Open Access

Prediction of severe community-acquired pneumonia: a systematic review and meta-analysis

  • Christophe Marti1Email author,
  • Nicolas Garin1, 2,
  • Olivier Grosgurin1,
  • Antoine Poncet3,
  • Christophe Combescure3,
  • Sebastian Carballo1 and
  • Arnaud Perrier1
Critical Care201216:R141

DOI: 10.1186/cc11447

Received: 16 May 2012

Accepted: 27 July 2012

Published: 27 July 2012

Abstract

Introduction

Severity assessment and site-of-care decisions for patients with community-acquired pneumonia (CAP) are pivotal for patients' safety and adequate allocation of resources. Late admission to the intensive care unit (ICU) has been associated with increased mortality in CAP. We aimed to review and meta-analyze systematically the performance of clinical prediction rules to identify CAP patients requiring ICU admission or intensive treatment.

Methods

We systematically searched Medline, Embase, and the Cochrane Controlled Trials registry for clinical trials evaluating the performance of prognostic rules to predict the need for ICU admission, intensive treatment, or the occurrence of early mortality in patients with CAP.

Results

Sufficient data were available to perform a meta-analysis on eight scores: PSI, CURB-65, CRB-65, CURB, ATS 2001, ATS/IDSA 2007, SCAP score, and SMART-COP. The estimated AUC of PSI and CURB-65 scores to predict ICU admission was 0.69. Among scores proposed for prediction of ICU admission, ATS-2001 and ATS/IDSA 2007 scores had better operative characteristics, with a sensitivity of 70% (CI, 61 to 77) and 84% (48 to 97) and a specificity of 90% (CI, 82 to 95) and 78% (46 to 93), but their clinical utility is limited by the use of major criteria.

ATS/IDSA 2007 minor criteria have good specificity (91% CI, 84 to 95) and moderate sensitivity (57% CI, 46 to 68). SMART-COP and SCAP score have good sensitivity (79% CI, 69 to 97, and 94% CI, 88 to 97) and moderate specificity (64% CI, 30 to 66, and 46% CI, 27 to 66). Major differences in populations, prognostic factor measurement, and outcome definition limit comparison. Our analysis also highlights a high degree of heterogeneity among the studies.

Conclusions

New severity scores for predicting the need for ICU or intensive treatment in patients with CAP, such as ATS/IDSA 2007 minor criteria, SCAP score, and SMART-COP, have better discriminative performances compared with PSI and CURB-65. High negative predictive value is the most consistent finding among the different prediction rules. These rules should be considered an aid to clinical judgment to guide ICU admission in CAP patients.

Introduction

Community-acquired pneumonia (CAP) is a major health problem. In the United States, 500,000 adults are hospitalized annually for CAP [1], of whom 10% to 20% are admitted to the intensive care unit (ICU) [2]. Because site of care is a major determinant of costs, clinical prediction rules have been developed to identify patients with low mortality who can be safely treated as outpatients [3].

Since 1993, efforts have been made to identify severe community-acquired pneumonia (SCAP) requiring admission to the ICU because the ICU is an expensive and scarce resource. Concurrently, delay in ICU admission of CAP patients has been shown to be associated with increased mortality [46]. In 1993, the American Thoracic Society (ATS) proposed a definition of severe CAP requiring ICU admission [7]. Since then, these criteria have been updated twice [8, 9], and various clinical prediction rules have been developed to predict SCAP [1012]. In addition to the variety of the prediction rules, the assessment of their validity is further hampered by the absence of a unique definition of SCAP and the inclusion of ICU admission or intensive treatment in the definition, which exposes prediction rules for SCAP to incorporation bias [13].

The goal of our search was to review systematically the performance of existing clinical prediction rules to identify in the Emergency Department patients with CAP requiring ICU admission or intensive treatment.

Materials and methods

Search strategy and study selection

We systematically searched Medline, Embase, and the Cochrane Controlled Trials registry by using the following key words: community-acquired pneumonia AND (decision tree OR clinical prediction rule OR clinical prediction score OR clinical decision rule OR clinical decision score OR management studies OR outcome studies OR ICU admission OR ICU need OR invasive management OR severity assessment). The search was performed for articles in English, French, Italian, Spanish, and German languages and limited to articles with an abstract and completed on the first of March 2012. To ensure a comprehensive literature search, we examined reference lists from retrieved articles and reference literature (guidelines and systematic reviews) and questioned experts in CAP for possible missing studies.

Study inclusion and data extraction

Eligible studies were prospective or retrospective studies evaluating clinical prediction rules in adult immunocompetent patients with CAP to predict the need for ICU admission, intensive treatment, or early mortality (< 14 days). The evaluation had to be performed during the first 24 hours after hospital admission. Studies addressing specific patient subgroups based on etiology or age were excluded. A prediction rule was defined as the combination of two or more clinical or biologic markers. Four investigators (CM, NG, SC, and OG) evaluated studies for possible inclusion. All studies were evaluated independently by at least two investigators. Nonrelevant studies were excluded based on title and abstract. For potentially relevant studies, the full text was obtained, and two investigators (CM, NG) independently assessed study eligibility and extracted the data on study design, patient characteristics, and outcomes. Disagreements were resolved by consensus or by discussion with a third reviewer (AP)

Quality assessment

We used modified quality criteria based on the guidelines for assessing quality in prognostic studies [14]. Two investigators (CM, NG) assessed study quality independently. Each of six items was scored from 0 to 2. Studies with a total quality score between 11 and 12 were considered "good," between 9 to 10, "moderate," and 8 or less was considered "poor."

Data analysis

For each score, the diagnostic performances (sensitivity, specificity, likelihood ratios, and diagnostic odds ratios) to predict different definitions of SCAP (ICU admission, early death, or intensive treatment) at the usual cut-off were pooled by using the method of the inverse of the variance. Random effects were systematically introduced [15]. Heterogeneity was measured by the I-square index [16] and tested with the Cochran test. Potential heterogeneity factors were explored by subgroup analyses for the Pneumonia Severity Index (PSI) (only score with more than 10 studies). A sensitivity analysis was conducted to check the robustness of the pooled sensitivities and specificities by removing each study, one by one. The R package "meta: Meta-analysis with R, version 1.6-1" was used for these analyses. For the PSI and CURB-65 (Confusion, Urea, Respiratory Rate, Blood pressure, Age > 65 years) scores, a summary ROC curve was assessed by the approach proposed by Moses et al.[17]. As several sensitivities and specificities were reported in the studies at different cut-offs, we used a linear mixed model with a correlation structure to take the dependence of the measures into account. The 95% confidence intervals of the areas under the curves were obtained by bootstrap. This analysis was performed with S-plus 8.0 for Windows. The significance level was 0.05 for all analyses. Forest plots of the Sensibility and Specificity were used for the graphic display of the results.

