamic variables in order to define profiles can improve diagnostic accuracy, there are few reports of improved outcomes resulting from such refinements in data analysis. Nevertheless, in the following discussion we consider the primary hemodynamic measures that are commonly used in critically ill patients.
Table 1

Clinical caveats for hemodynamic variables

Type of hemodynamic variable




Blood pressure

Hypotension is always pathological


Central venous pressure (CVP)

CVP is only elevated in disease


Pulmonary artery occlusion pressure (Ppao)

Ppao is the back-pressure to pulmonary blood flow


Cardiac output

There is no normal cardiac output, only an adequate or inadequate one


Mixed venous oxygen saturation (SvO2)

Decreasing SvO2 is a sensitive but nonspecific marker of cardiovascular stress


Volume challenge

Positive response defined as an increase in any of blood pressure, CVP, Ppao, cardiac output and/or SvO2, or a decrease in heart rate


Echocardiographic analysis of vena cavae collapse during positive pressure inspiration identifies CVP <10 mmHg if it detects

Complete inferior vena caval collapsea


>36% collapse in superior vena cavaa


Defining preload responsiveness

≥13% pulse pressure variation during positive pressure ventilationa


>1 mmHg decrease in CVP during spontaneous inspirationb

aRequires a fixed tidal volume of 6–8 ml/kg and complete adaptation to the ventilator. bRequires a spontaneous inspiratory effort greater than -2 mmHg to be valid.

Blood pressure

Arterial blood pressure is not a single pressure but a range of pressure values from systole and diastole. Mean arterial pressure best approximates the organ perfusion pressure in noncardiac tissues, as long as venous or surrounding pressures are not elevated. Arterial pressure is commonly measured noninvasively on an intermittent basis using a sphygmomanometer [10]. Indwelling arterial catheters permit continuous monitoring of arterial pressure. Because blood pressure is a regulated variable, a normal blood pressure does not necessarily reflect hemodynamic stability [11]. Organ systems also tend to autoregulate their blood flow such that organ-specific blood flow remains constant within a wide range of blood pressures if metabolic rate is unchanged, and varies with changes in local metabolic rate. The lower limit of this flow autoregulation, based on mean arterial pressure, varies between organs, patients (based on their underlying circulatory status, for example essential hypertension or peripheral vascular disease), their disease state, their metabolic activity, and associated vasoactive therapies.

Thus, there is no threshold blood pressure value that defines adequate organ perfusion among organs, between patients, or in the same patient over time [12]. However, because arterial pressure is a primary determinant of organ blood flow, hypotension (mean arterial pressure <65 mmHg) is always pathological.

Central venous pressure

CVP is the back-pressure to systemic venous return. Because CVP is usually very low, defining the appropriate hydrostatic zero level is important in estimating CVP, but such physiological zeroing can be difficult. Few absolutes can be stated regarding static measures of CVP. If CVP is 10 mmHg or less then cardiac output will uniformly decrease when 10 cmH2O positive end-expiratory pressure is given to ventilator-dependent patients [13], whereas a CVP above 10 mmHg has no predictive value. Demonstration, using echocardiographic techniques, of more than 36% superior vena caval collapse during positive-pressure inspiration [14] or complete inferior vena caval collapse [15, 16] identifies individuals whose CVP is below 10 mmHg. However, there is no threshold value of CVP that identifies patients whose cardiac output will increase in response to fluid resuscitation [17]. Importantly, CVP is only elevated in disease, but the clinical utility of CVP as a guide to diagnosis or therapy has not been demonstrated.

