Open Access

Training and experience are more important than the type of practitioner for intubation success

  • Pieter F Fouche1Email author,
  • Paul M Middleton2 and
  • Kristina M Zverinova1
Critical Care201317:412

DOI: 10.1186/cc11924

Published: 8 February 2013

A meta-analysis by Lossius and colleagues demonstrated a higher endotracheal intubation (ETI) success rate for physicians compared with nonphysicians [1]. We re-analysed the data from Lossius and colleagues to investigate the relationship between the clinicians' levels of training and ETI success rates, utilising the 'Overview of included studies' spreadsheet published online as an additional file. We created a variable for a high level of ETI training taken from the 'extensively trained' description in the 'EMS manning' column. We assumed the physicians to also be highly trained in ETI, as they entirely comprised emergency physicians and anaesthesiologists. Our multivariate meta-regression analysis, adjusting for the training and the type of practitioners, found a 2% difference in intubation success (P = 0.12) between physicians and nonphysicians (Table 1), which was not statistically significant. Rapid sequence induction was left out of the final model, because all physicians utilised this induction and its inclusion would needlessly inflate variance.
Table 1

Predictors of intubation success, univariate versus multivariate regression model

Predictor

Unadjusted success difference

P value

Adjusted success difference

P value

Training of clinician (reference: lower experience)

    

Extensively trained

0.08

0.0001

0.07

0.001

EMS manning (reference: physician)

    

Nonphysician

-0.04

0.01

-0.02

0.12

RSI (reference: no RSI)

    

RSI

0.07

0.0005

-

-

EMS, emergency medical service; RSI, rapid sequence induction.

Lossius and colleagues compared ETI success rates of highly trained anaesthesiologists and emergency physicians from northwestern Europe with paramedics and nurses from the rest of the world [1]. One study incorporated in their meta-analysis included rural basic emergency medical technicians with no previous ETI experience [2]. Perhaps unsurprisingly, the authors showed increased ETI success rates for highly trained physicians compared with such a varied group of nonphysicians. However, our re-analysis suggests that the differing level of ETI training between clinicians, not the type of clinician, might be more important for successful ETI.

Authors' response

Hans Morten Lossius, David J Lockey and Jo Røislien

Thank you for your interest in our recently published paper [1]. Our article makes it quite clear that we consider training of key importance in intubation success rates. While training may be more important than the type of provider, the type of provider is often a marker of level of training. We point out that there is considerable variability in paramedic training, and to a lesser extent physician training, in different emergency medicine systems. However, well-trained anaesthesiologists are likely to have more training and experience in advanced airway management than even well-trained paramedics. We looked at one large study involving paramedics trained to a high level, higher than most emergency medicine systems are able to resource, and found that intubation failure rates were higher than for undifferentiated physician intubation. We believe our publication provides evidence that paramedics, even with good training but limited regular exposure to advanced airway management, have a higher intubation failure rate than physicians with high exposure to in-hospital and out-of-hospital advanced airway management.

Advanced airway management must be considered a complex intervention in which the ETI procedure is one of many factors influencing outcome [3, 4]. In-hospital advanced airway management is the responsibility of specialist physicians working in high-dependency units. Prehospital advanced airway management is even more challenging due to environmental factors and arbitrary patient information. Successful airway management is dependent on a comprehensive understanding of pathophysiological processes and high competence in advanced corrective interventions, and is not equivalent to good technical skills in ETI alone.

Abbreviations

ETI: 

endotracheal intubation.

Declarations

Authors’ Affiliations

(1)
Ambulance Service of New South Wales
(2)
Discipline of Emergency Medicine, University of Sydney

References

  1. Lossius HM, Røislien J, Lockey DJ: Patient safety in pre-hospital emergency tracheal intubation: a comprehensive meta-analysis of the intubation success rates of EMS providers. Crit Care 2012, 16: R24. 10.1186/cc11189PubMed CentralView ArticlePubMedGoogle Scholar
  2. Bradley JS, Billows GL, Olinger ML, Boha SP, Cordell WH, Nelson DR: Prehospital oral endotracheal intubation by rural basic emergency medical technicians. Ann Emerg Med 1998, 32: 26-32. 10.1016/S0196-0644(98)70095-2View ArticlePubMedGoogle Scholar
  3. Lossius HM, Sollid SJ, Rehn M, Lockey DJ: Revisiting the value of pre-hospital tracheal intubation: an all time systematic literature review extracting the Utstein airway core variables. Crit Care 2011, 15: R26. 10.1186/cc9973PubMed CentralView ArticlePubMedGoogle Scholar
  4. Sollid SJ, Lockey DJ, Lossius HM, Pre-hospital advanced airway management expert group: A consensus-based template for uniform reporting of data from pre-hospital advanced airway management. Scand J Trauma Resusc Emerg Med 2009, 17: 58. 10.1186/1757-7241-17-58PubMed CentralView ArticlePubMedGoogle Scholar

Copyright