Results

The search retrieved a total of 5,249 references, among which 1,005 duplicates were identified. Of the 4,244 remaining articles, 3,966 were excluded based on title and abstract (Figure 1). Full texts were obtained for the remaining 278 articles. Ten did not contain original data, 13 concerned only ICU patients, 204 did not meet inclusion criteria, 18 were review articles, and 33 satisfied inclusion criteria. Three articles [1820] were identified by manual search of the references, leading to a total of 36 included articles [5, 1012, 1850]. Main characteristics of included studies are detailed in Table 1.
Figure 1

Study flow chart.

Table 1

Main characteristics of included studies

Study first author

Year of publication

Studied scores

End point

Prevalence end point

Uni- versus multicentric

Prospective/Retrospective

Ambulatory/Hospitalized

Number

NTBR

Major criteria excluded

Ananda-Rajah

2008

PSI/CURB-65

ICU admission

10.50%

Uni

R

H

408

Included

n

Angus

2002

PSI/ATS 2001/ATS 1993/URB

ICU admission

12.70%

Multi

P

H

1,339

Included

n

Brown

2009

IDSA-ATS 2007/SMART-COP/CURB-65/SCAP

IT in ICU/IT/ICU admission

12.35%

Uni

R

A+H

2,413

Excluded

n

Buising

2006

PSI/ATS 2001/CURB/CURB-65

ICU admission/death or ICU

6.60%

Uni

P

A+H

392

Included

n

Buising

2007

PSI/CURB/CURB-65/CORB

Vent/vaso ICU

6.30%

Uni

P

A+H

330

Included

n

Calbo

2004

PSI

ICU admission

3.30%

Uni

P

H

362

Included

n

Capelastegui

2006

CURB-65

ICU admission

4.09%

Uni

R

H

1,100

Included

n

Chalmers

2008

CRB-65/CURB-65

Vent/vaso

10.20%

Multi

P

A+H

1,007

Excluded

n

Chalmers

2011

IDSA-ATS 2007/SMART-COP/CURB-65/SCAP/CURB/CRB-65/PSI, ATS 2001 minor

Vent/vaso

ICU admission

6.4%

7.6%

Uni

P

H

1,062

Excluded

y

Charles

2008

SMART-COP/CURB-65/PSI

Vent/vaso

10.30%

Multi

P

A+H

882

Included

n

Davis

2010

SMART-COP/SMARTACOP

Vent/vaso

21.00%

Uni

P

H

184

Excluded

n

Escobar

2008

Abbreviated PSI

ICU admission

12.30%

Uni

R

H

6,147

Included

n

Espana

2006

ATS 2001/PSI/CURB-65/SCAP

Death, vent, or septic shock

7.10%

Uni

P

A+H

1,057

Included

n

   

Death, vent or septic shock

5.90%

Uni

P

A+H

719

Included

n

   

Death, vent or septic shock

11.9%

Multi

R

H

1,121

Included

n

Espana

2010

SCAP/PSI/CURB-65

ICU admission

NA

Multi

R

H

712

Included

n

Ewig

2000

PSI

ICU admission

9%

Uni

P

H

232

Included

n

Ewig

2004

ATS 2001/PSI/CURB/CRB/URB/

ICU admission

16.70%

Uni

p

H

696

Excluded

n

Ewig

1998

ATS 1993/ATS 2001

ICU admission

16%

Uni

p

H

332

Excluded

n

Feagan

2000

PSI

ICU admission

13.60%

Multi

r

H

858

Included

n

Frei

2004

SBP/pH/O2 saturation/pulse

ICU admission

17%

Uni

r

H

782

Included

n

Fukuyama

2011

IDSA-ATS 2007/SMART-COP/CURB-65/SCAP/PSI/A-DROP

ICU admission

Death, vent, or septic shock

7.6%

11.9%

Uni

p

H

505

Excluded

n

Garcia-Vidal

2008

PSI

Early death

2.3%

Uni

p

H

2,457

Included

 

Garau

2008

PSI

ICU admission

5%

Multi

r

H

3,233

Included

n

Kamath

2003

CURB

ICU admission

10%

Uni

p

H

100

Excluded

n

Lamy

2004

PSI

ICU admission

14%

Uni

r

H

152

Included

n

Liapikou

2009

IDSA-ATS 2007/ATS 2001/PSI

ICU admission

11%

Uni

p

H

2,102

Excluded

n(y)

Man

2007

PSI/CURB-65/CRB-65/ATS 2001

ICU admission

4%

Uni

p

H

1,016

Included

n

Marrie

2007

PSI/CURB-65

ICU admission

10%

Multi

p

H

3,675

Included

n

Neill

1996

CURB

Mortality and ICU

11%

Uni

p

H

251

Excluded

n

Phua

2009

PSI/CURB-65/IDSA-ATS 2007

ICU admission

15%

Uni

p

H

1,017

Excluded

y

Putinati

2003

PSI

ICU admission

10%

Uni

p

H

229

Included

n

Renaud

2007

PSI

ICU admission

3.70%

Multi

p

H

566

Included

n

  

PSI

ICU admission

10.50%

Multi

p

H

761

Included

n

Renaud

2009

REA-ICU

Early ICU admission < 3

4.40%

Multi

p

A+H

4,593

Excluded

y

Restrepo

2008

PSI/ATS 2001

ICU admission

19.90%

Multi

r

H

730

Excluded

n

Riley

2004

ATS 2001/PSI

ICU admission

23.69%

Uni

r

H

498

Excluded

n

Roson

2001

PSI

ICU admission

8%

Uni

p

H

533

Included

n

Shah

2010

CURB-65/PSI

ICU admission

23.30%

Uni

p

A+H

150

Included

n

Van der Eerden

2004

PSI

ICU admission

8.00%

Uni

p

H

260

Included

n

A, ambulatory; H, hospitalized; ICU, intensive care unit; IT, intensive treatment; P, prospective; R, retrospective; n, number; y, yes; NTBR, not to be resuscitated; Vent, ventilation; Vaso, vasopressor.

Scores

We identified 11 main severity scores based on 20 variables. Components of the main severity scores are illustrated in Figure 2. Sufficient data were available to perform a meta-analysis on eight scores: PSI, CURB-65, CRB-65 (Confusion, Respiratory Rate, Blood pressure, Age > 65), CURB (Confusion, Urea, Respiratory rate, Blood pressure), ATS 2001 criteria, ATS/Infectious Disease Society of America (IDSA) 2007 criteria, SCAP score (Severe Community-Acquired Pneumonia), and SMART-COP (Systolic Blood pressure, Multilobar infiltrate, Albumin, Respiratory Rate, Tachycardia, Confusion, low Oxygen, low PH) (Table 2) Score definitions are included in the Additional file 1. Forrest plots for specificity and sensitivity of the eight scores for the prediction of ICU admission are provided in Additional file 2.
Figure 2

Components of the main severity scores. Criteria used in the score appear as shaded areas. BP, blood pressure; HR, heart rate; RR, respiratory rate; T, temperature.