Pulmonary artery catheter

The pulmonary artery catheter (PAC) permits LV filling pressures to be estimated by measuring Ppao [18, 19]. However, Ppao values do not correlate with LV end-diastolic volume, and neither do they predict preload responsiveness [20]. Nevertheless, Ppao is the back-pressure to pulmonary blood flow, and it can be used to identify the presence of a hydrostatic component to pulmonary edema and to assess pulmonary vascular resistance. Using a rapid response thermistor, the PAC can be used to monitor RV end-diastolic volume based on measures of residual thermal signal. Measures of changes in RV end-diastolic volume are useful in cardiac surgery when trying to identify right-sided cardiac failure. If RV end-diastolic volume increases as cardiac output decreases, then the patient has cor pulmonale [21]. Using a transthoracic measure of thermal decay, one can estimate intrathoracic blood volume, global cardiac volume, and lung water. Of these three measures, intrathoracic blood volume is presently the most widely used technique, although intrathoracic lung water measures may be of interest in the management of patients with acute lung injury. Intrathoracic blood volume and its changes in response to fluid challenge reflect LV preload and changes in LV preload better than do more conventional measures, such as CVP or Ppao [17, 22]. However, the utility of any of these measures as static single-point values in predicting preload responsiveness or in improving outcome in unstable patients has not been documented.

Indicator dilution techniques using thermal, indocyanine green, and lithium can measure blood flow from both central venous and PAC [23]. LV stroke volume can be estimated using a beat-to-beat based, algorithmic analysis of arterial pulse pressure [24]. Several monitoring techniques use subtle variations in this concept to calculate stroke volume and cardiac output. The overall accuracy of these techniques varies. Esophageal Doppler techniques can be used to measure descending aortic flow [2527] and to estimate both stroke volume and cardiac output. Because accurate measurement of cardiac output is less important than accurate documentation of trends in flow, these measures may have profound clinical utility if they are accurate and stable over time.

Recall that there is no normal cardiac output; because cardiac output varies with metabolic demand, it is either able or unable to meet these demands. Measures of mixed venous oxygen saturation (SvO2) may reflect better the adequacy of oxygen delivery. The normal value for SvO2 is 75–70%. Exercise, anemia, hypoxemia, and decreased cardiac output all independently decrease SvO2. Although SvO2 above 70% does not necessarily reflect adequate tissue oxygenation, a persistently low SvO2 (>30%) is associated with tissue ischemia [28]. Measures of central venous oxygen saturation (ScvO2) tend to track SvO2. However, ScvO2 and SvO2 are not equal, and so use of ScvO2 to define thresholds of resuscitation requires one to give special attention to related clinical variables [29]. Splanchnic oxygen consumption can be estimated from hepatic venous oxygen saturation and hepatic venous blood flow measures [30]. However, this measure has not been shown to be superior to less invasive techniques in directing resuscitation or in improving outcome.

Hypoperfusion initially decreases blood flow but not oxidative phosphorylation; thus, tissue partial carbon dioxide tension (PCO2) reflects both local metabolism and regional blood flow. If blood flow decreases then tissue PCO2 will increase relative to arterial PCO2. Measurement of this PCO2 gap could allow one to assess whether tissue blood flow is effective. Measures of gastric [31] and sublingual [32, 33] PCO2 gaps identify tissue hypoperfusion. Gastric tonometry is useful in guiding resuscitation in critically ill patients [31], and sublingual PCO2 measures may have similar utility.

Functional hemodynamic monitoring: defining response to therapy

Although one may use hemodynamic monitoring to identify cardiovascular insufficiency before it results in clinical hypoperfusion or as a prognostic indicator of survival, its greatest potential role is in directing application of cardiovascular therapies that are of proven efficacy. Monitoring conducted to evaluated the effect of treatment can be referred to as functional monitoring, because it implies a therapeutic application. Although trends in specific variables over time are useful in defining hemodynamic stability, their rapid change in response to application of a therapy has greater clinical utility. Some examples of functional monitoring variables are given in Table 1. The most common example of functional monitoring is in a therapeutic trial. Below we list the various types of functional monitoring presently validated.

