Comparison of a fully automatic ARDSNet protocol and a feedback-controlled open lung management concept

  • A Pomprapa1,
  • D Schwaiberger2,
  • B Lachmann2 and
  • S Leonhardt1
Critical Care201317(Suppl 2):P97

DOI: 10.1186/cc12035

Published: 19 March 2013

Introduction

The aim of this study is to compare two ventilation strategies, the ARDSNet protocol and open lung management, using computer control for 6 hours. The standard therapy for patients with ARDS does typically apply a mechanical ventilator to support breathing. The cost of therapy is high and it requires much attention from physicians to adjust the proper ventilation settings in a timely manner. A closed-loop ventilation concept has therefore been developed and tested with two induced ARDS pigs.

Methods

The hardware system is composed of a ventilator (Servo 300), a spectrophotometry (CEVOX), a capnography device (CO2SMO+), an electrical impedance tomography device (GOE MF II) and a patient monitor (Sirecust). The software is developed with Labview 7.1. With approval from the ethical committee, two 27 kg pigs were exposed to surfactant depletion with a warm saline washout to induce ARDS (PaO2/FiO2 <200 mmHg). One pig model was ventilated with an automatic ARDSNet protocol and another was automatically ventilated with open lung management. Blood gas analysis (BGA) was carried out every half an hour.

Results

Artificial ventilation using the auto ARDSNet protocol successfully stabilized oxygenation, minimized plateau pressure (<30 cmH2O), and controlled the pH value for acidosis and alkalosis management. On the other hand, auto open lung management offers a distinctive result of ventilation. A significant improvement of oxygenation and lung compliance was observed within a few breaths after the recruitment maneuvers. Both subjects were ventilated at the same tidal volume of 6 ml/kg and the comparative results of automatic ventilation settings and BGA are provided in Table 1 for every 2 hours.
Table 1

Comparative results between auto ARDSNet protocol and auto open lung management

 

Before auto ventilation

After 2 hours

After 4 hours

After 6 hours

 

Auto ARDSNet

Auto OLM

Auto ARDSNet

Auto OLM

Auto ARDSNet

Auto OLM

Auto ARDSNet

Auto OLM

FiO2

1.0

1.0

0.40

0.25

0.40

0.25

0.50

0.25

RR (bpm)

22

40

23

38

23

39

23

35

I:E ratio

1:2

1:1

1:2

1:1

1:2

1:1

1:2

1:1

VT (ml/kg)

6

6

6

6

6

6

6

6

Pplat/PIP (mmHg)

22

22

18

20.6

15.8

17.7

14.6

15.4

PEEP (cmH2O)

5

10

8

15

2

13

10

11

pH

7.34

7.3

7.32

7.52

7.37

7.6

7.38

7.51

pO2/FiO2 (mmHg)

59

59

180

380.8

155

387.2

122

404

pCO2 (mmHg)

52

64

56

39.6

58

31.5

54

41

OLM, open lung management.

Conclusion

The auto open lung management concept gave much better gas exchange than the auto ARDSNet protocol. These preliminary results showed a necessity to evaluate the two different ventilation strategies. Therefore, further experiments with pig models will be implemented in the near future to obtain results with statistical significance and to ensure the safety of automation in a mechanical ventilation system.

Authors’ Affiliations

(1)
RWTH Aachen University
(2)
Charite Berlin

Copyright

© Pomprapa et al.; licensee BioMed Central Ltd. 2013

This article is published under license to BioMed Central Ltd. This is an Open Access article distributed under the terms of the Creative Commons Attribution License (http://creativecommons.org/licenses/by/2.0), which permits unrestricted use, distribution, and reproduction in any medium, provided the original work is properly cited.