Table 2

Operative characteristics of the principal scores to predict ICU admission at their usual cut-off (95% CI)

 

Sensitivity

Specificity

NLR

PLR

DOR

PSI ≥ 4

75.0 (71-78)

48.0 (44-52)

0.53 (0.46-0.6)

1.5 (1.4-1.6)

2.9 (2.4-3)

CURB-65 ≥ 3

56.2 (41-70)

74.2 (68-79)

0.64 (0.51-0.79)

2.1 (1.6-2.7)

3.6 (2.2-5.8)

CRB-65 ≥ 3

34.2 (18-55)

90.6 (89-92)

0.72 (0.54-0.97)

3.6 (1.9-6.9)

5.0 (2.0-12.7)

CURB ≥ 2

76.8 (48-92)

68.6 (53-81)

0.35 (0.18-0.70)

2.3 (1.9-2.7)

5.5 (3.7-8.2)

ATS 2001

69.5 (61-77)

90.1 (82-95)

0.37 (0.30-0.46)

7.3 (4.4-12.2)

24.6 (13.1-46.4)

ATS 2007

83.8 (48-97)

77.7 (46-93)

0.22 (0.08-0.66)

3.8 (1.7-8.6)

17.6 (13.1-24.1)

ATS 2007a

57.0 (46-68)

90.5 (84-95)

0.48 (0.38-0.6)

5.9 (3.8-9.3)

13.1 (7.7-22.3)

SCAP

93.8 (88-97)

45.6 (27-66)

0.13 (0.06-0.26)

1.8 (1.2-2.6)

14.9 (6.7-33.1)

SMART-COP

79.0 (69-87)

64.2 (30-66)

0.15 (0.03-0.91)

2.6 (1.3-5.3)

14.9 (8.6-25.7)

DOR, diagnostic odds ratio; NLR, negative likelihood ratio; PLR, positive likelihood ratio. aMinor criteria.

PSI

Twenty-four studies [5, 18, 19, 21, 22, 2426, 32, 34, 36, 3952], including 20,622 patients and 2,073 ICU admissions (10.1%), evaluated the performance of PSI to predict ICU need. A PSI score category of IV or more had a pooled sensitivity of 75% and a specificity of 48%. A cut-off of V increased specificity to 84% and decreased sensitivity to 38%. The global performance of PSI to predict ICU admission was modest, with an AUC of 0.69 (Figure 3). Significant heterogeneity was present. Performance of PSI to predict an alternative definition of SCAP, including mortality, was superior, with a pooled sensitivity of 92.4% (CI, 89 to 95) and specificity of 56.2% (CI, 43 to 69) in four cohorts including 3,195 patients [11, 49].
Figure 3

SROC curve and area under the curve (AUC) of Pneumonia Severity Index (PSI) and CURB-65 to predict ICU admission. Individual studies are represented by a number indicating the cut-off used. Their place on the diagram represents the sensitivity and specificity of the individual study. Diamonds represent meta-analytic test statistics for each cut-off.

CURB-65

CURB-65 was studied in nine cohorts [21, 24, 25, 27, 40, 42, 47, 49, 50] including a total of 5,773 patients and 479 ICU admissions (8.3%). At the usual cut-off value of 3 or more, pooled sensitivity was 56%, and specificity, 74%. The global performance of CURB-65 to predict ICU admission was similar to PSI with an AUC of 0.69 (Figure 3). Significant heterogeneity was present. The performance of CURB-65 to predict the need for ventilation or vasopressors was studied in three publications [10, 28, 50] including 2,951 patients, 264 requiring intensive treatment. Results were similar, with a pooled sensitivity of 57.2% (CI, 37 to 75) and specificity of 77.2% (CI, 73 to 81).

CRB-65

CRB-65 is a simplified version of the CURB-65 including only clinical predictors. Two studies [40, 50] including 2,078 patients and 122 ICU patients (5.8%) calculated the performance of CRB-65 to predict ICU admission. For a threshold of 3 or more, pooled sensitivity was 34%, and specificity, 91%.

CURB (original BTS rule)

Performance of CURB to predict ICU admission was studied in four cohorts [24, 25, 32, 38] totaling 1,418 patients and 161 ICU admissions (12.1%). Pooled sensitivity of a CURB score of 2 or more to predict ICU admission was 76.8%, and specificity, 68.6%. Significant heterogeneity was observed.

ATS 2001

The original ATS criteria for severe CAP published in 1993 [7] included 10 criteria. Some of these criteria were assessed at admission, and others, at any time during clinical course, limiting their use as a prediction rule. A new set of criteria was proposed by Ewig in 1998 [33] and adopted by the ATS in 2001. This prediction rule consists of two major (mechanical ventilation or shock) and three minor criteria (blood pressure < 90 mm Hg at admission, PaO2/FiO2 < 250 mm Hg, and multilobar involvement on chest radiograph). The prediction rule is considered positive in the presence of one major or two minor criteria.

We identified eight studies [5, 22, 24, 32, 33, 39, 40, 45, 49] including a total number of 7,116 patients with 908 ICU admissions (12.8%). The pooled sensitivity was 69.5%, and specificity, 90.1%. Pooled AUC could not be calculated because of insufficient data. Performance of the 2001 ATS criteria in comparison with PSI and CURB-65 is illustrated in Figure 4. A supplementary study [11] validated this rule to predict a composite definition of SCAP (in-hospital death, mechanical ventilation, or shock) in three cohorts including 2,897 patients and 252 SCAP (8.7%). Pooled sensitivity of the ATS 2001 criteria was 52.7%, and specificity, 95.1%. One study validated the use of the ATS 2001 minor criteria on a cohort excluding patients with therapeutic limitations or major criteria [50]. Sensitivity and specificity of two or more minor criteria to predict ICU admission were 47% and 91%.
Figure 4

Pooled discriminative performance of the principal scores for severe CAP compared with Pneumonia Severity Index (PSI) and CURB-65 ROC curve.

ATS-IDSA 2007

A new prediction rule was proposed by the ATS and the Infectious Disease Society of America (IDSA) in 2007. This rule, known as the ATS-IDSA rule, consists of two major (mechanical ventilation or shock) and nine minor criteria (Figure 2). The rule is considered positive in the presence of one major or three minor criteria. We found five publications evaluating this rule [23, 39, 42, 49, 50]. Two studies, including 2,400 patients and 266 ICU patients (11%), validated the original rule to predict ICU admission [39, 49]. Pooled sensitivity was 84%, and specificity, 78%.

Four studies [23, 39, 42, 50] evaluated the performance of the minor criteria in a total of 6,412 patients including 650 ICU patients (10.1%). Pooled sensitivity was 57%, and specificity, 90%. Significant heterogeneity was present. Performance of the 2007 ATS-IDSA rule in comparison with PSI and CURB-65 is illustrated in Figure 4.

SMART-COP

Charles et al.[10] developed a prediction rule based on eight weighted criteria (see Additional File 1). This rule was validated in five external cohorts of pneumonia patients and further adapted and validated in two cohorts [29, 50]. Pooled sensitivity of SMART-COP to predict the need for vasopressors or mechanical ventilation was 79%, and specificity, 68%.