Volume challenge

The time-honored method of assessing preload responsiveness is to administer a relatively small intravascular volume bolus rapidly and observe the subsequent hemodynamic response in terms of blood pressure, pulse, cardiac output, SvO2 and related measures. There is little agreement regarding what absolute volume and infusion rate defines an adequate fluid challenge. In a volume challenge trial estimates of improved circulatory status (e.g. increasing blood pressure and decreasing heart rate) and improved effective blood flow (e.g. increasing SvO2 and decreasing blood lactate) are used to document a beneficial response. The primary factor addressed by a fluid challenge is preload responsiveness; specifically, will cardiac output increase with fluid loading? Importantly, being preload responsive does not equate to requiring fluid resuscitation. Normal individuals are preload responsive but do not require resuscitation. Thus, a fluid challenge must be conducted within the context of known or suspected tissue hypoperfusion [34]. Furthermore, a volume challenge is not fluid resuscitation; it is merely a test to identify those who are preload responsive [35]. Volume responders can then be given additional fluid resuscitation with minimal risk for worsening cor pulmonale or inducing pulmonary edema.

However, a volume change, as a primary diagnostic approach in hemodynamically unstable patients, has important clinical drawbacks. First, only half of all hemodynamically unstable patients are preload responsive [36]. Second, it delays primary therapy in a setting where delayed appropriate treatment has consequences for survival. Finally, a volume challenge in an unresponsive patient may worsen or precipitate pulmonary edema or cor pulmonale. Therefore, several surrogate methods of creating reversible or transient volume challenges, including breathing and passive leg raising, have been advocated.

Passive leg raising

Passive leg raising to 30° transiently increases venous return [37] in patients who are preload responsive. Leg raising only transiently increases cardiac output in responders, and so it is not a treatment for hypovolemia. When coupled with measures of aortic flow, patients exhibiting a sustained (15 s) increase in mean aortic flow 30 s after leg raising were found to be preload responsive [38]. The advantages of this approach are that it is easy perform, induces only a transient and reversible volume challenge, yields volume challenges proportional to individual body size, and can be repeated as needed to reassess preload responsiveness. Limitations of the technique are that, presently, only measures of mean aortic flow, using esophageal Doppler, can assess preload responsiveness and that the blood volume mobilized by leg raising is dependent on total blood volume and so could be small in severely hypovolemic patients [39].

Changes in central venous pressure during spontaneous breathing

With spontaneous inspiration, venous return normally increases in association with the decrease in intrathoracic pressure [40]. If the right ventricle can transfer this transient bolus of blood into the pulmonary circulation, then CVP will correlate with intrathoracic pressure, decreasing with each spontaneous inspiratory effort. An inspiratory decrease in CVP of more than 1 mmHg in the setting of an intrathoracic pressure decrease of more than 2 mmHg accurately predicts preload responsiveness, whereas those patients whose CVP does not decrease do not increase their cardiac output in response to fluid challenge [41]. This simple approach requires central venous catheterization, and one must give close attention to CVP waveform analysis. During positive pressure ventilation, the interpretation of CVP as reflected by changes in inferior vena cava diameter is complex and of minimal diagnostic utility.

Changes in left ventricular output during positive pressure ventilation

If changes in both right and left ventricles induce changes in output, then one can use positive pressure ventilation to assess the dynamic and necessarily cyclic effect of ventilation on venous return by assessing dynamic swings in LV output. The greater the increase in tidal volume for the same lung compliance, the greater is the transient decrease in venous return and subsequently greater decrease in LV output [42]. Changes in systolic arterial pressure during a programmed series of increasing tidal breaths quantify the degree of preload responsiveness [43]. Furthermore, during fixed tidal volume positive pressure ventilation, variations in systolic pressure [44], pulse pressure [45], LV stroke volume [46], and aortic flow [47] are robust measures of preload responsiveness. Michard and coworkers [45] found that a systolic pressure or a pulse pressure variation of 13% or more in septic patients breathing with a tidal volume of 8 ml/kg is highly sensitive and specific for preload responsiveness. In contrast, no threshold values for either Ppao or CVP could be identified that were better than random chance in predicting preload responsiveness.