Two studies evaluated this rule to predict ICU admission [49, 50], with a pooled sensitivity of 79% and specificity of 64% on 1,567 patients including 112 ICU admissions (7.1%).

SCAP score

Espana et al.[11] derived and validated a prediction rule based on eight weighted criteria (see Additional File 1). Pooled performance of this rule on three cohorts totaling 3,402 patients (SCAP, 9%) to predict a composite definition of SCAP (in-hospital death, mechanical ventilation, or shock) was 92% (0.83 to 0.97) for sensitivity and 64% (0.5 to 0.76) for specificity.

Pooled performance of the SCAP score to predict ICU admission in two recent cohorts [49, 50] was similar in terms of sensitivity (94%) but lower regarding specificity (46%)

Other scores

Renaud et al.[12] proposed a prediction rule (REA-ICU index) based on 11 predictors (see Additional File 1). This rule was derived to predict early ICU admission (day 1 to day 3), excluding patients with an obvious ICU indication at admission and patients with therapeutic limitations (not to be resuscitated, NTBR order) and validated on four North American and European prospective multicenter cohorts including 6,560 patients. AUC was 0.81 (CI, 0.78 to 0.83) on the overall population.

An abbreviated version of the PSI was tested on an administrative database [30], and some authors proposed to modify the CURB rule to include an oxygenation parameter [25] (CORB). These scores have not been validated in other cohorts.

Quality assessment and sources of heterogeneity

We completed quality assessment according to the criteria developed by Hayden et al.[14].

Nine studies were considered good (scores ≥ 11); 23, moderate (9 to 10); and four poor (≤ 8) quality (see Additional file 3). Important differences were observed in terms of design, populations, and ICU admission rates. Twenty-nine studies included only hospitalized patients, and seven included both hospitalized and ambulatory patients. Patients with therapeutic limitations (NTBR order) were excluded in 15 of 36 studies, and patients with an obvious indication for ICU, in three [12, 42, 47]. ICU admission rate varied from 3.3% [26] to 23.9% [45], and receipt of intensive treatment in the ICU was highly variable (16% to 100% [25, 33]). When multiple measures of the same variable were available, as for vital signs, the measure introduced into the prediction rule was the first available one in four studies, the most abnormal one in five, and not specified in the majority of studies.

We performed an analysis to identify sources of heterogeneity for the only score (PSI) evaluated in at least 10 studies. Analyses did not reveal any significant factor of heterogeneity for sensitivity. Three factors influenced specificity: inclusion of NTBR patients decreased specificity; studies of good quality or high prevalence (≥ 15%) showed higher specificity. Nevertheless, the NTBR factor is highly correlated with the quality of the study (the NTBR patients were more frequently excluded in high-quality studies) and the prevalence (NTBR patients were more frequently excluded in studies with a high prevalence). NTBR exclusion is suspected to be the main factor explaining the heterogeneity observed in the specificity. The association between specificity and prevalence or quality may be caused by the correlation with NTBR exclusion.

Sensitivity analysis showed that the conclusions of the meta-analyses were robust and not caused by a single study. Sensitivity and specificity did not vary by more than 10%, except for CURB score: the study by Ewig [32] has a lower sensitivity and higher specificity than do other studies, but only four studies were included in this analysis.

Discussion

In this systematic review of clinical prediction rules to predict severe CAP, PSI and CURB-65 have similar performances to identify patients requiring ICU admission. A PSI score of 4 or more is more sensitive (75% versus 56%) but less specific (47% versus 74%) than a CURB-65 score of 3 or more. These two scores, derived and validated to predict 30-day mortality, perform poorly to predict ICU admission, with an estimated AUC of 0.69. This performance is inferior to their original application: AUC for prediction of 30-day mortality was 0.80 in a recent systematic review [53].

Some explanations are available for this difference. These prediction rules, and especially the PSI, are heavily weighted by comorbidities and age and tend to consider as severe, patients in whom CAP is a terminal event. In many cohorts, the mortality rate is higher than the ICU admission rate, suggesting that an important proportion of patients with CAP die without being considered for ICU admission [21, 39]. Age, a major component of these scores, is often negatively associated with ICU admission or intensive treatment [12, 25, 41] Exclusion of patients with therapeutic limitations may improve the specificity of these scores for ICU admission. In our heterogeneity analysis, PSI specificity increased from 45% to 56% when studies including NTBR patients were excluded (P = 0.008).

A new generation of scores, specifically developed to predict ICU admission, focuses on the severity of the pneumonia itself rather than on age and comorbid conditions. Overall, the performance of these scores seems superior to that of the PSI or CURB-65, particularly regarding specificity (90.1% and 90.5% for the ATS 2001 score and ATS/IDSA 2007 minor criteria versus 48.0% and 74.2% for PSI and CURB-65).

When considering scores derived over time, secular trends must be considered. Use of noninvasive ventilation (NIV) for severe CAP has increased over the last decade. This might have changed the proportion of patients receiving invasive mechanical ventilation. However, NIV was explicitly included in the definition of mechanical ventilation in most studies using that end point [10, 25, 29]. As the efficacy of NIV in alleviating respiratory failure for CAP patients is still debated, NIV is unlikely to have induced a major bias in the performance of the prediction rules over time. Furthermore, the principal scores (SCAP rule, SMART-COP, ATS 2007, and REA-ICU) have been proposed recently, at a time during which NIV for respiratory failure was routinely used.

Inclusion of major criteria in the ATS rules (mechanical ventilation and shock) improves their diagnostic performance but is not useful clinically, because these patients have an obvious indication for ICU admission. However, use of the minor criteria only does not seem to reduce the specificity of the ATS/IDSA 2007 score, as suggested by our results, with a pooled specificity of 90.5%. In a recent prospective cohort excluding patients with major criteria or therapeutic limitations, ATS-IDSA minor criteria had an AUC of 0.85 to predict ICU admission [50].

SMART-COP, REA-ICU, and SCAP scores seem to have operative characteristics similar to the ATS minor criteria but are less extensively validated. Also, confidence intervals are wide, and the use of different definitions of severe CAP limits comparison. Not surprisingly, these prediction rules share an important proportion of their predictive variables, as illustrated in Figure 2. Use of these three rules is more difficult than that of the ATS minor criteria, because of the weighting of the different predictive factors and the use of age-adjusted cut-offs.

Although the new generation of scores seems to have enhanced operative characteristics to predict ICU admission, their clinical utility is still debated. With a 10% prevalence of severe CAP and 57% sensitivity, even a specificity of 91% (as reached by the ATS-IDSA 2007 minor criteria) will lead to a positive predictive value of only 41%, leading to an important overuse of ICU resources. High negative predictive value is the most consistent finding among the different studies, suggesting that these scores could be more relevant to exclude the presence of a severe CAP than to aid in performing triage in patients for ICU admission. However, this high negative predictive value is mainly due to the low prevalence of severe patients. With a pooled sensitivity of 57%, the ATS-IDSA 2007 minor criteria would fail to identify almost one half of the patients with severe CAP, an unacceptably high proportion.

CAP is a complex and evolving inflammatory disease and critical clinical deterioration can result from various processes: respiratory failure, circulatory failure, destabilization of a preexisting comorbidity, appropriateness of initial antibiotic therapy, or hospital-acquired illnesses. It is not surprising that no single clinical rule has sufficient operating characteristics to be useful in this wide spectrum of evolution profiles.

The new generation of clinical prediction rules focuses on the early detection of respiratory and circulatory failure. Inclusion of various biomarkers such as procalcitonin [54], endothelin-1 [55], co-peptin [56], pro-atrial natriuretic peptide [57, 58], or adrenomedullin [59] is hoped to improve this detection. Nevertheless, these biomarkers will probably fail to predict clinical deterioration due to hospital-acquired complications or decompensated comorbidities. It is even dubious whether they will detect circulatory or respiratory failure in patients admitted in the very early course of their disease.

In our view, rather than a definitive response to severity assessment, clinical prediction rules in patients with CAP should be considered an aid to clinical judgment particularly useful for less-experienced clinicians. Some of the included variables are consistently associated with a grim evolution, and their identification in an individual patient should alert for this possibility and trigger its timely reassessment and a thorough evaluation for intensive care or high-dependency-unit admission.

Strengths

Our review incorporates the most recent published studies and updates the systematic review by Chalmers et al.[60]. We also used somewhat different inclusion criteria and data analysis. We did not include subgroups of patients based on age or pathogen, leading to the noninclusion of three studies [6163]. Among the 40 studies included in the two systematic reviews, 24 were included in both, 12 in our work only, and four in the study by Chalmers et al. only. We decided against aggregating different definitions of SCAP, in an attempt to limit heterogeneity, and computed the performances of the prediction rules for the different definitions of SCAP. We included large recent studies [49, 50] and could estimate the pooled performance of the more-recent prediction rules, SMART-COP and SCAP score. This allows direct comparison between recent prediction rules. Although included studies differed partially between our systematic review and that by Chalmers et al., our final results are very similar, mutually strengthening their validity.

Limitations

An important limitation of systematic reviews is their dependence on the quality of the included studies. Although the majority of included studies were considered of good or moderate quality, several pitfalls remain in the prediction of severe CAP. First, as already discussed, two biases are highly prevalent in these studies: inclusion in the studied population of patients not at risk for ICU admission (patients with therapeutic limitations); and use as a predictor of a surrogate of the outcome (use of mechanical ventilation and vasopressors, which are universally delivered only in an intensive or intermediate care unit).

Second, no universally accepted definition exists of severe CAP. The most frequently used proxy, ICU admission, is heavily influenced by ICU beds availability, local ICU admission policy, or subjectivity of the ICU specialist's evaluation. Use of a subjective decision such as ICU admission as a gold standard might lead to circular reasoning, because a perfect rule would be the one fitting usual practice. However, alternative definitions of SCAP, such as receipt of intensive treatment, do not seem to modify importantly operative characteristics of the prediction rules [23].

Third, definition of a "false positive" ICU admission is unclear: some patients might benefit from ICU admission even if they are not receiving vasopressors or mechanical ventilation (for example, through better fluid resuscitation).

Fourth, some of the studied rules have been fully incorporated in expert society recommendations. This might lead to contamination of ICU admission practices, further leading to an overestimation of their accuracy.

Finally, major heterogeneity was present among included studies, limiting the validity of the meta-analysis.

Conclusions

PSI and CURB-65 do not have sufficient operating characteristics to be useful for making ICU triage decisions in severe CAP. Newer rules, specifically conceived to aid in identifying severe CAP, perform better but still have insufficient test characteristics to be a major help in everyday decisions. Recent clinical prediction rules should be considered an aid to clinical judgment to guide ICU admission in CAP patients. Clinical trials evaluating this issue should exclude patients who are not candidates for ICU admission and predicting factors that make ICU admission mandatory. Inclusion of new biomarkers, dynamic reassessment of the severity scores, and impact studies evaluating their use would deserve evaluation in future clinical research.

Key messages

  • Identification of severe community-acquired pneumonia (SCAP) should allow admission of patients at the appropriate level of care.

  • Traditional severity scores, the PSI and CURB 65, perform poorly to identify patients requiring ICU admission.

  • New dedicated scores have better operative characteristics and could be useful adjuncts to clinical judgment.

Abbreviations

ATS: 

American Thoracic Society

AUC: 

area under the curve

CAP: 

community-acquired pneumonia

CURB: 

confusion, urea, respiratory rate, blood pressure

ICU: 

intensive care unit

IDSA: 

Infectious Disease Society of America

NIV: 

noninvasive ventilation

NTBR: 

not-to-be resuscitated

PSI: 

pneumonia severity index

ROC: 

receiver operating characteristic

SCAP: 

severe community-acquired pneumonia

SMART-COP: 

systolic blood pressure, multilobar infiltrate, albumin, respiratory rate, tachycardia, confusion, low oxygen, low pH.

Declarations

Authors’ Affiliations

(1)
Department of Internal Medicine, Rehabilitation and Geriatrics, University Hospitals of Geneva and Geneva
(2)
Division of Internal Medicine, Regional Hospital of Chablais
(3)
Department of Health and Community Medicine, University Hospitals of Geneva and Geneva