One can estimate LV stroke volume based on the arterial pressure pulse contour. Several studies conducted in patients undergoing surgery have documented a good relation between this measure of pulse contour derived stroke volume variation and preload responsiveness [48]. Unfortunately, the accuracy of the pulse contour algorithm used to calculate stroke volume is proprietary and has changed on commercially available devices since these validation studies were preformed [49]. Thus, the extent to which these measures accurately track real stroke volume fluctuations is unclear. Furthermore, because these various devices calculate stoke volume differently, the threshold values for each parameter in predicting preload responsiveness may be different between devices, and may exhibit different degrees of robustness under varying clinical conditions. Changes in vasomotor tone [50] will also alter the observed changes in each parameter and may do so to proportionally different degrees. Thus, more clinical validation work must be done on these measures before they may become standard measures in most intensive care units.

Standardization of care

The application of evidence-based guidelines to clinical practice is rational. This approach often reduces health care costs by reducing practice variations, medical errors, and length of stay [51]. Fluid optimization as an end-point of resuscitation reduces length of hospital stay and important complications in patients undergoing a variety of major surgical procedures that routinely require postoperative resuscitation, but the degree of this effect varies among patients [5254]. Cost-effectiveness analyses of specific types of treatment directed by hemodynamic monitoring, namely SvO2 monitoring to identify adequacy of treatment for hemodynamic instability [55] and preoptimization in high-risk surgery patients [56], have demonstrated benefit. These studies underscore the importance of examining the utility of monitoring systems within the context of a specific disease process coupled to effective treatment protocols.


Fundamentally, one may ask just three questions regarding the cardiovascular system during resuscitation from shock [57]: will blood flow to the body increase with fluid resuscitation?; is arterial hypotension due to inadequate blood flow or loss of vasomotor tone, or both?; and is the heart capable of maintaining effective blood flow without going into failure? If the answer to the first question is 'yes', then treatment must include volume expansion. However, if the patient is also hypotensive and has reduced vasomotor tone, then vasopressor therapy may be started simultaneously because arterial pressure will not increase with volume expansion alone, even though cardiac output will increase. If the patient is not preload responsive but has reduced vasomotor tone associated with hypotension, then a vasopressor alone is indicated. If the patient is neither preload responsive nor exhibiting reduced vasomotor tone and hypotension, then the problem is the heart, and both diagnostic and therapeutic actions must be taken to address these specific problems (e.g. echocardiography, dobutamine). Protocolized cardiovascular management based on functional hemodynamic monitoring has the added advantages of being intuitively obvious (facilitating buy-in by stakeholders), pleuripotential (many different monitoring devices can all drive the same protocol) and scalable (alter intensity of resuscitation), and lends itself to automation.



central venous pressure


left ventricular


pulmonary artery catheter


partial carbon dioxide tension


pulmonary arterial occlusion pressure


right ventricular


central venous oxygen saturation


mixed venous oxygen saturation.



Dr Pinsky was supported in part by NIH grants HL67181 and HL07820. This review is based on a ESICM/SCCM sponsored Roundtable Meeting, held in 27–29 March 2004 in Brussels, Belgium. Chairpersons of the meeting were Michael R Pinsky and Didier Payen. Participants included David Bennett, Jacques Creteur, Daniel De Backer, Phillip Dellinger, Luciano Gattinoni, Alwin Goetz, Johan Groeneveld, Jessie Hall, Can Ince, Sheldon Magder, John Marini, Monty Mythen, Azriel Perel, Peter Radermacher, Konrad Reinhart, Andrew Rhodes, Mervyn Singer, Michel Slama, Jean-Louis Teboul, Jean-Louis Vincent, Max H Weil, and Julia Wendon.

Authors’ Affiliations

Professor of Critical Care Medicine, Bioengineering and Anesthesiology, Department of Critical Care Medicine, University of Pittsburgh Medical Center
Professor of Anesthesiology, Department of Anesthesiology and Critical Care Medicine, Lariboisière Hospital, University of Paris VII