References

  1. Marston BJ, Plouffe JF, File TM Jr, Hackman BA, Salstrom SJ, Lipman HB, Kolczak MS, Breiman RF: Incidence of community-acquired pneumonia requiring hospitalization: results of a population-based active surveillance study in Ohio: The Community-Based Pneumonia Incidence Study Group. Arch Intern Med 1997, 157: 1709-1718. 10.1001/archinte.1997.00440360129015PubMedView ArticleGoogle Scholar
  2. Ewig S, Torres A: Severe community-acquired pneumonia. Clin Chest Med 1999, 20: 575-587. 10.1016/S0272-5231(05)70237-9PubMedView ArticleGoogle Scholar
  3. Fine MJ, Auble TE, Yealy DM, Hanusa BH, Weissfeld LA, Singer DE, Coley CM, Marrie TJ, Kapoor WN: A prediction rule to identify low-risk patients with community-acquired pneumonia. N Engl J Med 1997, 336: 243-250. 10.1056/NEJM199701233360402PubMedView ArticleGoogle Scholar
  4. Chalfin DB, Trzeciak S, Likourezos A, Baumann BM, Dellinger RP: Impact of delayed transfer of critically ill patients from the emergency department to the intensive care unit. Crit Care Med 2007, 35: 1477-1483. 10.1097/01.CCM.0000266585.74905.5APubMedView ArticleGoogle Scholar
  5. Restrepo MI, Mortensen EM, Velez JA, Frei C, Anzueto A: A comparative study of community-acquired pneumonia patients admitted to the ward and the ICU. Chest 2008, 133: 610-617. 10.1378/chest.07-1456PubMedView ArticleGoogle Scholar
  6. Renaud B, Santin A, Coma E, Camus N, Van Pelt D, Hayon J, Gurgui M, Roupie E, Herve J, Fine MJ, Brun-Buisson C, Labarere J: Association between timing of intensive care unit admission and outcomes for emergency department patients with community-acquired pneumonia. Crit Care Med 2009, 37: 2867-2874. 10.1097/CCM.0b013e3181b02dbbPubMedView ArticleGoogle Scholar
  7. Niederman MS, Bass JB Jr, Campbell GD, Fein AM, Grossman RF, Mandell LA, Marrie TJ, Sarosi GA, Torres A, Yu VL: Guidelines for the initial management of adults with community-acquired pneumonia: diagnosis, assessment of severity, and initial antimicrobial therapy: American Thoracic Society, Medical Section of the American Lung Association. Am Rev Respir Dis 1993, 148: 1418-1426.PubMedView ArticleGoogle Scholar
  8. Niederman MS, Mandell LA, Anzueto A, Bass JB, Broughton WA, Campbell GD, Dean N, File T, Fine MJ, Gross PA, Martinez F, Marrie TJ, Plouffe JF, Ramirez J, Sarosi GA, Torres A, Wilson R, Yu VL: Guidelines for the management of adults with community-acquired pneumonia: diagnosis, assessment of severity, antimicrobial therapy, and prevention. Am J Respir Crit Care Med 2001, 163: 1730-1754.PubMedView ArticleGoogle Scholar
  9. Mandell LA, Wunderink RG, Anzueto A, Bartlett JG, Campbell GD, Dean NC, Dowell SF, File TM Jr, Musher DM, Niederman MS, Torres A, Whitney CG: Infectious Diseases Society of America/American Thoracic Society consensus guidelines on the management of community-acquired pneumonia in adults. Clin Infect Dis 2007,44(Suppl 2):S27-S72.PubMedView ArticleGoogle Scholar
  10. Charles PG, Wolfe R, Whitby M, Fine MJ, Fuller AJ, Stirling R, Wright AA, Ramirez JA, Christiansen KJ, Waterer GW, Pierce RJ, Armstrong JG, Korman TM, Holmes P, Obrosky DS, Peyrani P, Johnson B, Hooy M, Grayson ML: SMART-COP: a tool for predicting the need for intensive respiratory or vasopressor support in community-acquired pneumonia. Clin Infect Dis 2008, 47: 375-384. 10.1086/589754PubMedView ArticleGoogle Scholar
  11. Espana PP, Capelastegui A, Gorordo I, Esteban C, Oribe M, Ortega M, Bilbao A, Quintana JM: Development and validation of a clinical prediction rule for severe community-acquired pneumonia. Am J Respir Crit Care Med 2006, 174: 1249-1256. 10.1164/rccm.200602-177OCPubMedView ArticleGoogle Scholar
  12. Renaud B, Labarere J, Coma E, Santin A, Hayon J, Gurgui M, Camus N, Roupie E, Hemery F, Herve J, Salloum M, Fine MJ, Brun-Buisson C: Risk stratification of early admission to the intensive care unit of patients with no major criteria of severe community-acquired pneumonia: development of an international prediction rule. Crit Care (London) 2009, 13: R54. 10.1186/cc7781View ArticleGoogle Scholar
  13. Worster A, Carpenter C: Incorporation bias in studies of diagnostic tests: how to avoid being biased about bias. Canadian J Emerg Med 2008, 10: 174-175.Google Scholar
  14. Hayden JA, Cote P, Bombardier C: Evaluation of the quality of prognosis studies in systematic reviews. Ann Intern Med 2006, 144: 427-437.PubMedView ArticleGoogle Scholar
  15. DerSimonian R, Laird N: Meta-analysis in clinical trials. Control Clin Trials 1986, 7: 177-188. 10.1016/0197-2456(86)90046-2PubMedView ArticleGoogle Scholar
  16. Higgins JP, Thompson SG, Deeks JJ, Altman DG: Measuring inconsistency in meta-analyses. BMJ 2003, 327: 557-560. 10.1136/bmj.327.7414.557PubMedPubMed CentralView ArticleGoogle Scholar
  17. Moses LE, Shapiro D, Littenberg B: Combining independent studies of a diagnostic test into a summary ROC curve: data-analytic approaches and some additional considerations. Stat Med 1993, 12: 1293-1316.PubMedView ArticleGoogle Scholar
  18. Lamy O, Van Melle G, Cornuz J, Burnand B: Clinical management of immunocompetent hospitalized patients with community-acquired pneumonia. Eur J Intern Med 2004, 15: 28-34. 10.1016/j.ejim.2003.11.003PubMedView ArticleGoogle Scholar
  19. Ewig S, Seifert K, Kleinfeld T, Goke N, Schafer H: Management of patients with community-acquired pneumonia in a primary care hospital: a critical evaluation. Respir Med 2000, 94: 556-563. 10.1053/rmed.1999.0775PubMedView ArticleGoogle Scholar
  20. Neill AM, Martin IR, Weir R, Anderson R, Chereshsky A, Epton MJ, Jackson R, Schousboe M, Frampton C, Hutton S, Chambers ST, Town GI: Community acquired pneumonia: aetiology and usefulness of severity criteria on admission. Thorax 1996, 51: 1010-1016. 10.1136/thx.51.10.1010PubMedPubMed CentralView ArticleGoogle Scholar
  21. Ananda-Rajah MR, Charles PGP, Melvani S, Burrell LL, Johnson PDR, Grayson ML: Comparing the pneumonia severity index with CURB-65 in patients admitted with community acquired pneumonia. Scand J Infect Dis 2008, 40: 293-300. 10.1080/00365540701663381PubMedView ArticleGoogle Scholar