  1. Weil MH, Shubin H: Shock following acute myocardium infarction: current understanding of hemodynamic mechanisms. Prog Cardiovasc Dis 1968, 11: 1-17.View ArticlePubMedGoogle Scholar
  2. Schmidt-Nielsen K: Circulation. In Animal Physiology. 4th edition. Cambridge, UK: Cambridge University Press; 1983:97-133.Google Scholar
  3. Lush CW, Kvietys PR: Microvascular dysfunction in sepsis. Microcirculation 2000, 7: 83-101. 10.1038/ ArticlePubMedGoogle Scholar
  4. Buwalda M, Ince C: Opening the microcirculation: can vasodilators be useful in sepsis? Intensive Care Med 2002, 28: 1208-1217. 10.1007/s00134-002-1407-2View ArticlePubMedGoogle Scholar
  5. Rivers E, Nguyen B, Havstad S, Ressler J, Muzzin A, Knoblich B, Peterson E, Tomlanovich M: Early goal-directed therapy in the treatment of severe sepsis and septic shock. N Engl J Med 2001, 345: 1368-1377. 10.1056/NEJMoa010307View ArticlePubMedGoogle Scholar
  6. Bland RD, Shoemaker WC, Abraham E, Cobo JC: Hemodynamic and oxygen transport patterns in surviving and nonsurviving postoperative patients. Crit Care Med 1985, 13: 85-90.View ArticlePubMedGoogle Scholar
  7. Gattinoni L, Brazzi L, Pelosi P, Latini R, Tognoni G, Pesenti A, Fumagalli R: A trial of goal-oriented hemodynamic therapy in critically ill patients. SvO 2 Collaborative Group. N Engl J Med 1995, 333: 1025-1032. 10.1056/NEJM199510193331601View ArticlePubMedGoogle Scholar
  8. Heyland DK, Cook DJ, King D, Kernerman P, Brun-Buisson C: Maximizing oxygen delivery in critically ill patients: a methodologic appraisal of the evidence. Crit Care Med 1996, 24: 517-524. 10.1097/00003246-199603000-00025View ArticlePubMedGoogle Scholar
  9. Kern JW, Shoemaker WC: Meta-analysis of hemodynamic optimization in high-risk patients. Crit Care Med 2002, 30: 1686-1692. 10.1097/00003246-200208000-00002View ArticlePubMedGoogle Scholar
  10. Bur A, Hirschl MM, Herkner H, Oschatz E, Kofler J, Woisetschlager C, Laggner AN: Accuracy of oscillometric blood pressure measurement according to the relation between cuff size and upper-arm circumference in critically ill patients. Crit Care Med 2000, 28: 371-376. 10.1097/00003246-200002000-00014View ArticlePubMedGoogle Scholar
  11. Partrick DA, Bensard DD, Janik JS, Karrer FM: Is hypotension a reliable indicator of blood loss from traumatic injury in children? Am J Surg 2002, 184: 555-559. 10.1016/S0002-9610(02)01052-8View ArticlePubMedGoogle Scholar
  12. LeDoux D, Astix ME, Carpati C, Rackow EC: Effects of perfusion pressure on tissue perfusion in septic shock. Crit Care Med 2000, 28: 2729-2732. 10.1097/00003246-200008000-00007View ArticlePubMedGoogle Scholar
  13. Jellinek H, Krafft P, Fitzgerald RD, Schwartz S, Pinsky MR: Right atrial pressure predicts hemodynamic response to apneic positive airway pressure. Crit Care Med 2000, 28: 672-678. 10.1097/00003246-200003000-00012View ArticlePubMedGoogle Scholar
  14. Vieillard-Baron A, Chergui K, Rabiller A, Peyrouset O, Page B, Beauchet A, Jardin F: Superior vena caval collapsibility as a gauge of volume status in ventilated septic patients. Intensive Care Med 2004, 30: 1734-1739.PubMedGoogle Scholar
  15. Vieillard-Baron A, Augarde R, Prin S, Page B, Beauchet A, Jardin F: Influence of superior vena caval zone condition on cyclic changes in right ventricular outflow during respiratory support. Anesthesiology 2001, 95: 1083-1088. 10.1097/00000542-200111000-00010View ArticlePubMedGoogle Scholar
  16. Barbier C, Loubieres Y, Schmit C, Hayon J, Ricome JL, Jardin F, Vieillard-Baron A: Respiratory changes in inferior vena cava diameter are helpful in predicting fluid responsiveness in ventilated septic patients. Intensive Care Med 2004, 30: 1740-1746.PubMedGoogle Scholar
  17. Kumar A, Anel R, Bunnell E, Habet K, Zanotti S, Marshall S, Neumann A, Ali A, Cheang M, Kavinsky C, Parrillo JE: Pulmonary artery occlusion pressure and central venous pressure fail to predict ventricular filling volume, cardiac performance, or the response to volume infusion in normal subjects. Crit Care Med 2004, 32: 691-699. 10.1097/01.CCM.0000114996.68110.C9View ArticlePubMedGoogle Scholar
  18. Pinsky MR, Vincent JL, DeSmet JM: Estimating left ventricular filling pressure during positive end-expiratory pressure in humans. Am Rev Respir Dis 1991, 143: 25-31.View ArticlePubMedGoogle Scholar
  19. Teboul JL, Pinsky MR, Mercat A, Nadia A, Bernardin G, Achard J-M, Boulain T, Richard C: Estimating cardiac filling pressure in mechanically ventilated patients with hyperinflation. Crit Care Med 2000, 28: 3631-3636. 10.1097/00003246-200011000-00014View ArticlePubMedGoogle Scholar
  20. Michard F, Boussat S, Chemla D, Anguel N, Mercat A, Lecarpentier Y, Richard C, Pinsky MR, Teboul J-L: Relation between respiratory changes in arterial pulse pressure and fluid responsiveness in septic patients with acute circulatory failure. Am J Respir Crit Care Med 2000, 162: 134-138.View ArticlePubMedGoogle Scholar
  21. Hines R, Rafferty T: Right ventricular ejection fraction catheter: toy or tool? Pro: a useful monitor. J Cardiothorac Vasc Anesth 1993, 7: 236-240. 10.1016/1053-0770(93)90224-9View ArticlePubMedGoogle Scholar
  22. Payen DM, Brun-Buisson CJ, Carli PA, Huet Y, Leviel F, Cinotti L, Chiron B: Hemodynamic, gas exchange, and hormonal consequences of LBPP during PEEP ventilation. J Appl Physiol 1987, 62: 61-70.PubMedGoogle Scholar
  23. Snyder JV, Powner DJ: Effects of mechanical ventilation on the measurement of cardiac output by thermodilution. Crit Care Med 1982, 10: 677-682.View ArticleGoogle Scholar
  24. Wesseling KH, Jansen JRC, Settels JJ, Schreuder JJ: Computation of aortic flow from pressure in humans using a nonlinear, three-element model. J Appl Physiol 1993, 74: 2566-2573.PubMedGoogle Scholar
  25. Singer M, Clarke J, Bennett D: Continuous hemodynamic monitoring by esophageal Doppler. Crit Care Med 1989, 17: 447-452.View ArticlePubMedGoogle Scholar
  26. Cariou A, Monchi M, Joly LM, Bellenfant F, Claessens YE, Thebert D, Brunet F, Dhainaut JF: Noninvasive cardiac output monitoring by aortic blood flow determination: evaluation of the Sometec Dynemo-3000 system. Crit Care Med 1998, 26: 2066-2072. 10.1097/00003246-199812000-00043View ArticlePubMedGoogle Scholar
  27. Valtier B, Cholley BP, Belot JP, de la Coussaye JE, Mateo J, Payen DM: Noninvasive monitoring of cardiac output in critically ill patients using transesophageal Doppler. Am J Respir Crit Care Med 1998, 158: 77-83.View ArticlePubMedGoogle Scholar
  28. Rady MY, Rivers EP, Martin GB, Smithline H, Appelton T, Nowak RM: Continuous central venous oximetry and shock index in the emergency department: use in the evaluation of clinical shock. Am J Emerg Med 1992, 10: 538-543. 10.1016/0735-6757(92)90178-ZView ArticlePubMedGoogle Scholar