  22. Angus DC, Marrie TJ, Scott Obrosky D, Clermont G, Dremsizov TT, Coley C, Fine MJ, Singer DE, Kapoor WN: Severe community-acquired pneumonia: use of intensive care services and evaluation of American and British Thoracic Society diagnostic criteria. Am J Respir Crit Care Med 2002, 166: 717-723. 10.1164/rccm.2102084PubMedView ArticleGoogle Scholar
  23. Brown SM, Jones BE, Jephson AR, Dean NC: Validation of the Infectious Disease Society of America/American Thoracic Society 2007 guidelines for severe community-acquired pneumonia. Crit Care Med 2009, 37: 3010-3016. 10.1097/CCM.0b013e3181b030d9PubMedPubMed CentralView ArticleGoogle Scholar
  24. Buising KL, Thursky KA, Black JF, MacGregor L, Street AC, Kennedy MP, Brown GV: A prospective comparison of severity scores for identifying patients with severe community acquired pneumonia: reconsidering what is meant by severe pneumonia. Thorax 2006, 61: 419-424. 10.1136/thx.2005.051326PubMedPubMed CentralView ArticleGoogle Scholar
  25. Buising KL, Thursky KA, Black JF, MacGregor L, Street AC, Kennedy MP, Brown GV: Identifying severe community-acquired pneumonia in the emergency department: a simple clinical prediction tool. Emerg Med Australas 2007, 19: 418-426. 10.1111/j.1742-6723.2007.01003.xPubMedView ArticleGoogle Scholar
  26. Calbo E, Ochoa de Echaguen A, Rodriguez-Carballeira M, Ferrer C, Garau J: Hospital admission, duration of stay and mortality in community-acquired pneumonia in an acute care hospital: correlation between a pneumonia prognosis index and conventional clinical criteria for assessing severity. Enferm Infecc Microbiol Clin 2004, 22: 64-69. 10.1157/13056884PubMedView ArticleGoogle Scholar
  27. Capelastegui A, Espana PP, Quintana JM, Areitio I, Gorordo I, Egurrola M, Bilbao A: Validation of a predictive rule for the management of community-acquired pneumonia. Eur Respir J 2006, 27: 151-157. 10.1183/09031936.06.00062505PubMedView ArticleGoogle Scholar
  28. Chalmers JD, Singanayagam A, Hill AT: Systolic blood pressure is superior to other haemodynamic predictors of outcome in community acquired pneumonia. Thorax 2008, 63: 698-702. 10.1136/thx.2008.095562PubMedView ArticleGoogle Scholar
  29. Davis JS, Cross GB, Charles PG, Currie BJ, Anstey NM, Cheng AC: Pneumonia risk stratification in tropical Australia: does the SMART-COP score apply? Med J Aust 2010, 192: 133-136.PubMedGoogle Scholar
  30. Escobar GJ, Fireman BH, Palen TE, Gardner MN, Lee JY, Clark MP, Kipnis P: Risk adjusting community-acquired pneumonia hospital outcomes using automated databases. Am J Manag Care 2008, 14: 158-166.PubMedGoogle Scholar
  31. Espana PP, Capelastegui A, Quintana JM, Bilbao A, Diez R, Pascual S, Esteban C, Zalacain R, Menendez R, Torres A: Validation and comparison of SCAP as a predictive score for identifying low-risk patients in community-acquired pneumonia. J Infect 2010, 60: 106-113. 10.1016/j.jinf.2009.11.013PubMedView ArticleGoogle Scholar
  32. Ewig S, de Roux A, Bauer T, Garcia E, Mensa J, Niederman M, Torres A: Validation of predictive rules and indices of severity for community acquired pneumonia. Thorax 2004, 59: 421-427. 10.1136/thx.2003.008110PubMedPubMed CentralView ArticleGoogle Scholar
  33. Ewig S, Ruiz M, Mensa J, Marcos MA, Martinez JA, Arancibia F, Niederman MS, Torres A: Severe community-acquired pneumonia: assessment of severity criteria. Am J Respir Crit Care Med 1998, 158: 1102-1108.PubMedView ArticleGoogle Scholar
  34. Feagan BG, Marrie TJ, Lau CY, Wheeler SL, Wong CJ, Vandervoort MK: Treatment and outcomes of community-acquired pneumonia at Canadian hospitals. CMAJ 2000, 162: 1415-1420.PubMedPubMed CentralGoogle Scholar
  35. Frei CR, Burgess DS: Community-acquired pneumonia: objective criteria to predict severe disease. J Infect Dis Pharmacother 2004, 6: 39-48.View ArticleGoogle Scholar
  36. Garau J, Baquero F, Perez-Trallero E, Perez JL, Martin-Sanchez AM, Garcia-Rey C, Martin-Herrero JE, Dal-Re R: Factors impacting on length of stay and mortality of community-acquired pneumonia. Clin Microbiol Infect 2008, 14: 322-329. 10.1111/j.1469-0691.2007.01915.xPubMedView ArticleGoogle Scholar
  37. Garcia-Vidal C, Fernandez-Sabe N, Carratala J, Diaz V, Verdaguer R, Dorca J, Manresa F, Gudiol F: Early mortality in patients with community-acquired pneumonia: causes and risk factors. Eur Respir J 2008, 32: 733-739. 10.1183/09031936.00128107PubMedView ArticleGoogle Scholar
  38. Kamath A, Pasteur MC, Slade MG, Harrison BD: Recognising severe pneumonia with simple clinical and biochemical measurements. Clin Med 2003, 3: 54-56.View ArticleGoogle Scholar
  39. Liapikou A, Ferrer M, Polverino E, Balasso V, Esperatti M, Piner R, Mensa J, Luque N, Ewig S, Menendez R, Niederman MS, Torres A: Severe community-acquired pneumonia: validation of the Infectious Diseases Society of America/American Thoracic Society guidelines to predict an intensive care unit admission. Clin Infect Dis 2009, 48: 377-385. 10.1086/596307PubMedView ArticleGoogle Scholar
  40. Man SY, Lee N, Ip M, Antonio GE, Chau SS, Mak P, Graham CA, Zhang M, Lui G, Chan PK, Ahuja AT, Hui DS, Sung JJ, Rainer TH: Prospective comparison of three predictive rules for assessing severity of community-acquired pneumonia in Hong Kong. Thorax 2007, 62: 348-353. 10.1136/thx.2006.069740PubMedPubMed CentralView ArticleGoogle Scholar
  41. Marrie TJ, Shariatzadeh MR: Community-acquired pneumonia requiring admission to an intensive care unit: a descriptive study. Medicine (Baltimore) 2007, 86: 103-111. 10.1097/MD.0b013e3180421c16View ArticleGoogle Scholar
  42. Phua J, See KC, Chan YH, Widjaja LS, Aung NW, Ngerng WJ, Lim TK: Validation and clinical implications of the IDSA/ATS minor criteria for severe community-acquired pneumonia. Thorax 2009, 64: 598-603. 10.1136/thx.2009.113795PubMedView ArticleGoogle Scholar
  43. Putinati S, Ballerin L, Piattella M, Ritrovato L, Zabini F, Potena A: Clinical policy for management and risk stratification of community-acquired pneumonia in patients hospitalized on the basis of conventional admission criteria. Rec Progr Med 2003, 94: 199-203.Google Scholar
  44. Renaud B, Coma E, Hayon J, Gurgui M, Longo C, Blancher M, Jouannic I, Betoulle S, Roupie E, Fine MJ: Investigation of the ability of the Pneumonia Severity Index to accurately predict clinically relevant outcomes: a European study. Clin Microbiol Infect 2007, 13: 923-931. 10.1111/j.1469-0691.2007.01772.xPubMedView ArticleGoogle Scholar