  29. Scheinman MM, Brown MA, Rapaport E: Critical assessment of use of central venous oxygen saturation as a mirror of mixed venous oxygen in severely ill cardiac patients. Circulation 1969, 40: 165-172.View ArticlePubMedGoogle Scholar
  30. Creteur J, De Backer D, Vincent JL: A dobutamine test can disclose hepatosplanchnic hypoperfusion in septic patients. Am J Respir Crit Care Med 1999, 160: 839-845.View ArticlePubMedGoogle Scholar
  31. Tang W, Weil MH, Sun S, Noc M, Gazmuri RJ, Bisera J: Gastric intramural PCO 2 as a monitor of perfusion failure during hemorrhagic and anaphylactic shock. J Appl Physiol 1994, 76: 572-577.PubMedGoogle Scholar
  32. Nakagawa Y, Weil MH, Tang W, Sun S, Yamaguchi H, Jin X, Bisera J: Sublingual capnometry for diagnosis and quantitation of circulatory shock. Am J Respir Crit Care Med 1998, 157: 1838-1843.View ArticlePubMedGoogle Scholar
  33. Marik PE, Bankov A: Sublingual capnometry versus traditional markers of tissue oxygenation in critically ill patients. Crit Care Med 2003, 31: 818-822. 10.1097/01.CCM.0000054862.74829.EAView ArticlePubMedGoogle Scholar
  34. Dellinger RP, Carlet JM, Masur H, Gerlach H, Calandra T, Cohen J, Gea-Banacloche J, Keh D, Marshall JC, Parker MM, for the Surviving Sepsis Campaign Management Guidelines Committee, et al.: Crit Care Med. 2004, 32: 858-873. 10.1097/01.CCM.0000117317.18092.E4View ArticlePubMedGoogle Scholar
  35. Pinsky MR: Using ventilation-induced aortic pressure and flow variation to diagnose preload responsiveness. Intensive Care Med 2004, 30: 1008-1010. 10.1007/s00134-004-2208-6View ArticlePubMedGoogle Scholar
  36. Michard F, Teboul JL: Predicting fluid responsiveness in ICU patients. A critical analysis of the evidence. Chest 2002, 121: 2000-2008. 10.1378/chest.121.6.2000View ArticlePubMedGoogle Scholar
  37. Thomas M, Shillingford J: The circulatory response to a standard postural change in ischaemic heart disease. Br Heart J 1965, 27: 17-27.PubMed CentralView ArticlePubMedGoogle Scholar
  38. Boulain T, Achard JM, Teboul JL, Richard C, Perrotin D, Ginies G: Changes in blood pressure induced by passive leg raising predict response to fluid loading in critically ill patients. Chest 2002, 121: 1245-1252. 10.1378/chest.121.4.1245View ArticlePubMedGoogle Scholar
  39. Monnet X, Rienzo M, Osman D, Anguel N, Richard C, Pinsky MR, Teboul JL: Esophageal Doppler monitoring predicts fluid responsiveness in critically ill ventilated patients. Intensive Care Med 2005, 31: 1195-1201. 10.1007/s00134-005-2731-0View ArticlePubMedGoogle Scholar
  40. Pinsky MR: Determinants of pulmonary artery flow variation during respiration. J Appl Physiol 1984, 56: 1237-1245. 10.1063/1.334058View ArticlePubMedGoogle Scholar
  41. Magder SA, Georgiadis G, Tuck C: Respiratory variations in right atrial pressure predict response to fluid challenge. J Crit Care 1992, 7: 76-85. 10.1016/0883-9441(92)90032-3View ArticleGoogle Scholar
  42. Reuter DA, Bayerlein J, Goepfert MS, Weis FC, Kilger E, Lamm P, Goetz AE: Influence of tidal volume on left ventricular stroke volume variation measured by pulse contour analysis in mechanically ventilated patients. Intensive Care Med 2003, 29: 476-480.PubMedGoogle Scholar
  43. Perel A: Analogue values from invasive hemodynamic monitoring. In Applied Cardiovascular Physiology. Edited by: Pinsky MR. Berlin, Germany: Springer-Verlag; 1997:129-140.View ArticleGoogle Scholar
  44. Perel A: Assessing fluid responsiveness by the systolic pressure variation in mechanically ventilated patients. Systolic pressure variation as a guide to fluid therapy in patients with sepsis-induced hypotension. Anesthesiology 1998, 89: 1309-1310. 10.1097/00000542-199812000-00005View ArticlePubMedGoogle Scholar
  45. Michard F, Boussat S, Chemla D, Anguel N, Mercat A, Lecarpentier Y, Richard C, Pinsky MR, Teboul JL: Relation between respiratory changes in arterial pulse pressure and fluid responsiveness in septic patients with acute circulatory failure. Am J Respir Crit Care Med 2000, 162: 134-138.View ArticlePubMedGoogle Scholar
  46. Slama M, Masson H, Teboul JL, Arnout ML, Susic D, Frohlich E, Andrejak M: Respiratory variations of aortic VTI: a new index of hypovolemia and fluid responsiveness. Am J Physiol Heart Circ Physiol 2002, 283: H1729-H1733.View ArticlePubMedGoogle Scholar
  47. Feissel M, Michard F, Mangin I, Ruyer O, Faller JP, Teboul JL: Respiratory changes in aortic blood velocity as an indicator of fluid responsiveness in ventilated patients with septic shock. Chest 2001, 119: 867-873. 10.1378/chest.119.3.867View ArticlePubMedGoogle Scholar
  48. Stroke volume variation as a predictor of fluid responsiveness in patients undergoing brain surgery Anesth Analg 2001, 92: 984-989. 10.1097/00000539-200104000-00034
  49. Pinsky MR: Probing the limits of arterial pulse contour analysis to predict preload responsiveness. Anesth Analg 2003, 96: 1245-1247. 10.1213/01.ANE.0000055821.40075.38View ArticlePubMedGoogle Scholar
  50. Pinsky MR: Using ventilation-induced aortic pressure and flow variation to diagnose preload responsiveness. Intensive Care Med 2004, 30: 1008-1010. 10.1007/s00134-004-2208-6View ArticlePubMedGoogle Scholar
  51. McKee M, Clarke A: Guidelines, enthusiasms, uncertainty and the limits to purchasing. BMJ 1995, 310: 101-104.PubMed CentralView ArticlePubMedGoogle Scholar
  52. McKendry M, McGloin H, Saberi D, Caudwell L, Brady AR, Singer M: Randomised controlled trial assessing the impact of a nurse delivered, flow monitored protocol for optimisation of circulatory status after cardiac surgery. BMJ 2004, 329: 438-443. 10.1136/bmj.38156.767118.7CView ArticleGoogle Scholar
  53. Mythen MG, Webb AR: Perioperative plasma volume expansion reduces the incidence of gut mucosal hypoperfusion during cardiac surgery. Arch Surg 1995, 130: 423-429.View ArticlePubMedGoogle Scholar
  54. Venn R, Steele A, Richardson P, Poloniecki J, Grounds M, Newman P: Randomised controlled trial to investigate influence of the fluid challenge on duration of hospital stay and perioperative morbidity in patients with hip fractures. Br J Anaesth 2002, 88: 65-71. 10.1093/bja/88.1.65View ArticlePubMedGoogle Scholar
  55. Jastremski MS, Chelluri L, Beney KM: Analysis of the effects of continuous on-line monitoring of mixed venous oxygen saturation on patient outcome and cost-effectiveness. Crit Care Med 1989, 17: 148-153.View ArticlePubMedGoogle Scholar
  56. Fenwick E, Wilson J, Sculpher M, Claxton K: Pre-operative optimisation employing dopexamine or adrenaline for patients undergoing major elective surgery: a cost-effectiveness analysis. Intensive Care Med 2002, 28: 599-608. 10.1007/s00134-002-1257-yView ArticlePubMedGoogle Scholar
  57. Pinsky MR: Functional hemodynamic monitoring: applied physiology at the bedside. In Yearbook of Emergency and Intensive Care Medicine 2001. Edited by: Vincent JL. Berlin, Germany: Springer-Verlag; 2002:537-552.Google Scholar


© BioMed Central Ltd 2005