  45. Riley PD, Aronsky D, Dean NC: Validation of the 2001 American Thoracic Society criteria for severe community-acquired pneumonia. Crit Care Med 2004, 32: 2398-2402. 10.1097/01.CCM.0000147443.38234.D2PubMedView ArticleGoogle Scholar
  46. Roson B, Carratala J, Dorca J, Casanova A, Manresa F, Gudiol F: Etiology, reasons for hospitalization, risk classes, and outcomes of community-acquired pneumonia in patients hospitalized on the basis of conventional admission criteria. Clin Infect Dis 2001, 33: 158-165. 10.1086/321808PubMedView ArticleGoogle Scholar
  47. Shah BA, Ahmed W, Dhobi GN, Shah NN, Khursheed SQ, Haq I: Validity of pneumonia severity index and CURB-65 severity scoring systems in community acquired pneumonia in an Indian setting. Indian J Chest Dis 2010, 52: 9-17.Google Scholar
  48. van der Eerden MM, de Graaff CS, Bronsveld W, Jansen HM, Boersma WG: Prospective evaluation of pneumonia severity index in hospitalised patients with community-acquired pneumonia. Respir Med 2004, 98: 872-878. 10.1016/j.rmed.2004.02.022PubMedView ArticleGoogle Scholar
  49. Fukuyama H, Ishida T, Tachibana H, Nakagawa H, Iwasaku M, Saigusa M, Yoshioka H, Arita M, Hashimoto T: Validation of scoring systems for predicting severe community-acquired pneumonia. Intern Med 2011, 50: 1917-1922. 10.2169/internalmedicine.50.5279PubMedView ArticleGoogle Scholar
  50. Chalmers JD, Taylor JK, Mandal P, Choudhury G, Singanayagam A, Akram AR, Hill AT: Validation of the Infectious Diseases Society of America/American Thoratic Society Minor Criteria for Intensive Care Unit Admission in Community-Acquired Pneumonia Patients Without Major Criteria or Contraindications to Intensive Care Unit Care. Clin Infect Dis 2011, 53: 503-511. 10.1093/cid/cir463PubMedView ArticleGoogle Scholar
  51. Putinati S, Ballerin L, Zabini F, Piattella M, Ritrovato L, Simoni M, Potena A: Predictive rules for prognosis in community-acquired pneumonia. Minerva Pneumol 2005, 44: 85-91.Google Scholar
  52. Renaud B, Coma E, Labarere J, Hayon J, Roy PM, Boureaux H, Moritz F, Cibien JF, Guerin T, Carre E, Lafontaine A, Bertrand MP, Santin A, Brun-Buisson C, Fine MJ, Roupie E: Routine use of the Pneumonia Severity Index for guiding the site-of-treatment decision of patients with pneumonia in the emergency department: a multicenter, prospective, observational, controlled cohort study. Clin Infect Dis 2007, 44: 41-49. 10.1086/509331PubMedView ArticleGoogle Scholar
  53. Chalmers JD, Singanayagam A, Akram AR, Mandal P, Short PM, Choudhury G, Wood V, Hill AT: Severity assessment tools for predicting mortality in hospitalised patients with community-acquired pneumonia: systematic review and meta-analysis. Thorax 2010, 65: 878-883. 10.1136/thx.2009.133280PubMedView ArticleGoogle Scholar
  54. Schuetz P, Suter-Widmer I, Chaudri A, Christ-Crain M, Zimmerli W, Mueller B: Prognostic value of procalcitonin in community-acquired pneumonia. Eur Respir J 2011, 37: 384-392. 10.1183/09031936.00035610PubMedView ArticleGoogle Scholar
  55. Schuetz P, Christ-Crain M, Zimmerli W, Mueller B: Repeated measurements of endothelin-1 precursor peptides predict the outcome in community-acquired pneumonia. Intensive Care Med 2011, 37: 970-980. 10.1007/s00134-011-2208-2PubMedView ArticleGoogle Scholar
  56. Kruger S, Ewig S, Kunde J, Hanschmann A, Marre R, Suttorp N, Welte T: C-terminal provasopressin (copeptin) in patients with community-acquired pneumonia: influence of antibiotic pre-treatment: results from the German competence network CAPNETZ. J Antimicrob Chemother 2009, 64: 159-162. 10.1093/jac/dkp148PubMedView ArticleGoogle Scholar
  57. Christ-Crain M, Breidthardt T, Stolz D, Zobrist K, Bingisser R, Miedinger D, Leuppi J, Tamm M, Mueller B, Mueller C: Use of B-type natriuretic peptide in the risk stratification of community-acquired pneumonia. J Intern Med 2008, 264: 166-176. 10.1111/j.1365-2796.2008.01934.xPubMedView ArticleGoogle Scholar
  58. Kruger S, Ewig S, Kunde J, Hartmann O, Suttorp N, Welte T: Pro-atrial natriuretic peptide and pro-vasopressin for predicting short-term and long-term survival in community-acquired pneumonia: results from the German Competence Network CAPNETZ. Thorax 2010, 65: 208-214. 10.1136/thx.2009.121178PubMedView ArticleGoogle Scholar
  59. Christ-Crain M, Morgenthaler NG, Stolz D, Muller C, Bingisser R, Harbarth S, Tamm M, Struck J, Bergmann A, Muller B: Pro-adrenomedullin to predict severity and outcome in community-acquired pneumonia [ISRCTN04176397]. Crit Care 2006, 10: R96. 10.1186/cc4955PubMedPubMed CentralView ArticleGoogle Scholar
  60. Chalmers JD, Mandal P, Singanayagam A, Akram AR, Choudhury G, Short PM, Hill AT: Severity assessment tools to guide ICU admission in community-acquired pneumonia: systematic review and meta-analysis. Intensive Care Med 2011, 37: 1409-1420. 10.1007/s00134-011-2261-xPubMedView ArticleGoogle Scholar
  61. Spindler C, Ortqvist A: Prognostic score systems and community-acquired bacteraemic pneumococcal pneumonia. Eur Respir J 2006, 28: 816-823. 10.1183/09031936.06.00144605PubMedView ArticleGoogle Scholar
  62. Kontou P, Kuti JL, Nicolau DP: Validation of the Infectious Diseases Society of America/American Thoracic Society criteria to predict severe community-acquired pneumonia caused by Streptococcus pneumoniae . Am J Emerg Med 2009, 27: 968-974. 10.1016/j.ajem.2008.07.037PubMedView ArticleGoogle Scholar
  63. Feldman C, Alanee S, Yu VL, Richards GA, Ortqvist A, Rello J, Chiou CC, Chedid MB, Wagener MM, Klugman KP: Severity of illness scoring systems in patients with bacteraemic pneumococcal pneumonia: implications for the intensive care unit care. Clin Microbiol Infect 2009, 15: 850-857. 10.1111/j.1469-0691.2009.02901.xPubMedView ArticleGoogle Scholar

Copyright

© Marti et al.; licensee BioMed Central Ltd. 2012

This article is published under license to BioMed Central Ltd. This is an open access article distributed under the terms of the Creative Commons Attribution License (http://creativecommons.org/licenses/by/2.0), which permits unrestricted use, distribution, and reproduction in any medium, provided the original work is properly cited.